Default New NextSeq: First run. A few issues.

Hi All,
We're running standalone with our new NextSeq500 and the very first user came in with 30 multiplexed NEB Next Ultra2 libraries, with BIOO single index 12bp long adapters, for us to run on a MID-flowcell as genomic DNA (2x150) using an Illumina kit.
We spent some time checking with NEB, and then BIOO, and Illumina, before proceeding. We set up a plate with the 12bp indices on the I7 sequence primer. In short, although slightly overclustered (233M/mm2) the run looked okay (89% >Q30), and all samples demultiplexed. A few things not specifically discussed on these forums have appeared in fastqc and the flowcell summary.

Newbie Question: Roughly 10% of my reads went to "Undetermined". This is not far off from our 10% PhiX spike in. Of those that were PhiX, at least 90% had "GGGGGGGGGGGG" as the barcode. Is it normal for bcl2Fastq to keep these dark indices? It also kept over a million reads with a barcode of "NNNNNNNNNNNN". Normal?

PROBLEMS:
For the "good" demultiplexed sequences of ONE sample (~4.7M reads):
1. In read-pair file reads #1, there were ~42K consisting of a stretch of Ns, exactly 35 bp long. Also >37K were PhiX reads ending in a long run of Gs. They showed the correct index.
2. Same sample, in read-pair file #2, about 243 sequences were reads dominated by mostly Ns, with some a string consisting of only 35 Ns. There were also about 1500 PhiX reads ending in long runs of Gs. They also showed the correct index.

How would this happen? It doesn't seem like a tile edge effect, and while I know that PhiX makes it into the "good" reads sometimes, and I know that "G"s can be heavily over represented at the ends of read sequences, why so many, and why so different depending on the read pairs?

Thanks.
-p
hoytpr is offline   Reply With Quote
Old 04-20-2017, 08:49 AM   #2
hoytpr
Member
 
Location: Stillwater

Join Date: Dec 2009
Posts: 62
Default

Answering my own question: The long "G"s are likely adapter/primer dimers, where the read one primer bound, but not the read two primer. Not yet sure if this is due to the use of NEB+BIOO+Illumina kits. The 35 "N"s are from adapter/primer dimers being N-masked.
-p
hoytpr is offline   Reply With Quote
Old 04-20-2017, 09:06 AM   #3
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,091
Default

Illumina's phiX is not indexed (if I recall right). The GGGG's basically is NextSeq saying I see no signal = G basecalls (2-color chemistry).
GenoMax is offline   Reply With Quote
Old 04-20-2017, 10:13 AM   #4
hoytpr
Member
 
Location: Stillwater

Join Date: Dec 2009
Posts: 62
Default

I'm sure you are correct, thanks. The weirdness (among other things) was the numerical disparity between read 1 and read 2. Maybe because the read2 index in the plate file wasn't entered (which causes the read2 to be automatically 'masked'), and also the libraries had very low complexity (e.g long runs of Gs). In the end, the run worked fine but may need some cleanup.

After a couple days with Illumina, they said :
Quote:
I pulled the 47,000 sequences that gave Ns in R1 from R2 and it looks like pretty much all of them have 100% G calls. A G call in NextSeq is usually associated with dark cycles (or no image). This tells me that the R2 primer may not have bound correctly to these obviously adapter dimer sequences (since R1 is 35bp of Ns) and that is why we don't see the adapters in R2 region and hence no subsequence conversion to Ns.
This is still a little murky to me, so I plan to draw out the whole NEB+12bp-BIOO+Illumina indexing on paper.
hoytpr is offline   Reply With Quote
Old 04-20-2017, 12:45 PM   #5
jdk787
josh kinman
 
Location: Austin

Join Date: Apr 2014
Posts: 71
Default

Quote:
Originally Posted by hoytpr View Post
I'm sure you are correct, thanks. The weirdness (among other things) was the numerical disparity between read 1 and read 2.
I don't have any hands on experience with the NextSeq, but there seems to be a tendency for NextSeq data to have more polyG sequences in R2 than in R1. Possibly due to some inefficiency in the Read 2 re-synthesis. Maybe this is worse for adapter dimers?
http://seqanswers.com/forums/showthr...t=40741&page=5
http://seqanswers.com/forums/showthread.php?t=45130

bcl2fastq default will also mask reads that are less than 22 bases after trimming with Ns along with adapter dimers.
https://support.illumina.com/content...5051736-01.pdf

Quote:
Maybe because the read2 index in the plate file wasn't entered (which causes the read2 to be automatically 'masked')
If you don't enter an I5 index than the sequencer will just read I7 and not I5. So for paired end single index there will be reads for R1, I7, and R2. The I5 read will be skipped and the I7 will be used to demux.
http://www.bea.ki.se/documents/Illum...de_Aug2015.pdf
__________________
Josh Kinman
jdk787 is offline   Reply With Quote
Old 04-20-2017, 12:57 PM   #6
hoytpr
Member
 
Location: Stillwater

Join Date: Dec 2009
Posts: 62
Default

Thanks very much for the links and clarification.

-p
hoytpr is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 04:19 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2021, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO