  • DNA library shoulder peak on bioanalyzer

    Hi Everyone,

    We recently just processed a few DNA libraries using the Illumina Truseq PCR free kit. When we ran them on Bioanalyzer, one of the libraries came out perfect, with a single peak centered around 500 bp. The others, however, have a secondary shoulder peak ( ~ 375 bp) attached to the main library curve. We have never had this happen to our previous libraries. Has anyone else seen this/know the underlying cause and how to avoid it? Thanks!
    Attached Files

  • #2
    Few points:

    1- Chip is overloaded
    2- If shoulder is not consistent among the samples it could be due to vibration during Chip run.
    3- Is the perfect library concentration is lower than the others. If true it would indicate artefact of overloading Chip.
    4- The trace is not similar to TruSeq PCR-free library profiles (350 or 550 bp insert)
    5- If shoulder is real, could be the result of size selection step

    Comment

    Latest Articles

    Collapse

    • seqadmin
      Current Approaches to Protein Sequencing
      by seqadmin


      Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
      04-04-2024, 04:25 PM
    • seqadmin
      Strategies for Sequencing Challenging Samples
      by seqadmin


      Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
      03-22-2024, 06:39 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by seqadmin, 04-11-2024, 12:08 PM
    0 responses
    18 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 10:19 PM
    0 responses
    22 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 09:21 AM
    0 responses
    17 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-04-2024, 09:00 AM
    0 responses
    49 views
    0 likes
    Last Post seqadmin  
    Working...
    X