  • TruSight One - no library

    Hi everyone,
    I'm running the TruSight One panel, and the samples look great on Qubit and Bioanalyzer after the 1st PCR. However, at the end of the process I am validating the library and there is nothing there.....well, maybe a TINY little bit, but nothing in terms of the proper amounts.

    Any ideas? TechSupport is trying to help, and they say the thermocycler may be the problem. We use the SensoQuest or the Biometra. They are saying the hyb and capture steps are very sensitive, and they have only approved 3 thermocyclers for this use. They say this ramping down step is extremely important, but cannot tell me what specs to test any other PCR machines with. In essence, they are saying it's a guessing game, and to just try multiple machines if we can't buy the recommended cyclers. 1 of the 3 recommended cylcers is now discontinued, 1 of the tetrads is on backorder for 6-8 months, and the last one is extremely expensive.

    Does anyone have any experience with non-recommended cyclers?

    Thanks,
    Jim

  • #2
    Hi Jim,

    we are using the biometra machines (TPROFESSIONAL BASIC) to generate TS1 libraries and the are doing fine. You have to pay attention that your thermocycler is suitable for 100µl. Otherwise the temperature will not be correct in your fluids. As far as I know there are different blocks for the biometra machines. The basic block ist for up to 70µl while the basic xl is suitable for 100µl reaction volume.

    http://www.biometra.de/index.php/thermoblock.html

    Do you perform the first hybridazation overnight?

    kind regards,

    -mike-

    Comment


    • #3
      Hi Mike,

      Thanks for the reply. We are not doing the first hyb overnight. We used the SensoQuest, and not the Biometra, and we got bad results (after full library prep, only 1.1 ng/ul). Illumina sent new oligos, and we tried the SensoQuest with the old and new oligos, and an Eppendorf Nexus with old and new oligos. Illumina Field Apps was here while we did the lab protocols and saw nothing wrong. All libraries failed again, but the Eppendorf was about 3 ng/ul. Illumina says we should have at least 10 ng/ul as a bare minimum, but more likely 20-30 or even 50 ng/ul. We're really at a loss here.......

      Jim

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Current Approaches to Protein Sequencing
        by seqadmin


        Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
        04-04-2024, 04:25 PM
      • seqadmin
        Strategies for Sequencing Challenging Samples
        by seqadmin


        Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
        03-22-2024, 06:39 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 04-11-2024, 12:08 PM
      0 responses
      30 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 10:19 PM
      0 responses
      32 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 09:21 AM
      0 responses
      28 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-04-2024, 09:00 AM
      0 responses
      53 views
      0 likes
      Last Post seqadmin  
      Working...
      X