Old 09-02-2011, 06:30 AM   #1
hoisinjl
Member
 
Location: St. Louis, MO

Join Date: Jun 2009
Posts: 23
Default Loss of tiles with v3 chemistry

Ever since we have switched to v3 chemistry, we have noticed that we are loosing tiles randomly during imaging. There seem to be 3 specific phenotypes:

1. Loosing a few tiles for a cycle, and then it recovers. This can happen only once during the runs or several times.
2. Loosing the same tile many tiles during the run, causing the tile not to be analyzed at all, reducing the pass filter rate.
3. Loosing an entire surface (top or bottom) for a cycle.

Has anyone else seen this? The bottom, middle surface (8 tiles) seems to be the most common to be lost, but have seen it in other tiles.
hoisinjl is offline   Reply With Quote
Old 09-02-2011, 09:57 AM   #2
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,177
Default

We have just recently had our HiSeq installed and during the test run I noted a number of phenotype 3 occurrences. They were always the bottom surface, swath #2 (middle); they occurred in several different lanes (6 of the 8) and from cycles throughout the run (it was a 2x100 run). In all there were 25 of these bad scans over course of our test run. I haven't been able to talk to our FAS yet about this to determine if this frequency is normal.

During the run our FAS explained that this can occur if there is a bubble near the inlet port of the lane. This causes a focus error at the beginning of the scan so the entire scan for that swath will be out of focus. It makes sense that this would be more likely for scans of the bottom surface since it would be trying to focus through the bubble. I'm not sure why they were uniformly in the middle swath.

I'm worried about the frequency we observed. Considering that a bad swath blows 16.7% of the clusters in that lane for that cycle that's huge. Have that happen a couple of times in the same lane/swath and before you know it you've lost 1/6th of your reads from that lane for having too many N's.
kmcarr is offline   Reply With Quote
Old 09-02-2011, 11:40 AM   #3
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,059
Default

We have several HiSeq's and have very rarely had problems with dropped tiles. It sounds like a visit from an illumina engineer may be needed to check your HiSeq.
GenoMax is offline   Reply With Quote
Old 09-04-2011, 09:12 AM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

We saw dropped tiles of this sort in our HiScanSQ prior to complete blockages in some of our lines solved by a few service calls to iron out fluidics issues. I have heard these are endemic to new HiSeqs and HiScanSQs, but not old ones. Since then I mainly see dropped tiles in bar code cycles where low numbers of indexes are in the lane.

Strangely, some of the dropped tiles did not seem to have focus problems -- the clusters were quite clear.

--
Phillip
pmiguel is offline   Reply With Quote
Old 09-12-2011, 10:27 AM   #5
GW_OK
Senior Member
 
Location: Oklahoma

Join Date: Sep 2009
Posts: 411
Default

Well I've been on the phone with tech support about this for two runs now. There's an actual name for the phenotype: "BMS" (Bottom Middle Swath). You can call up tech support and say "I've got BMS" and they'll know what you're talking about. I've been informed that Illumina has their best people on it, but as long as you get "enough" data it's going to be a shoulder-shrugging issue.

Reminds me a little of the v1 hiseq flowcells...
GW_OK is offline   Reply With Quote
Old 09-12-2011, 11:25 AM   #6
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,177
Default

Quote:
Originally Posted by GW_OK View Post
Well I've been on the phone with tech support about this for two runs now. There's an actual name for the phenotype: "BMS" (Bottom Middle Swath). You can call up tech support and say "I've got BMS" and they'll know what you're talking about. I've been informed that Illumina has their best people on it, but as long as you get "enough" data it's going to be a shoulder-shrugging issue.

Reminds me a little of the v1 hiseq flowcells...
Thanks for the report GW_OK. It's nice to know we're not the only ones.
kmcarr is offline   Reply With Quote
Old 09-13-2011, 07:23 AM   #7
hoisinjl
Member
 
Location: St. Louis, MO

Join Date: Jun 2009
Posts: 23
Default

We have figured out that for a tile not to be completed analyzed, you have to loose focus twice in the first 24 tiles (i.e. before the chastity filter). We just had the y-axis on our HiSeqs adjusted so that the imaging is more towards the back of the flowcell. We are monitoring the next couple of runs closely to see if we see a decrease in the number of occurrences. I will update when I know more.
hoisinjl is offline   Reply With Quote
Old 09-13-2011, 12:58 PM   #8
josdegraaf
Member
 
Location: Germany

Join Date: Mar 2010
Posts: 33
Default

Same here, is indeed a known problem due to bubbles in the lanes during scanning. Maybe due to bigger lanes?
josdegraaf is offline   Reply With Quote
Old 10-05-2011, 05:26 AM   #9
sehrrot
Member
 
Location: USA

Join Date: Jul 2010
Posts: 58
Default

Yes it's due to the physically wider lane, which introduce more bubbles. When I first ran V3 fc, I changed gaskets 10 times. But still substantial bubbles. We also found some phenotype issues (spikes) but recovered after few cycles (but luckily we didn't lose any data and got over 300Gb output).

This issue only happens in bottom middle surface, as Illumina admits. Hope it will be fixed soon. BTW at this moment we can do is only removing bubbles as much as we possible. Also, for the first time V3 user, do 1 maintenence wash + 2-3 times water washes + double priming before you start actual run. It will remove and make V3 environment in your Hiseq syringes (remove all V2 reagents in there)
sehrrot is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:11 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO