Old 09-18-2013, 06:00 AM   #1
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 199
Default MiSeq Cluster Density Revisited

This is discussed ad nauseum, but I can't seem to find the answer I'm looking for so here goes! Over the last two months my cluster densities have gone from 800-1000 K/mm2 to 500-600 K/mm2. Same v2 chemistry, same library prep (TruSeq-like), same KAPA quantification. I've been clustering with 8 pM libraries and tried going up to 8.5 pM but it hasn't made a difference.

What concentration are most of you clustering at? I'd like to get back up to the ~900 K/mm2. And can NaOH go bad? I'm thinking that maybe it's not denaturing properly. Is it possible to buy it in small aliquots so that I can open a new tube/container every couple of runs?
microgirl123 is offline   Reply With Quote
Old 09-18-2013, 06:18 AM   #2
mcnelson.phd
Senior Member
 
Location: Connecticut

Join Date: Jul 2011
Posts: 162
Default

Our clustering is a bit variable, but we're generally in the 700-1000K range for all runs, with the average over our last few being ~850K. This is a bit lower than what the MiSeq can obviously support, but we'd rather under cluster a little than over cluster and get terrible data back.

As for the NaOH, according to Illumina it can "go bad". I've been storing 25ul aliquots of 2N NaOH in our freezer and each time we go to denature libraries we pull out a new aliquot and dilute it to 0.1N instead of 0.2N. We used to have a ton of problems getting consistent clustering, but since we've switched to this method we've been more consistent even though we're still working with lots of different library types that require different final pM concentrations in order to get appropriate clustering.

I'll note that we still don't use the Kappa kit for quantification and only do Qubit HS DNA and Bioanalyzer, so that obviously introduces some variability in our clustering since we're not quantifying just fragments that would bind to the flow-cell.
mcnelson.phd is offline   Reply With Quote
Old 09-18-2013, 07:26 AM   #3
clintp
Member
 
Location: Georgia

Join Date: Apr 2013
Posts: 19
Default

Regarding NaOH, I wonder what the best way is to store it. Does anyone store it under mineral oil at room temperature? Or is freezing an effective method to prevent interaction with CO2 in the air?

Also, mcnelson.phd, have you had any conversations with Illumina about your better/more consistent clustering with reduced [NaOH]?
clintp is offline   Reply With Quote
Old 09-18-2013, 07:50 AM   #4
mcnelson.phd
Senior Member
 
Location: Connecticut

Join Date: Jul 2011
Posts: 162
Default

Quote:
Originally Posted by clintp View Post
Also, mcnelson.phd, have you had any conversations with Illumina about your better/more consistent clustering with reduced [NaOH]?
The recommendation to use 0.1N NaOH as the starting point for denaturing actually came from our FAS when we were discussing our problems of inconsistent clustering. At first we thought we were having fluidics problems because some runs had densities in the 100-200K range, but that was ruled out as a problem. Our FAS told us that the 0.2N value is really overkill, and that they were recommending the 0.1N starting value to their other sites.

Since we've started using 0.1N for everything, and using the frozen 25ul aliquots I mentioned above, we've had very consistent clustering for runs that only feature a single library type (e.g. only 16S amplicons or only Nextera genomes). Most of our variation comes up when we mix different library types and try to compensate for the fact that smaller fragments cluster better by diluting libraries to different final pM concentrations before pooling them to load into the cartridge.

Last words I really have are that we have also given up entirely on using the XT library normalization and denaturing process and instead take the post-PCR cleaned libraries and treat them like standard Nextera libraries. Since we've started doing that we haven't had the horribly uneven pooling or over clustering issues we had when using the official normalization and denaturing protocol.
mcnelson.phd is offline   Reply With Quote
Old 09-23-2013, 05:22 AM   #5
sweetph3
Member
 
Location: New Jersey

Join Date: Jun 2013
Posts: 22
Default

We have been freezing 1 ml aliquots of 1N NaOH (no mineral oil). We toss the aliquot after the 2nd thaw. Dilute to 0.2 N according to Illumina's protocol. Never see any problems with clustering doing it this way. Good luck.
sweetph3 is offline   Reply With Quote
Old 09-24-2013, 04:30 AM   #6
yaximik
Senior Member
 
Location: Oregon

Join Date: Apr 2011
Posts: 205
Default

Quote:
Originally Posted by clintp View Post
Regarding NaOH, I wonder what the best way is to store it. Does anyone store it under mineral oil at room temperature? Or is freezing an effective method to prevent interaction with CO2 in the air?

Also, mcnelson.phd, have you had any conversations with Illumina about your better/more consistent clustering with reduced [NaOH]?
I stored original Illumina's aliquiot of 2N NaOH just on the bench, tightly closed, making single use 20 ul dilutions every time. Did not notice much difference after a year or so. Indeed, how much CO2 is in the air over solution in the capped tube even if it is opened and recapped? Hard to imagine something less complicated than NaOH, may be salt or water...
yaximik is offline   Reply With Quote
Old 09-25-2013, 04:51 PM   #7
LVAndrews
Member
 
Location: Flagstaff, AZ

Join Date: Sep 2012
Posts: 55
Default

I made some 2N NaOH when our miseq arrived a year ago and it has lived on bench since then in a 1.5mL snap top tube. I dilute fresh 0.2N from this stock each time. No trouble!!

Also, last time I denatured phix was May. I just keep the same 20pM stock in the freezer. Probably 25 runs or so since then, no problems with PhiX either.

And to the original question about clustering, we used to load at lower concentrations (~8-12pM depending on application), but around June our cluster density started to drop off. Illumina came out and cleaned our lines and upgraded the valve and back in business. We now load between 14-18pM and typically see clustering between 900-1200k. Every now and then we botch it a little, usually when a user supplies a library with adapters remaining. For amplicon runs, we load a bit lower since these seem to cluster more densely. My picogreen and qPCR quants are usually in good agreement. I wouldn't trust a bioanalyzer quant on a library that has been amplified at all (unreliable due to pipet error - just 1ul, and the nonvisualizable heteroduplex component). Generally increasing load concentrations by just a hair (0.5pM) has little to no effect, probably due to variability in library quantification. If you are clustering at 500k with 8pM, I would bump up to 12pM for the next run and see where that puts you. It's not exactly a linear relationship between loading concentration and cluster density, but not far off either. You might want to go as high as 16pM. For Kozarewa and Turner preps (truseq like), we load at 18pM.

For qPCR we don't use the Kapa kit. Great company (check out the 2G Fast polymerase), but how hard is it to make qPCR standard, especially when you are making Illumina-compatible amplicons? Just take a bit of a previous amplicon library (length should be known - I use 16S v4 construct), and reamplify it to produce more. Bead clean it (see this post: http://enggen-nau.blogspot.com/2013/...-cleanups.html), pico-quantify, and std made. Use P5/P7 sequences as primers. Can also use these for QC checks during other preps (TSCA, Nextera, etc).
LVAndrews is offline   Reply With Quote
Old 09-25-2013, 04:57 PM   #8
LVAndrews
Member
 
Location: Flagstaff, AZ

Join Date: Sep 2012
Posts: 55
Default

One more thing regarding NaOH. It is ridiculously easy to make. 400g NaOH pellets, dissolve in H2O (exothermic), bring to 1L (10N solutioN). Cap tightly and it is good for years.
LVAndrews is offline   Reply With Quote
Old 10-11-2013, 11:02 PM   #9
ramujana
Member
 
Location: Vilnius

Join Date: Apr 2012
Posts: 11
Default

Dear All,
i'm running TruSight Cancer Nextera Enrichment protocol for libraries (genomic DNA) on MiSeq and end up in quite low (although very good quality reads) cluster density 180K with 0,2N NaOH. The length of libraries on average was 1000 bp which is quite long i think, what could impede the generation of clusters on flowcell. The input of libraries I made according the illumina protocol suggestion ~20nM. What would be the suggestions to improve cluster density? Should i use 0,1N NaOH (this is actually the quantity which is in the arrays preparation protocol)?
ramujana is offline   Reply With Quote
Old 10-12-2013, 06:30 AM   #10
yaximik
Senior Member
 
Location: Oregon

Join Date: Apr 2011
Posts: 205
Default

In regard to eliminating effect of remaining NaOH I found a very simple solution. Make solution of 1M acetic acid in water, it perfectly stable on a bench. Add 1 ul to the final dilution of your library in HT before loading into the library well - cluster density jumps several fold. I overclustered a few times before I realized I got to be more conservative with this modification.
BTW, I prefer not to waste library, so I use exact amount recalculated back to 600 ul of final volume from the protocol recommendations. So, for example, to get 600 ul of 50 pM I simply combine 3 ul 10 nM with 3 ul of 0.2 M NaOH, wait, then dilute straight with 594 ul of HT, add 1 ul of 1 M AcOH and load the well. One can hit density 3000 at this concentration, though.
yaximik is offline   Reply With Quote
Old 12-11-2013, 09:45 PM   #11
chen
Member
 
Location: HK

Join Date: Sep 2009
Posts: 14
Default

Quote:
Originally Posted by ramujana View Post
Dear All,
i'm running TruSight Cancer Nextera Enrichment protocol for libraries (genomic DNA) on MiSeq and end up in quite low (although very good quality reads) cluster density 180K with 0,2N NaOH. The length of libraries on average was 1000 bp which is quite long i think, what could impede the generation of clusters on flowcell. The input of libraries I made according the illumina protocol suggestion ~20nM. What would be the suggestions to improve cluster density? Should i use 0,1N NaOH (this is actually the quantity which is in the arrays preparation protocol)?

We've got similar issue with the Miseq, low density at 122K with 0.2N NaOH according to Miseq user guide. It's a polyA mRNA lib using the NEB kit, with 750bp length on average. The final concentration of loading lib is 11pM.

Is it the long lib that cause the problem or the NaOH concentration?
any suggestion?
__________________
Cheers,
Tanya
chen is offline   Reply With Quote
Old 12-13-2013, 06:33 AM   #12
ychang
Member
 
Location: philly

Join Date: Oct 2010
Posts: 17
Default

when we get the illumina training one year before, the FAS have told us to use 0.1N NaOH. it is interesting that yaximik use 1M acetic acid before loading on FC, have anyone tried HCl?
ychang is offline   Reply With Quote
Old 12-13-2013, 06:39 AM   #13
yaximik
Senior Member
 
Location: Oregon

Join Date: Apr 2011
Posts: 205
Default

HCl is too strong acid I think, it will skew pH significantly. Obviously, one ul of 1M acetic acid is in excess to remaining NaOH, but it is much weaker and rather create acetate buffer.
yaximik is offline   Reply With Quote
Old 03-27-2014, 12:30 PM   #14
LVAndrews
Member
 
Location: Flagstaff, AZ

Join Date: Sep 2012
Posts: 55
Default

Yaximik, I am suddenly finding myself interested in your acetate mod. I have some libraries that just don't want to reach 4nM. Your mod suggests adequate clustering can be had with lower concentration libraries, however, I notice your final NaOH concentration is the same as the rest of us (0.001N), and yet you are still adding acetic acid. Has this improved your clustering consistency? What is the lowest concentration library you have denatured and achieved useful cluster density?
LVAndrews is offline   Reply With Quote
Old 03-27-2014, 01:04 PM   #15
yaximik
Senior Member
 
Location: Oregon

Join Date: Apr 2011
Posts: 205
Default

Addition of acetic acid likely will not, or very slightly may improve clustering above what is achieved with final NaOH at 1 mM. The mod was designed to take care of higher NaOH concentrations, which happens if you mix libraries at low concentration with equal volume of 0.2 M NaOH. The lowest concentration I used was 2.5 nM (for v3 kits). I used 3 ul of library and 3 ul of 200 mM NaOH. After 5 min, I hung 1 ul of 1 M AcOH on the tube wall, then diluted all straight with 600 ul of HT, and loaded into the well. In two cases I got ~1400 cl/mm2. Sometimes I intentionally dilute 10 nM libraries two or three times to use more reliable volumes than, say 0.75 or 1 ul, and use accordingly 2-3 times more 0.2 M NaOH. Perhaps you could use 1.5 ul of your 4 nM libraries. I do not see why one cannot use much lower library concentrations, as AcOH is added in huge excess to NaOH anyway.
yaximik is offline   Reply With Quote
Old 03-27-2014, 01:20 PM   #16
LVAndrews
Member
 
