  • MiSeq still underclustered at 18 pM

    Hello dear people,
    We are doing some whole genome sequencing on a plant genome using the MiSeq with v3 reagents for 2x250 paired-end. Library size was measured using a Bioanalyzer and concentration by Qubit.

    On the first run, we loaded 15pM library (Nextera), spiked with 1% phiX. We got nice quality (88% >Q30), but a fairly low clustering density (639k/mm2), resulting in 7.3 Gbp of non-index yield.
    Surprisingly, the metrics showed 2.3% ‘aligned’ reads, though only 1% should be phiX.

    Our second MiSeq run used the same library as before. We tried 18pM but went down to 0.5% phiX. The clustering density has increased a bit, but is still on the low side (738k/mm2), quality has not dropped (it’s about 91% >Q30), and we got 8.26 Gbp non-index reads.

    This makes me wonder whether we are somehow overestimating the effective library concentration due to either underestimating average fragment length (I had quite a broad distribution) or incomplete adapter ligation, causing some fragments to wash off without binding to the flowcell.

    In the second run, ‘aligned’ reads are still 1.3% despite using 0.5% phiX. Does that mean the effective library concentration that is being sequenced isn’t actually 18pM, but more like 6.9 pM (because 0.5% of 18 pM = 1.3% of 6.9pM) and we should go with a significantly higher concentration?

    If with v3 reagents we are aiming for a cluster density of 1200-1400, and 12-15 Gbp of yield, are we seriously underclustering and could get much more out of one run by nearly doubling the library concentration? Is there any downside to loading (supposedly) 30pM if more than half of it washes off the flowcell anyway?

    Would love to hear what you think.

  • #2
    Update:

    We went ahead with loading 30pM, by using 7.5 uL of 4 nM library and 2.5 ul 0.4N NaOH and then proceeding with the rest of the protocol as if we were handling 20 pM.

    The run worked very well, giving us 1152 k/mm2 cluster density, with 87.3% passing the filter. The yield was 11.26 Gbp with 87.5% > Q30, which is barely lower than the quality we had before.

    TL;DR:
    If your library adapter ligation seems to be incomplete, loading 20 pM on a MiSeq v3 should work fine to compensate.

    Comment


    • #3
      Try to measure concentration through qPCR targeting the p7/p5 adapters. This will give you the a good measurement of the concentration that can amplify on the flowcell. You can a kit from NEB for this if you want.

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Techniques and Challenges in Conservation Genomics
        by seqadmin



        The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

        Avian Conservation
        Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
        03-08-2024, 10:41 AM
      • seqadmin
        The Impact of AI in Genomic Medicine
        by seqadmin



        Artificial intelligence (AI) has evolved from a futuristic vision to a mainstream technology, highlighted by the introduction of tools like OpenAI's ChatGPT and Google's Gemini. In recent years, AI has become increasingly integrated into the field of genomics. This integration has enabled new scientific discoveries while simultaneously raising important ethical questions1. Interviews with two researchers at the center of this intersection provide insightful perspectives into...
        02-26-2024, 02:07 PM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 03-14-2024, 06:13 AM
      0 responses
      32 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 03-08-2024, 08:03 AM
      0 responses
      72 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 03-07-2024, 08:13 AM
      0 responses
      80 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 03-06-2024, 09:51 AM
      0 responses
      68 views
      0 likes
      Last Post seqadmin  
      Working...
      X