  • Illumina second read quality drop

    Hello,
    For several months now, we observed on our GAIIx that the intensities for the second read in a paired end run (V4 kits) are often below the intensities of the last cycle of the first read. Before we always obtained a first cycle intensity for read 2 between 100 to 70% of the the first cycle intensity for read 1 (as it is normally expected).
    We have had some extreme runs where the 1st cycle intensity of read 1 is 386 and for the second read it drop at 53 for the phiX lane. The engineer from Illumina checked already several times the peltier and re calibrated its temperature and alignment. Until now the problem isn't solved.
    The quality of the results for read 2 for the phiX lane still fall within the specifications of Illumina on % error Rate but are typically twice as high as the error rate for read 1.
    I checked all the runs we did before this phenomenon started and we always observed almost exactly the same error rate for read 1 and 2.
    As the results are within specifications Illumina pretends their is no problem, but we have the very strong feeling that the machine is running below optimal.
    Does anyone else has observed similar problems and if yes how were they solved?

  • #2
    Did you think about the second read primer hyb, maybe something with the PE module?

    Comment


    • #3
      This happened to us once, and our FAS has told us that they are investigating a cause for this type of situation. I'll keep you posted if I hear anything.

      Comment


      • #4
        rgrande

        Did the Illumina’s FAS verified the intensity of laser beams of the device? We had a similar problem with our instrument and it was solved by changing the laser engine because one of the two laser beams intensity was low (which identifies G and T bases).

        Comment


        • #5
          Did the Illumina’s FAS verified the intensity of laser beams of the device? We had a similar problem with our instrument and it was solved by changing the laser engine because one of the two laser beams intensity was low (which identifies G and T bases).

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Essential Discoveries and Tools in Epitranscriptomics
            by seqadmin




            The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
            04-22-2024, 07:01 AM
          • seqadmin
            Current Approaches to Protein Sequencing
            by seqadmin


            Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
            04-04-2024, 04:25 PM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, Yesterday, 08:47 AM
          0 responses
          12 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-11-2024, 12:08 PM
          0 responses
          60 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 10:19 PM
          0 responses
          59 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 04-10-2024, 09:21 AM
          0 responses
          54 views
          0 likes
          Last Post seqadmin  
          Working...
          X