Location: Flagstaff, AZ

Join Date: Sep 2012
Posts: 55
Default

Doing it!! My difficult library is 2.7nM. Too lazy to bead-concentrate and re-quant. Plus I like to try useful mods. Loading a 600 cycle kit now. Thanks!!
LVAndrews is offline   Reply With Quote
Old 03-27-2014, 05:42 PM   #17
austinso
Member
 
Location: Bay area

Join Date: Jun 2012
Posts: 77
Default

Yaximik, are you saying that when you neutralize with acetic acid, you don't dilute 100-fold with HT1? You just make it up to 600 ul final?

Just wanted clarification...

Best

Austin
austinso is offline   Reply With Quote
Old 03-27-2014, 06:55 PM   #18
rnaeye
Member
 
Location: East Cost

Join Date: May 2011
Posts: 79
Default

Hi!,
I make 1 N NaOH in 15 mL tube. And make fresh 0.2 N NaOH every time. I have not tested longer times but 1 N NaOH is stable in a tightly capped 15 mL conical for at least 6 weeks at room temperature in my hands. I do not let other people touch my tube If the cap is left open long time, CO2 in air may change pH in long term.

One important thing is that do not use more than 1 mm M final concentration of NaOH in denatured libraries. I would recommend do not be even close to it. I think hybridization is very sensitive to the amount of NaOH especially around 1mM. I start with 4 nM library to make sure I am not getting close to 1 mM NaOH limit. If you have lower concentrations just use freshly prepared 0.1 N or 0.15N NaOH instead of 0.2 N, it works without any problem. Since I try to be far from max recommended 1 mM NaOH I have never had any problem (previously had 2-3 failures).

Hope helps!
rnaeye is offline   Reply With Quote
Old 03-28-2014, 09:53 AM   #19
LVAndrews
Member
 
Location: Flagstaff, AZ

Join Date: Sep 2012
Posts: 55
Default

austinso, Yaximiks recipe does diulte 100fold with hyb buffer. Since he(?) does 3ul NaOH and 3ul of pool then the final dilution achieves this.

Thanks rnaeye for the clarification. I almost tried this instead yesterday to get my concentration up, but figured I should let the denature go a bit longer if I used a lower NaOH concentration. I found this paper (http://onlinelibrary.wiley.com/doi/1....201200934/pdf) which seemed to indicate that NaOH as low as 0.5mM might not completely denature the sample.

I wanted to report my run metrics since I was attempting Yaximiks neutralization mod. Raw library was 2.73nM, so following this protocol, my final (loading) pool concentration was 13.65pM + 10uL 20pM PhiX = 13.98 (call it 14pM loaded). Density is 1546 +/- 46 k/mm, 81.5%PF (28M reads), 17G projected yield, 93.2% >q30 after 220 cycles. In a word, fantastic!!
LVAndrews is offline   Reply With Quote
Old 03-28-2014, 10:54 AM   #20
austinso
Member
 
Location: Bay area

Join Date: Jun 2012
Posts: 77
Default

I was referring to this statement, AKrohn:
Quote:
Sometimes I intentionally dilute 10 nM libraries two or three times to use more reliable volumes than, say 0.75 or 1 ul, and use accordingly 2-3 times more 0.2 M NaOH.
I wasn't sure if that meant that he had tried denaturing a larger volume of a more dilute library solution, with a corresponding increase in volume of 0.2N NaOH, then neutralizing with 1 ul of 1M HOAc.

As the net effect with HT1 is neutralization in the absence of HOAc, so in the presence of HOAc, maybe that volume is not as important (i.e. just make up to 600 ul).

Expensive experiment to do, so I thought I'd get clarification than do it myself.

Neither here nor there, really...
austinso is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:49 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO