Hi !
We want to sequence 12 plasmids. 10 libraries were constructed using Ninblegen Seq Cap EZ Library SR (single index) and 2 using Nextera (dual index). We are planning to pool all these libraries and performing sequencing using a MiSeq Nano Kit v2 flowcell. Could I sequence libraries constructed using kits from different manufactors in a single lane ? Thanks Last edited by frossit; 10-14-2016 at 07:58 AM. |
![]() |
![]() |
![]() |
#2 |
Jafar Jabbari
Location: Melbourne Join Date: Jan 2013
Posts: 1,238
|
![]()
You can, but note following:
1- Libraries with shorter insert size will sequence more efficiently so you have to load more from larger libraries if you want relatively equal number of reads 2- Index read and demultiplexing: if your Nextera libraries index 1 are different then you can do only one index read otherwise you have to do two index reads. I am not sure if you can mask the index 2 for libraries with only one index with MiSeq software but it can be done with bcl2fastq |
![]() |
![]() |
![]() |