Default RIN of 6- should I run sample?

I have RNA from hard-to-come-by samples. Some RINs are 8+ but some are 6.3-6.9. I would need comparative RNA-Seq work on this, as well as transcriptome assembly, as this is an unsequenced genome.
Should I chance the samples with the low RIN on an Illumina?
Noa is offline   Reply With Quote
Old 03-04-2012, 09:59 AM   #2
pbluescript
Senior Member
 
Location: Boston

Join Date: Nov 2009
Posts: 224
Default

Go ahead. The first step in many RNA library prep protocols is fragmenting the RNA, so a lower RIN is generally not a problem. RINs in the 6 range really aren't that low.
pbluescript is offline   Reply With Quote
Old 03-04-2012, 10:39 AM   #3
mnkyboy
Member
 
Location: Seattle, WA

Join Date: Mar 2009
Posts: 87
Default

If it is whole transcriptome it will not matter, if you are doing some mRNA-seq you may see some loss but if it is for assembly you should be fine since you are probably doing a lot of reads.
mnkyboy is offline   Reply With Quote
Old 03-05-2012, 12:58 PM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

RINs that low will cause a problem in most use cases.

(1) If you plan to do a polyA+ RNA purification (as in TruSeq RNA prep libraries), then only the segments of your RNA still attached to their polyA tails will be isolated. The result will be 3' bias in your sequence reads.

(2) If you plan to ribo-deplete your sample using a hybridization-based method, then some ribosomal RNA will "bleed" through the process because the ribosomal RNA will be somewhat fragmented. Hybridization-based methods generally rely on a set of biotinylated oligonucleotides complementary to certain regions on the ribosomal RNA. Extensive fragmentation will create fragments without these binding sites on them and allow them to escape the depletion process. The result will be a higher level of ribosomal RNA contamination of your reads. The result will be fewer useful reads. Epicentre's "ribozero" is more effective at removing slightly degraded ribosomal RNA -- presumably by supplying a more generous diversity of binding oligos.

(3) If you plan to use normalization to ribo-deplete your samples, then everything is likely fine. Except that normalization is trickier than other methods of ribo-depletion.

--
Phillip
pmiguel is offline   Reply With Quote
Reply

Tags
illumina, rin, rna-seq

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 12:55 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2021, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO