  • Data recovery from aborted HiSeq2500 run

    Hi, we were running a 1x126bp high-output run on our HiSeq, but a comms issue caused the run to fail at cycle 109. Basecalling was complete however for cycles 1-107, although for some reason Q30% drops to about 50% after cycle 99. However, since the user only needs ~100bp of data, I would like to process the basecalls as they stand.

    What would I need to do in order for bcl2fastq to correctly process the data assuming I treat it as a 1x101bp run? Modify the runinfo/runparameters.xml files to alter read length and set usebasesmask to y101nnnnn (or something like that)? Or would I also need to copy the various RTAcomplete and BasecallingComplete.txt files from a successful run?

    Thanks,

    Matt

  • #2
    1) Fake the ....complete.txt files (just copy an old one and put a reasonable time stamp in there)
    2) Correct the read length in the xml files (you might need to go one less than you think, because last called cycle might be treated just for the phasing/prephasing calc.)
    3) Correct read length in your samplesheet (I did that, don't know is neccessary)

    That are the things I remember from the last time we needed it. Hope my list is complete.

    Comment


    • #3
      Thanks - it seems to have generated a makefile OK just running bcl2fastq with --use-bases-mask set to y101nnnnnn (since it had processed up to base 107), will know for sure once the make script has finished.

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Current Approaches to Protein Sequencing
        by seqadmin


        Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
        04-04-2024, 04:25 PM
      • seqadmin
        Strategies for Sequencing Challenging Samples
        by seqadmin


        Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
        03-22-2024, 06:39 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 04-11-2024, 12:08 PM
      0 responses
      30 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 10:19 PM
      0 responses
      32 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 09:21 AM
      0 responses
      28 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-04-2024, 09:00 AM
      0 responses
      52 views
      0 likes
      Last Post seqadmin  
      Working...
      X