Default MiSeq Failed Run

Hello,

I got this:

"Your run xxxxxxxx on instrument xxxxxxx recently encountered the following error during sequencing:

<Internal Illumina Command Exception>"

Any ideas?

Thanks
DRYTCYV is offline   Reply With Quote
Old 12-18-2015, 04:59 PM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,978
Default

Don't mean to state the obvious but have you contacted tech support after taking a screen shot?
GenoMax is offline   Reply With Quote
Old 12-22-2015, 02:53 AM   #3
DRYTCYV
Member
 
Location: Portugal

Join Date: Apr 2011
Posts: 50
Default

Quote:
Originally Posted by GenoMax View Post
Don't mean to state the obvious but have you contacted tech support after taking a screen shot?
MiSeq Run started: 20:30 (Friday)
Error msg : 21:30

Who to call?

A) Illumina techsupport
B) SEQanswers amazing tech support (and sometimes the only tech support)
C) Ghostbusters

The correct option : B

The error msg was just from the e-mail, when something bad happened, but in the machine I got this one:

"No usable signal found in the images; it is possible that clustering has failed"

Possible Cause: NaOH concentration too high
DRYTCYV is offline   Reply With Quote
Old 12-22-2015, 06:26 AM   #4
Jessica_L
Senior Member
 
Location: Washington, D.C. metro area

Join Date: Feb 2010
Posts: 118
Default

If you spiked in PhiX and you're relatively certain your libraries should have been fine, I think NaOH is at least worth a look. Do you normally check the pH of your 0.2N NaOH before using it to denature? I used to not do it and had so many runs fail because of bad NaOH from Illumina (came in our HiSeq kits) that I started checking it every single time.

Illumina tech support will ask you to run some of the machine tests (volumetric tests, etc.) so you can do some of that preemptively. I'd still contact Illumina because you should be able to get a replacement kit if everything else checks out.

I've seen cluster generation failures when there's just not enough signal (i.e. the loading concentration of the library is too low) so it might be worth double checking any quantitation measurements that were made, too.
Jessica_L is offline   Reply With Quote
Old 12-22-2015, 06:27 AM   #5
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 6,978
Default

So was that the cause of the failure? It sounds like you were able to re-run with correct concentration of NaOH?
GenoMax is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 11:45 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO