I accidentally set my index read as 10x as many cycles as I would like to sequence, and I am through most of my 1st read. Does anyone know how to change the parameters for the index read while a run is in progress?
Seqanswers Leaderboard Ad
Collapse
Announcement
Collapse
No announcement yet.
X
-
I understand it is too late for you now, but in principle changing paramaters mid-run is easy: you kill HCS, change parameters in the config files, restart. Since you discovered problem during cycle 1 and the barcodes are read at the end changing these parameters should create no issues. It is more tricky if you want to change parameters that are important early on later in the run. If you are still interested, let me know to post details about config files to be changed
Comment
-
Yes, MiSeq should be in the same boat. If you kill the run, it shouldn't write out to the RFID's so the cartridge and flow cell would still be useable. However, you'll have to change the recipe fragments so that when it goes to perform cluster generation it only does something mild like a PR2 delivery. MiSeq won't let you resume a run like HiSeq will.
Comment
Latest Articles
Collapse
-
by seqadmin
Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...-
Channel: Articles
04-04-2024, 04:25 PM -
-
by seqadmin
Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...-
Channel: Articles
03-22-2024, 06:39 AM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, 04-11-2024, 12:08 PM
|
0 responses
27 views
0 likes
|
Last Post
by seqadmin
04-11-2024, 12:08 PM
|
||
Started by seqadmin, 04-10-2024, 10:19 PM
|
0 responses
30 views
0 likes
|
Last Post
by seqadmin
04-10-2024, 10:19 PM
|
||
Started by seqadmin, 04-10-2024, 09:21 AM
|
0 responses
26 views
0 likes
|
Last Post
by seqadmin
04-10-2024, 09:21 AM
|
||
Started by seqadmin, 04-04-2024, 09:00 AM
|
0 responses
52 views
0 likes
|
Last Post
by seqadmin
04-04-2024, 09:00 AM
|
Comment