  • Quantitation of Nextera XT library

    I have searched the forum without success so apologies in advance if I have missed this elsewhere.

    We are sequencing pooled Nextera XT preps of bacterial genomes, using the bead normalisation. Until now we have not carried out any post-normalisation quantitation and have been reasonably happy with clustering on the MiSeq V3 of between 1000K and 1600K/mm2

    Recently, we have had a couple of failed runs with low clustering (<300K), possibly as a result of changes in staffing. We would like to use the Qubit ssDNA kit to quantify the pooled libraries. Does anyone have experience/suggestions/protocols?

  • #2
    I think you'll need the dsDNA assay kit since these are libraries. It's fairly straightforward and you can just follow the manufacturer's protocol. Nothing particularly special needs to be done for library quantification.

    Comment


    • #3
      We use the HS dsDNA kit for the Nextera XT input sample. But to measure the pooled normalised library, I think we need the ssDNA kit as the NaOH elution produces ss. Or am I wrong here?

      Comment


      • #4
        Link to Qubit ssDNA manual below. You might need to use higher volume of pool for quantification. Start with 5 ul and if signal is below detection it can be increased up to 20 ul.

        assets.thermofisher.com/TFS-Assets/LSG/manuals/Qubit_ssDNA_Assay_UG.pdf

        You will need to set up a small volume PCR reaction to estimate pool size for calculating molar concentration. Primers should be complementary to P5 and P7 flow cell binding motives such as primer cocktail form a TruSeq PCR or KAPA qPCR kit.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        24 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        25 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        21 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        52 views
        0 likes
        Last Post seqadmin  
        Working...
        X