  • Why and how: PhiX SPike-in?

    Hi Seqanswer team, and forum members,

    Pardon me if this seems to be a broad and naive question.

    I do not use sequencing kits, neither I do the quality checks of the data that immediately comes out of the sequencer. I deal with the data that comes after all these checks, and thus, I remain oblivious to what goes on behind the scenes before and during sequencing.

    I'm seeing PhiX term in 16s rRNA papers, and its concentration.

    I found these urls:.





    Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)


    Concentration varies from sequencing platform, and what you're sequencing. For low diversity sample, sometimes ~10% is suggested sometimes ~50%.

    My query:
    - I'd like a basic (read as superficial) understanding/definition what is the importance of PhiX?
    How does an external element helps to have a better quality data?

    - Doesn't it contaminate primers, bar code, indices?
    Bioinformaticscally calm

  • #2
    PhiX serves multiple functions:

    1) technical control for clustering reaction (spiking in a known amount of phiX should yield a known number of phiX clusters).
    2) technical control for sequencing accuracy (on-the-fly alignment to the phiX reference is used to calculate the sequencing error rate).
    3) introduction of sequencing diversity in low-complexity libraries (diversity is needed to discriminate clusters and create signal thresholds for base-calling). As the software has improved, the recommended amount of phiX spike-in has decreased.

    The phiX library does not contain an index; these reads are assigned to the Undetermined_indices directory.

    Comment


    • #3
      Hi HESmith,

      Thanks for valuable points and your reply.

      Would introducing higher % (~30-50) of PhiX for amplicons, at MiSeq platform would cause bias?
      Bioinformaticscally calm

      Comment


      • #4
        Older basecalling software on the MiSeq used to require a 50% phiX spike for low-diversity (amplicon samples). More recently, the software has been updated and only requires a 5-10% spike.

        Comment


        • #5
          Hi microgirl,
          Thanks for your reply.
          Bioinformaticscally calm

          Comment


          • #6
            Originally posted by bio_informatics View Post
            Hi HESmith,

            Thanks for valuable points and your reply.

            Would introducing higher % (~30-50) of PhiX for amplicons, at MiSeq platform would cause bias?
            Why do you think that would happen? phiX does compete for spots on the flowcell (but it should not out-compete specific amplicons) so you do end up losing some sequencing capacity.

            Comment


            • #7
              Hi Genomax,
              Thanks for your reply.

              I'm reading posts related to PhiX, joining pieces from here and there, and probably the question is incorrect/naive.

              I read:
              "Depending on what MiSeq software version was used for the sequencing, an inadequate amount of PhiX combined with too high a cluster density will result in an excessive number of sequencing errors leading to a large number of unique sequences."

              I asked wrongly on PhiX concentration, however, it was intended with cluster density.
              Again, I do not know much of these terms, and traversing for the sake of my knowledge, and thus muttering these queries.

              Originally posted by GenoMax View Post
              Why do you think that would happen? phiX does compete for spots on the flowcell (but it should not out-compete specific amplicons) so you do end up losing some sequencing capacity.
              Last edited by bio_informatics; 06-03-2015, 10:28 AM.
              Bioinformaticscally calm

              Comment


              • #8
                As @microgirl123 pointed out above there was a period in early MiSeq days when there were problems sequencing low complexity libraries. Most of these issues have been mitigated via software updates for MiSeq Control Software (MCS) over time.

                Comment


                • #9
                  Hi Genomax,

                  That clears many doubts.
                  Merci!
                  Bioinformaticscally calm

                  Comment


                  • #10
                    are my phiX QC measures ok?

                    I've just used phiX (v3) for the first time on a miseq run. I was using the 500-cycle reagent kit (running 241*2 cycles), and using 1% PhiX.

                    According to my run summary:

                    aligned = 1.3%
                    error rate (read 1) = 1.66 (+/- 0.13)
                    error rate (read 4) = 2.19 (+/- 0.14)

                    NB for both read 1 and read 2, error rate doubles between cycles 35 - 100

                    >=Q30 was 83.7% overall

                    Would these phiX measures fall within acceptable parameters, or should I be concerned? I can't find any clear guidance online as to what they actually mean.

                    Thanks in advance for any help!

                    Comment


                    • #11
                      The error rate sounds a little high to me, but that is dependent on other factors, including the cluster density for the run. Can you post the run summary chart and the intensity plot from the data by cycle section (IVC plot)?

                      Comment


                      • #12
                        thanks jessica_L
                        Here's the charts screen - is this what you were looking for?
                        Attached Files

                        Comment


                        • #13
                          Run looks to be within illumina spec. Have you had a chance to look at the data? Some of the libraries may have short inserts. A typical characteristic manifestation of this is dropping Q-scores towards the end of reads. I assume the run has completed without any problems.

                          Comment


                          • #14
                            Haven't looked at the data beyond running it on spades (where there seemed to decently few contigs). But when I ran the library on bioanalyser, the average fragment size for all samples were within the 600-700bp range.
                            The run finished without any problems.
                            Glad to hear it may be within illumina spec. Where could I find this information (recommended values/thresholds for QC measures)?
                            Thanks

                            Comment


                            • #15
                              As you do more runs you will start to develop a feel for what looks good and what does not. There isn't hard and fast thresholds (but a more continuous gradient) as to what constitutes good/bad runs. If you using your MiSeq in your own lab then you will have more control over samples/libraries. This is not possible when the instrument is being used in a core.

                              In general if you find a significant deviation (in terms of % Pass Filter, Q-scores) beyond the published spec (link below) then you stand a good chance of getting free replacement reagents from Illumina (once Illumina tech support determines that the problem was instrument/reagent related).

                              You can find published performance specification for MiSeq here: http://www.illumina.com/systems/mise...fications.html

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Essential Discoveries and Tools in Epitranscriptomics
                                by seqadmin




                                The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                                04-22-2024, 07:01 AM
                              • seqadmin
                                Current Approaches to Protein Sequencing
                                by seqadmin


                                Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                                04-04-2024, 04:25 PM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Yesterday, 08:47 AM
                              0 responses
                              14 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-11-2024, 12:08 PM
                              0 responses
                              60 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 10:19 PM
                              0 responses
                              60 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 09:21 AM
                              0 responses
                              54 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X