  • Illumina NextSeq shorter fragment size than expected

    Hi,

    We were preparing library for NextSeq whole exome sequencing. Our starting concentration before library preparation for three samples were 27ng/µL, 37ng/µL and 42ng/µL. We pooled library and when we checked our fragment size before running on sequencer by Bioanalyzer, we saw fragment size of 239 base pair which is smaller than expected. I want to know on which step we are making mistake which resulted in smaller insert size than expected?

  • #2
    If you can post Bioanalyser result then it will be easier to diagnose the issue.

    Comment


    • #3
      Bioanalyzer image

      https://drive.google.com/file/d/0ByA...ew?usp=sharingWe have repeated procedure of tagmentation and we have got same results of shorter fragment length
      This is the image of bioanalyzer peak.

      Comment


      • #4
        The fragment size distribution is within optimal for sequence capture if that is what you are doing. To obtain larger fragments (if you are using Nextera not Nextera flex) input DNA need to be increased. If you are comparing the library size with Illumina user guide the difference could be also due to Bioanlyser assay (note Illumina have used HS DNA Chip).

        Comment


        • #5
          This is size of library after adding adapters. Is 200-400bp normal? This will affect concentration or not?

          Comment


          • #6
            Which library prep and capture kit are you using?

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Current Approaches to Protein Sequencing
              by seqadmin


              Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
              04-04-2024, 04:25 PM
            • seqadmin
              Strategies for Sequencing Challenging Samples
              by seqadmin


              Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
              03-22-2024, 06:39 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 04-11-2024, 12:08 PM
            0 responses
            18 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 10:19 PM
            0 responses
            22 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 09:21 AM
            0 responses
            17 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-04-2024, 09:00 AM
            0 responses
            49 views
            0 likes
            Last Post seqadmin  
            Working...
            X