  • RNA integrity number Drosophila

    Hello @all
    as we want to start an RNA-Seq analysis we try to determine the quality of our RNA (Drosophila heads) first. When we did this via Bioanalyzer an error occurred as expected. The measurement in a TapeStation results in a RIN though. I can`t explain why because still no 28S peak is visible. The RIN is the ration of 28S to 18S as I understood this.
    Therefore I don`t know if I can rely on that. Does someone have experience with that?
    Thank you!
    Jule

    ps: attached you see the trace of one sample with both machines
    Attached Files

  • #2
    It's a result of the 28S from insects being split. See this article: https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3016993/

    Comment


    • #3
      As consequence from what Eric wrote, one often has to eyeball the RNA integrity for insect samples on the Bioanalyzer (I am looking ratio of the rRNA peak size to the smaller debris fragments).
      Your example trace looks great.

      Comment


      • #4
        Thanks - I try it with these samples now. Fingers crossed

        Comment


        • #5
          Hi,
          I'd suggest using the Bioanalyzer over the Tapestation in the future. It gives a more granular electropherogram, whereas the Tapestation directly images the gel image from the run. This results in a less detailed electropherogram. Your trace looks good, but the RIN/RINe values are effectively redundant for insect material. Eyes are the best measure.

          The link above is a great explanation!

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Current Approaches to Protein Sequencing
            by seqadmin


            Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
            04-04-2024, 04:25 PM
          • seqadmin
            Strategies for Sequencing Challenging Samples
            by seqadmin


            Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
            03-22-2024, 06:39 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, 04-11-2024, 12:08 PM
          0 responses
          22 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 10:19 PM
          0 responses
          24 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 09:21 AM
          0 responses
          19 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-04-2024, 09:00 AM
          0 responses
          50 views
          0 likes
          Last Post seqadmin  
          Working...
          X