  • Doing multiplex in one lane only

    We want to do a little multiplex sequencing using the Illumina kit and recipes, but we don't need to do a lot. We'd like to do a single lane on a flow cell with all the other lanes containing standard PE libraries. Problem is the multiplex recipes use a different read 2 sequencing primer. Illumina tech support tells me that some customers are just using an equimolar mixture of the multiplex and standard read 2 sequencing primer. Seems reasonable to me. Has anyone tried it, or something similar?

  • #2
    I don't think this could work. The primer for the index read is complementary to the same strand as the primer for read 1. The index read is supposed to be performed prior to cluster resynthesis (if you are doing a paired read). I have attached a picture from one of Illumina's publications showing the relationship of the read1, index read and read 2 primers and order of operations.
    Attached Files

    Comment


    • #3
      Hmmm... Perhaps I wasn't clear. I didn't intend to mix the index sequencing primer with the PE read 2 sequencing primer. I meant to mix the multiplex read 2 sequencing primer with the PE read 2 sequencing primer (because the multiplex library uses a different read 2 sequencing primer, supplied with the kit). I don't really see any show stopper there so long as the two primers don't anneal efficiently to the ends from the other library type (i.e. the PE read 2 sequencing primer doesn't anneal to the multiplex kit library ends and the multiplex read 2 sequencing primer doesn't anneal to the PE kit library ends). Does that make more sense?

      Comment


      • #4
        In the immortal words of Emily Litella...

        Never mind.

        You were making sense.

        (Must remember not to post before my morning coffee.)

        Comment


        • #5
          Has anyone tried this yet? Or heard anything from Illumina about it?

          Comment


          • #6
            Is anyone able to comment on whether mixing standard PE read 2 and multiplex PE read 2 primers works for getting read 2 in a flow cell with mixed libraries?

            I am also interested in this topic-- we are only going to do single lanes most of the time, so I am trying to make the libraries as flexible as possible so that I can jump into a SE, PE, or PE multiplex run. My libraries will be multiplex PE.

            Comment


            • #7
              Yes, you can mix the primers. Run it like a multiplex run... Just be sure to get the volume (hence concentration) of primers right according to the volume in the multiplex rd2 primer tube. Use multiplexing PhiX if you're using phi control lanes. We have done this many times. I think yhere may be another thread about this too...

              Comment


              • #8
                Thanks for the response. Are you saying to use the multiplexing PhiX control just as a control for the multiplex reads, or because the software has a multiplexing mode that will throw errors if it doesn't see indexes from the PhiX reads? (i.e. would the core have to do that to make the run work at all, or is it just a recommendation?)

                Comment


                • #9
                  Hi,

                  Sorry, I should have been more clear... It's just a recommendation, not a requirement. I simply meant that, if you're using a PhiX control, don't forget that there's a special version of the PhiX library for multiplexing. The software doesn't strictly require it.

                  Comment


                  • #10
                    Ah, that is great to hear-- thanks for the advice. Ultimately it will be up to the core here regarding running the two in the same flow cell, but at least it is feasible.

                    Comment


                    • #11
                      Mixing index and non-index lanes on a flowcell is fine. However, there are a few considerations to ensure success.

                      The multiplex and standard READ 2 primers can be mixed and in fact are mixed in the new TruSeq kits (HP7 is the mixed reagent). If you are using TruSeq, read 2 will sequence either the index or standard adaptors. If you are not using TruSeq, just mix the primers with the usual amount of each in the final primer mix.

                      The VERY important consideration is related to the focus of the instrument and this is a known problem on the HiSeq. When the index read is being performed in the lanes that do not contain any clusters since they are not indexed, it will throw the focus off for read2. This issue is not 100% consistent and can manifest itself in some odd ways but the quality of the run is definitely affected.

                      The easy fix is to include index PhiX at 1% in all lanes. That will allow all lanes to have some active clusters for every cycle (Read 1, Index read, and read 2). That will maintain the focus and prevent any surprises.
                      HudsonAlpha Institute for Biotechnology
                      http://www.hudsonalpha.org/gsl

                      Comment


                      • #12
                        Thanks csquared-- this is extremely valuable and helpful information. Our core spikes in 1% PhiX into all lanes, but I will need to be sure that they use index PhiX for my runs.

                        If I am in a SE run and want to read my index, I won't need to worry about the focusing issue, correct? I am mainly interested in using index adapters for convenience in the library protocol I am working up. Being able to jump into a PE run is just a nice side effect.

                        Comment


                        • #13
                          Correct. For a SE run, the focus issue is not a problem. Only affects read 2. In our experience, it has been quite variable. Everything from killing read 2 on a whole flowcell to no effect. Typically it results in a few lanes to drop out on read 2.

                          Including the index PhiX at 1% or more completely solves the issue. Really important to note that this issue is not like a control lane. You need something detectable for the index read in ALL lanes to avoid the focus issue.
                          HudsonAlpha Institute for Biotechnology
                          http://www.hudsonalpha.org/gsl

                          Comment


                          • #14
                            Is this focus issue a problem with the GAIIx? I'm just completing a small indexing run and was unaware of the indexed PhiX. We had two lanes drop out for read two. The coincidence may be that these two lanes were prepped in a different way and may have some other issues...

                            Comment


                            • #15
                              A related question-- is it possible to feed a custom sequencing primer to just one lane, or is the whole flow cell fed with the same mixes?

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Current Approaches to Protein Sequencing
                                by seqadmin


                                Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                                04-04-2024, 04:25 PM
                              • seqadmin
                                Strategies for Sequencing Challenging Samples
                                by seqadmin


                                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                                03-22-2024, 06:39 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, 04-11-2024, 12:08 PM
                              0 responses
                              25 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 10:19 PM
                              0 responses
                              28 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 09:21 AM
                              0 responses
                              24 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-04-2024, 09:00 AM
                              0 responses
                              52 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X