Old 08-12-2017, 08:08 AM   #1
patkrat
Member
 
Location: London

Join Date: Apr 2017
Posts: 13
Default Role/Importance of adapter contamination in cDNA libraries

I have created cDNA libraries from Drosophila melanogaster mRNA using the KAPA mRNA HyperPrep kit. The size distribution on the Bioanalyzer looks good, as I was expecting a mean of ~320 bp. I attached a screenshot of an example trace for one of my samples: My question concerns the two peaks very close to the upper marker (left-hand side): Are these adapter dimers (probably not, given their small size), or left-over primers? If not, what could they be? And, would this pose a problem for sequencing, given how much of them is present compared to the cDNA fragments? Thanks for any help!
Attached Images
File Type: png Screen Shot 2017-08-12 at 15.54.48.png (78.0 KB, 24 views)

Last edited by patkrat; 08-12-2017 at 09:03 AM.
patkrat is offline   Reply With Quote
Old 08-12-2017, 09:12 AM   #2
jdk787
josh kinman
 
Location: Austin

Join Date: Apr 2014
Posts: 57
Default

Those peaks are most likely leftover primers, which should not be a problem since they won't cluster.
__________________
Josh Kinman
jdk787 is offline   Reply With Quote
Old 08-12-2017, 01:10 PM   #3
patkrat
Member
 
Location: London

Join Date: Apr 2017
Posts: 13
Default

Quote:
Originally Posted by jdk787 View Post
Those peaks are most likely leftover primers, which should not be a problem since they won't cluster.
Thanks, Josh. So, I could increase the cycle number during library amplification by 1 or 2, and see whether the peaks disappear.
patkrat is offline   Reply With Quote
Old 08-12-2017, 09:26 PM   #4
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,074
Default

I wonder how you have cleaned up the library. Those peaks should not remain after clean up and they can reduce cluster density/read number by blocking flow cell oligos.
nucacidhunter is offline   Reply With Quote
Old 08-13-2017, 03:50 AM   #5
patkrat
Member
 
Location: London

Join Date: Apr 2017
Posts: 13
Default

Quote:
Originally Posted by nucacidhunter View Post
I wonder how you have cleaned up the library. Those peaks should not remain after clean up and they can reduce cluster density/read number by blocking flow cell oligos.
I do a 1X bead-based cleanup with KAPA Pure Beads.
patkrat is offline   Reply With Quote
Old 08-13-2017, 07:37 PM   #6
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,074
Default

1x bead clean up should have removed any fragment below 100 nt. I would suggest to rerun the Bioanalyzer to rule out possibility of issues with Chip run and contact KAPA tech support for their explanation.
nucacidhunter is offline   Reply With Quote
Old 08-16-2017, 04:32 AM   #7
patkrat
Member
 
Location: London

Join Date: Apr 2017
Posts: 13
Default

Quote:
Originally Posted by nucacidhunter View Post
1x bead clean up should have removed any fragment below 100 nt. I would suggest to rerun the Bioanalyzer to rule out possibility of issues with Chip run and contact KAPA tech support for their explanation.
I did re-run the chip and contacted KAPA tech support. The latter suggested making sure after the library amplification to be very stringent with the 1X clean-up; i.e. remove all supernatant, and esp. all EtOH, etc. Now, I did a whole new library prep on the same samples yesterday, and made absolutely sure to follow all those steps, but still see this extra peak at ~38-40 bp. The 50 bp peak seems to be primer-dimer bleed-through that tends to happen with this protocol apparently (even visible in KAPA's tech manual..), but I don't know what these 38-40 bp peaks are. Could they be a chip artefact? I have read that the Bioanalyzer can be off by up to 5 bp at times, but even if that was the case, this wouldn't explain my situation, since the concentration of the 35 bp peak (upper marker) is 125.00 pg/ul. Or maybe I have a false understanding of how this concentration is measured..
patkrat is offline   Reply With Quote
Old 08-16-2017, 09:09 AM   #8
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,074
Default

If those are artifact of Bioanalyser then they should be present in the ladder as well.
nucacidhunter is offline   Reply With Quote
Old 08-21-2017, 12:08 PM   #9
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,237
Default

If you are running these on a non-patterned-flowcell Illumina sequencer, ignore the primer peaks. If you are running them on a patterned-flowcell Illumina sequencer (HiSeq 3000/4000/X and NovaSeq) then do an extra ampure (or whatever small fragment clean-up you have available to you.)

--
Phillip
pmiguel is offline   Reply With Quote
Old 08-23-2017, 11:57 AM   #10
dylanfofylan
Junior Member
 
Location: Austin, TX

Join Date: Dec 2016
Posts: 2
Default

Quote:
Originally Posted by pmiguel View Post
If you are running these on a non-patterned-flowcell Illumina sequencer, ignore the primer peaks. If you are running them on a patterned-flowcell Illumina sequencer (HiSeq 3000/4000/X and NovaSeq) then do an extra ampure (or whatever small fragment clean-up you have available to you.)

--
Phillip
Hello Phillip,
Can you explain your reasoning behind the difference between patterned and non-patterned flowcells?

I can see how primers that contain an index sequence might lead to more index hopping on patterned flowcells, but what would be the issue with these primers that donít contain an index sequence?
I have noticed these primer peaks with Kapa kits before since they use a high concentration of PCR primers to prevent exhaustion, but I have never sequenced the libraries on a patterned flowcell. Now Iím curious to see if this is going to be an issue on the NovaSeq.

Thank You
dylanfofylan is offline   Reply With Quote
Old 08-23-2017, 06:15 PM   #11
luc
Senior Member
 
Location: US

Join Date: Dec 2010
Posts: 305
Default

Hi dylanfofylan,

Have a look at this video: https://www.youtube.com/watch?v=pfZp5Vgsbw0
Please note how the ends of the library molecules are double-stranded when settling onto the flowcell. Any PCR primers will compete with the oligos present in the reagents - and your PCR primers will not be blocked from extension.
luc is offline   Reply With Quote
Reply

Tags
adapter dimer, adapters, bioanalyzer, bioanalyzer 2100, cdna library

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 08:03 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO