#1 |
Junior Member
Location: Germany Join Date: Jan 2014
Posts: 5
|
![]()
Hello,
in some of my prepared sequencing libraries I observe multiple peaks in the bioanalyzer run after library pcr (see attached file). Many of the peaks have an average distance of arround 60 bp. Are these adaptor concatameres or something like this? Is there any way to avoid these peaks? I know the peak at 126 bp are adaptor dimers... But this is a different story. Best Harry |
![]() |
![]() |
![]() |
#2 |
Senior Member
Location: USA Join Date: Apr 2009
Posts: 482
|
![]()
That is the strangest library I've ever seen. What type of library is this?
|
![]() |
![]() |
![]() |
#3 |
Junior Member
Location: Germany Join Date: Jan 2014
Posts: 5
|
![]()
It's a ChIP-Seq library prepared with Illumina's TrueSeq ChIP Prep kit - prepared according to the protocol. I have it in both input samples and IP samples. However, from around 50 samples it happened only in 6.
My first idea was that it depends on the initial amount of DNA used for library preparation, but there is absolutely no correlation (it happend both for the maximal amount 10 ng and for a lower input amount 5-6 ng). The bioanalyzer traces are reproducible, excluding problems with the run. I personally can not believe that these are adaptor concatamers - since these peaks appear up to 1800 bp (would basically mean 30 adaptors ligated...). |
![]() |
![]() |
![]() |
#4 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Did you happen to use the spike-in standards for library construction? Those can give some strange patterns.
-- Phillip |
![]() |
![]() |
![]() |
#5 |
Junior Member
Location: Germany Join Date: Jan 2014
Posts: 5
|
![]()
No, nothing was added during library preparation.
Samples were prepared in parallel from the same fragmented chromatin and only few of them showed this pattern. All other libraries look fine. It also does not depend on particular adaptors. Could these peaks be the result of contaminations in the samples? |
![]() |
![]() |
![]() |
#6 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
-- Phillip |
|
![]() |
![]() |
![]() |
#7 |
Junior Member
Location: Germany Join Date: Jan 2014
Posts: 5
|
![]()
No the gel purification was performed after the ligation step. The strange samples were not near the markers (50 bp ladder) on the gels.
|
![]() |
![]() |
![]() |
#8 |
Member
Location: Nottingham, England Join Date: Nov 2009
Posts: 26
|
![]()
Silly question first. Have you run them on the chip a second time? It almost looks like bleed through from the ladder to be honest. Adapter concatamers is also possible but we have done a lot of chip preps and I have never seen this.
Last edited by barrmur; 01-28-2014 at 12:50 PM. |
![]() |
![]() |
![]() |
#9 |
Junior Member
Location: Germany Join Date: Jan 2014
Posts: 5
|
![]()
Yes, the pattern is absolutely reproducible and not dependent on a particular position on the chip or anything like this.
|
![]() |
![]() |
![]() |
#10 |
Senior Member
Location: USA Join Date: Apr 2009
Posts: 482
|
![]()
The 126 bp peak is an adapter dimer. The other peaks are likely concatamers of the adapter dimer.
I wouldn't bother sequencing this library. I would just remake it. |
![]() |
![]() |
![]() |
#11 |
Junior Member
Location: Hong Kong Join Date: Nov 2014
Posts: 3
|
![]()
how about sequencing the sample??
Were there any strange result or it is just a problem on bioanalyzer?? |
![]() |
![]() |
![]() |