Default PhiX spike-in issue

Hi all,

I have come across issues with the Illumina PhiX v3 in my project. I need to sequence my low diverse library on NextSeq 500. To balance out the diversity, I tried to spike-in 50% of the Illumina PhiX v3 control along with my library. But, the run wasn't good and I only managed to see a 9% alignment in my first run. At first, I thought it was resulted from the degradation of my PhiX. Thereby, I quantified my PhiX using picrogreen and the reading seems fine after conversion and the bioanalyzer showed no sign of degradation. I repeated my run the 2nd time. This time the Illumina FAS came over and supervised my library denaturation method to ensure I followed exactly the instructions recommended in the protocol, ie. library denaturation->library dilution to 1.8pM followed by PhiX denaturation->dilution of denatured PhiX to 1.8pm and, eventually, mixed both library and PhiX at 50:50 ratio. Everything was fine but the results of the 2nd run was still the same as in we couldn't achieved near 50% of PhiX spike-in. In addition, we observed a high error rate too. Does anyone have any suggestion on this issue? Thank you in advance for the guidance.
edwardwong1070584 is offline   Reply With Quote
Old 10-16-2016, 12:50 PM   #2
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,238
Default

Three issues to consider:
1- your library and PhiX
2- sequencing chemistry
3- sequencing hardware

For correct diagnosis you need to provide some information from runs SAV metrics such as cluster density, PF%, % aligned to PhiX, Q30 scores
nucacidhunter is offline   Reply With Quote
Old 10-16-2016, 05:56 PM   #3
edwardwong1070584
Junior Member
 
Location: Singapore

Join Date: Aug 2016
Posts: 5
Default

Hi @nucacidhunter,

Thank you for your reply. Please see below for the metrics. Thank you in advance for your guidance.
Attached Images
File Type: png Run Metrics 1.PNG (49.4 KB, 36 views)
File Type: png Run Metrics 2.PNG (43.4 KB, 13 views)
edwardwong1070584 is offline   Reply With Quote
Old 10-16-2016, 07:48 PM   #4
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,238
Default

The issues that is apparent:

1- Low cluster density that could be due to errors in quantification, denaturation or loading resulting in under loading or overloading (over clustering where several cluster are identified as a single cluster). Over clustering possibility can be checked by looking at images and also selecting intensity in Flow Cell Chart from analysis page which should show higher values (yellowish) in upper tiles in comparison to lower ones. In standard clustering lower tiles has higher density than the upper ones.

2- PF% very low: indicates low diversity (as you have mentioned) that has not been improved by PhiX spike-in because PhiX% is not high enough (this can be checked by looking at %Base in Data By Cycle plot in analysis tab).

3- High error rate: could be result of over clustering or poor sequencing primer compatibility (if custom primer was used)

4- If you can try quantifying PhiX by qPCR

I hope this and looking to similarities in second run helps you to identify the issue. If you need more help please post images from those analysis page plots.
nucacidhunter is offline   Reply With Quote
Old 10-16-2016, 11:54 PM   #5
edwardwong1070584
Junior Member
 
Location: Singapore

Join Date: Aug 2016
Posts: 5
Default

Thank you very much for your reply, @nucacidhunter. The Illumina FAS is currently checking through the run metrics. He is also suspecting that the library sample itself might be overclustered, resulting in the template generation to perform poorly and also exlaining on why the alignment didn't make sense. We had actually run the 3rd time, in which we lowered the library sample loading concentration from 1.8 to 1.5pM and it did improve the cluster density. But, the error rate was still high.

We did quantify our PhiX and the concentration was fine. So I guess we could omit the possibility of the degraded PhiX being used. I have attached the plot of data by cycle for this run for your reference.

Thank you
Attached Files
File Type: pdf Data by cycle_pdf.pdf (440.2 KB, 32 views)
edwardwong1070584 is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:44 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO