SEQanswers

Go Back   SEQanswers > Applications Forums > Sample Prep / Library Generation



Similar Threads
Thread Thread Starter Forum Replies Last Post
Qubit library quantitation variations Smriti Sample Prep / Library Generation 5 11-02-2013 10:57 AM
qubit/picogreen seqgirl123 Sample Prep / Library Generation 4 11-02-2013 08:06 AM
Qubit Quantification addyblanch Sample Prep / Library Generation 2 07-02-2013 08:29 AM
Qubit data? jminich444 Sample Prep / Library Generation 1 04-25-2012 04:43 PM
QuBit fluorimeter vs. Epoch? sehrrot Sample Prep / Library Generation 3 09-16-2010 12:09 PM

Reply
 
Thread Tools
Old 06-29-2016, 12:56 PM   #1
tmm343
Member
 
Location: New York

Join Date: Dec 2012
Posts: 14
Unhappy Quantitation Variations with Qubit

Hello All,

I have run into quite a strange issue recently.

I have successfully and consistently been running 16S and 18S libraries and quantifying the final multiplexed pool using the Qubit. Recently I used a different assay then normal on the Qubit (BR instead of HS) and had a failed sequencing run. After going back and re-quantifying I saw that the BR and HS were yielding two different concentrations and always a twofold difference. BR always 2x higher than the HS reading. From our sequencing results, it also seems as the HS is accurate for the 18S sample while the BR is accurate for the 16S.

Has anyone encountered this before or have any idea as to what could cause this?

Thank you!
tmm343 is offline   Reply With Quote
Old 06-29-2016, 03:36 PM   #2
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,083
Default

It is most likely due to concentration of library. HS assay is recommended for library quantification because of its accuracy in low range.
nucacidhunter is offline   Reply With Quote
Old 06-30-2016, 11:30 AM   #3
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 189
Default

I don't get the same results from the BR and HS assays either. I think the HS assay is just not correct at the top end of the standard curve. I usually pool samples based on the HS assay and then use qPCR to quantify the pooled library - my pooled library is often twice as concentrated as it should be based on the Qubit.

Was the quantification of the pooled 18s library very different from the 16s library or were they both in the middle of the standard curve? And you are converting the values to nM using the size (in bp) of the library?
microgirl123 is offline   Reply With Quote
Old 06-30-2016, 02:44 PM   #4
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,083
Default

Quote:
Originally Posted by microgirl123 View Post
I don't get the same results from the BR and HS assays either. I think the HS assay is just not correct at the top end of the standard curve. I usually pool samples based on the HS assay and then use qPCR to quantify the pooled library - my pooled library is often twice as concentrated as it should be based on the Qubit.
Qubit and qPCR quantify different properties of library. Qubit quantifies dsDNA regardless of adapter presence in fragment termini while qPCR quantifies the fragments flanked by adapters that are amplifiable. The quantity measured by these methods will be closer to each other if all fragments were double stranded and flanked by adapter sequences.

Qubit will be fine for amplicon libraries quantification where the PCR input was low so there is not much carry over of non-adapted DNA.

Last edited by nucacidhunter; 06-30-2016 at 03:03 PM.
nucacidhunter is offline   Reply With Quote
Old 07-01-2016, 06:12 AM   #5
MU Core
Member
 
Location: Columbia, Missouri

Join Date: Apr 2008
Posts: 50
Default

Consistency can be an issue. Our group performs two independent HS Qubit readings per library. Doing so will help catch any outlier measurements due to a pipetting error, etc. The average is then used between the two readings for nM calculation. Using the average fragment size has been shown in our hands to be more consistent as well. Although this is less of an issue for amplicons.
MU Core is offline   Reply With Quote
Old 07-01-2016, 08:29 AM   #6
tmm343
Member
 
Location: New York

Join Date: Dec 2012
Posts: 14
Default

Hi All,

We normally qPCR but for these 16S and 18S libraries we have found Qubit to be more than sufficient and have been using it successfully and consistently and just recently ran into this issue and seeing this discrepancy.

How we came across the issue was using the Quanit BR reagents and a plate reader for the individual libraries and doing a final QC with the Qubit on the pooled sample as a double check. The Qubit assay initially used was the HS. While the concentration and pooling was done with the Quantit BR to a 4 nM pool the Qubit HS told us the pool was at 2 nM. We decided to use the 2 nM concentration and overclustered our run leading us to believe the BR assay was initially correct. We decided to go back and do this comparison and found the variation to be between the BR and HS assays which is showing using both Qubit and Quantit reagents. It becomes even more complicated the for 16S and 18S samples we see this discrepancy but seems as if the HS is accurate for the 18S samples and BR is accurate for the 16S samples.

@microgirl123 we are using fragment size to do the conversion to nM. The sizes are pretty consistent between all the samples.
tmm343 is offline   Reply With Quote
Old 12-01-2017, 06:06 AM   #7
DPVT
Junior Member
 
Location: Acela Corridor

Join Date: Dec 2017
Posts: 1
Default qPCR/ Qubit discrepency

Our lab has often encountered trouble quantifying our Illumina pools. We sequence all manner of samples and find that some pools cluster more robustly then other others even when loaded at the same molarity. As a work around we load pools at different concentrations according how different the qPCR concentration and qubit HS concentration is.
DPVT is offline   Reply With Quote
Reply

Tags
16s, 18s, quantification, qubit, variations

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:37 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO