Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Illumina MiSeq-run: Bad quality per tile. How should i interpret this fastqc-image?

    I recently got very bad sequencing results from a Miseq Illumina run. The average Sequence qualities declined very early, leading to low read lengths after quality clipping.

    When I looked at the respective FastQC-results i noticed a very strange pattern in the "Quality per tile"-Plot (see attached .png file):

    I get less than average quality for cycles 80-280 almost exclusively at tiles 1112-1119 and then again at tiles 2115-2119.
    What I find extremely strange is the apparent "smear" of bad quality from tiles 1112-1119, which then seems to stop abruptly. I do not quite understand what is up with tiles 1120-2100: Are they simply not shown in this plot, or should I assume that Tiles 1119 and 2101 are neighbors?

    usually when I see examples of bad quality-per-tile patterns, bad quality tiles seem to be more distributed along the flow cell and have consistently bad quality during all cycles (e.g. this example).
    In my case, however i seem to have a double smear-pattern of temporarily bad quality tiles, broadly around cycle 175.

    Does anybody have experience in interpreting such plots?
    I would like to understand what the tile-designations "1101"-"2119" mean, exactly.
    Could something like bubbles have formed themselves at cycle ~80 and then just burst sometime around cycle 280?
    In that case, how can it be that this pattern seems to occur twice on the flow cell around roughly the same time? Pure coincidence?
    Attached Files

  • #2
    Those tiles are on opposite sides of the flow cell (the first number in the tile denotes whether it's on the top or bottom). So I suppose it could be a bubble, since that might affect the camera focusing on the other side. Having said that, the core facility and Illumina can probably give you a better answer, since they deal with the machines daily.

    Comment


    • #3
      Thanks for the reply. So as i understand it, the FASTQC-plot shows two opposide sides of my flow cell in part only, in my case?

      I'll also try the illumina customer support. However, at least from my previous experience, you often do not get much info on what actually caused any reported problem and/or why. Often you just simply get a replacement for the reagents if Illumina thinks there might be a problem with the reagents or the machine, in the hope that it doesn't occur again next time.

      Therfore I thought I'd ask the community first for detailed infos suggestions. Not sure if i'm looking at a problem caused by (e.g.) a faulty flow cell or by something that could have been prevented during sample preperation/loading.

      Comment


      • #4
        Originally posted by jov14 View Post
        Thanks for the reply. So as i understand it, the FASTQC-plot shows two opposide sides of my flow cell in part only, in my case?
        The tile numbers are of a format:

        XYZZ

        Where "X" is the surface of the flowcell -- 1= top, 2=bottom;
        "Y" is the swath number. MiSeqs only have a single swath, so you can ignore this position;
        "ZZ" is the tile -- should be numbered "01" at one end of the flowcell, and "19" at the other end (for a v2 MiSeq flowcell).

        So it looks like your issue occurs mid-run, impacting one end of your flowcell. Both surfaces are affected, but the upper surface, more than the bottom.

        I have no idea what would cause this. A bubble in the flowcell during scanning should cause more issues for the bottom surface as there will be issues focusing through the bubble.

        How do the thumbnail images of the tiles look during the low quality cycles?

        --
        Phillip

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM
        • seqadmin
          Techniques and Challenges in Conservation Genomics
          by seqadmin



          The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

          Avian Conservation
          Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
          03-08-2024, 10:41 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, Yesterday, 06:37 PM
        0 responses
        10 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, Yesterday, 06:07 PM
        0 responses
        9 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 03-22-2024, 10:03 AM
        0 responses
        50 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 03-21-2024, 07:32 AM
        0 responses
        67 views
        0 likes
        Last Post seqadmin  
        Working...
        X