  • Amplicon resolving differently between bioanalyzer and tapestation

    Hello,
    We are seeing a discrepancy between the Bioanalyzer and Tapestation for a large amplicon product that we received for sequencing. We received the PCR reaction for this amplicon and did an Ampure XP bead clean up. The PCR product was run on an agarose gel before giving us and a single band of the expected size (~665bp) and primer dimers (~100bp) was seen. When we ran in the bioanalyzer (attached ppt), there are two peaks showing up. One around 370bp and the other 664bp. On the tapestation (attached ppt) we only see a single peak around 670bp. What's causing this difference? Is the amplicon making secondary structure and migrating slower in the bioanalyzer? But in tapestation, why isn't the same thing showing up? Have anyone else seen this before?
    Just wanted to mention that the primers were ~30bp long and didn't have any ssDNA structures like the Y adapters in Illumina. This is a cDNA amplicon product that we are planning to ligate adapters to.
    Thanks in advance!
    Attached Files
    Last edited by Pab77; 01-25-2016, 12:13 PM.

  • #2
    I have run a lot of amplicon samples on the Bioanalyzer and never seen that! Since it's a high sensitivity chip, is it possible that the previous run contaminated the electrodes with something 370 bp in length? I'd pull out the electrode cartridge, clean it thoroughly, and try again.

    Comment


    • #3
      Thanks for the reply! We actually ran two other samples in this same chip and those were not contaminated with this same peak. It's unlikely that only this pin was contaminated. We can try to redo the run but yes, it's really weird.

      Comment


      • #4
        I say re-run it and see what happens. With something this strange I always like to make sure the result is repeatable before looking too much into it.

        Also, if you still have the rest of the 1:20 dilution run it on an agarose gel and see if the banding pattern repeats. There is a chance something strange happened during the dilution.

        Edit: One more thing. Was anything of this size (or expected size) run on the same chip? Maybe the well got double-loaded.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM
        • seqadmin
          Techniques and Challenges in Conservation Genomics
          by seqadmin



          The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

          Avian Conservation
          Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
          03-08-2024, 10:41 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, Yesterday, 06:37 PM
        0 responses
        11 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, Yesterday, 06:07 PM
        0 responses
        10 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 03-22-2024, 10:03 AM
        0 responses
        51 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 03-21-2024, 07:32 AM
        0 responses
        67 views
        0 likes
        Last Post seqadmin  
        Working...
        X