Old 12-21-2017, 09:27 AM   #1
evobio721
Junior Member
 
Location: Arkansas

Join Date: Dec 2017
Posts: 5
Default ddRAD Post-PCR Bioanalyzer Has Multiple Peaks

Hello,

I have a problem with a post-PCR ddRAD library that was prepared using Peterson (2012). It was prepared along with another index for sequencing on one Illumina lane, and the pre-PCR size selection for both indices was supposed to be 454-509 bp.

However, the Bioanalyzer results show two smaller peaks for one of the indices that shouldn't be there (see attached images), and the smaller peaks should be outside the size-selection range. Thus, it must have post size selection.

Does anyone know what might have caused the two smaller peaks to occur, and should I worry about it during sequencing?

Thank you,

-Bradley
Attached Images
File Type: jpg L1712-02-I8.jpg (76.4 KB, 16 views)
File Type: jpg L1712-02-I9.jpg (81.5 KB, 17 views)
evobio721 is offline   Reply With Quote
Old 12-21-2017, 10:26 AM   #2
BioKiwi
Member
 
Location: KiwiTeritory

Join Date: Sep 2014
Posts: 19
Default

I wonder how many PCR cycles you have done and what was the total yield.
BioKiwi is offline   Reply With Quote
Old 12-21-2017, 10:31 AM   #3
evobio721
Junior Member
 
Location: Arkansas

Join Date: Dec 2017
Posts: 5
Default

Thank you for your prompt reply. I did 12 PCR cycles. The post-PCR yield was 45.6 ng/uL, which come to think of it does seem high. Pre-PCR DNA was ~1 ng/uL.
evobio721 is offline   Reply With Quote
Old 12-21-2017, 12:57 PM   #4
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,171
Default

If you have run size selected tags on BA or still have remaining to run you can check for the presence of those peaks. I think the size selection has been sub optimal.

Other weak possibility is over amplification. I wonder what was the PCR input and final library elution volumes.

If it turns out to be over-amplification then a new library needs to be prepped. If it is result of size selection then you just will loose some reads and library can be sequenced.
nucacidhunter is offline   Reply With Quote
Old 12-21-2017, 01:09 PM   #5
evobio721
Junior Member
 
Location: Arkansas

Join Date: Dec 2017
Posts: 5
Default

Quote:
Originally Posted by nucacidhunter View Post
If you have run size selected tags on BA or still have remaining to run you can check for the presence of those peaks. I think the size selection has been sub optimal.

Other weak possibility is over amplification. I wonder what was the PCR input and final library elution volumes.

If it turns out to be over-amplification then a new library needs to be prepped. If it is result of size selection then you just will loose some reads and library can be sequenced.
Thanks for your input. The size selection was done on a Pippin Prep, and the elution volume was what the Pippin Prep manual recommends (can't remember exact amount at the moment, but I think 40uL?). PCR input was 5 uL repeated four times, followed by post-PCR AMPure XP pooling/cleanup of PCR product.

That's a good idea about running the size selected DNA on the BA. I will try that to see if the issue is pre or post PCR. I imagine that's about the only thing I can do at this point outside of re-doing library prep.
evobio721 is offline   Reply With Quote
Old 12-21-2017, 01:33 PM   #6
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,171
Default

So you have used in total 20 ng of size selected tags for PCRs and got 45.6 ng/ul after pooled PCR clean up. I am interested to know what was the elution volume from beads to look at possibility of over-amplification.
nucacidhunter is offline   Reply With Quote
Old 12-21-2017, 01:42 PM   #7
evobio721
Junior Member
 
Location: Arkansas

Join Date: Dec 2017
Posts: 5
Default

Quote:
Originally Posted by nucacidhunter View Post
So you have used in total 20 ng of size selected tags for PCRs and got 45.6 ng/ul after pooled PCR clean up. I am interested to know what was the elution volume from beads to look at possibility of over-amplification.
After the post-PCR AMPure cleanup, I eluted in 30 uL. So there was 45.6 ng/uL in 30 uL.

I should note that when I did the PCR, both the index that worked well (first image) and the index that contained the multiple peaks were done at the same time.
evobio721 is offline   Reply With Quote
Old 12-21-2017, 02:14 PM   #8
nucacidhunter
Senior Member
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,171
Default

It does not seem to be result of over-amplification. Most likely cassette lane has been faulty or other weak cause could be amplicon contamination from the lab if you have recently processed amplicons in those sizes recently in the same location.
nucacidhunter is offline   Reply With Quote
Old 12-22-2017, 07:40 AM   #9
evobio721
Junior Member
 
Location: Arkansas

Join Date: Dec 2017
Posts: 5
Default

Quote:
Originally Posted by nucacidhunter View Post
It does not seem to be result of over-amplification. Most likely cassette lane has been faulty or other weak cause could be amplicon contamination from the lab if you have recently processed amplicons in those sizes recently in the same location.
That sounds reasonable to me. Thank you very much for your time and help. I greatly appreciate it.
evobio721 is offline   Reply With Quote
Reply

Tags
bioanalyzer, ddrad, peaks

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 09:39 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO