We do a small number of nextera libraries. Not enough to justify really delving into the kit and doing lots of experimentation to reduce costs. But I'm always wanting to cut the easy costs. So was thinking of just doing 1/5 reactions (follow their protocol but 1/5 the volumes for everything).

I'd guess that this is fairly common way to cut costs. So am posting to request any tips or cautions from people doing this before we start?
thermophile is offline   Reply With Quote
Old 08-04-2016, 02:55 PM   #2
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,238
Default

There are few papers on this subject. Generally reduce reaction volumes and input in proportion, strip transposase by adding SDS directly followed by PCR with mix of Illumina index primer and P5, P7 primers (flow cell binding motives). PCR reactions need to be done with a third paty polymerase at 50 ul for good yield.
nucacidhunter is offline   Reply With Quote
Old 08-05-2016, 05:44 AM   #3
Friesac
Junior Member
 
Location: Columbus, OH

Join Date: Sep 2009
Posts: 1
Default

I've been looking at this paper to do the exact same thing... http://journals.plos.org/plosone/art...l.pone.0128036. My reasoning is small genomes/high throughput. I'd be interested to hear how it goes for you thermophile.
Friesac is offline   Reply With Quote
Old 08-05-2016, 07:32 AM   #4
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 243
Default

We don't do enough to justify rewriting the protocol that much. Our last run with 1/2 reactions looked identical to full reactions and used the same procedural steps so I think I'll stick to that unless we suddenly are getting full plates of genomes to prep.

thanks for the replies
thermophile is offline   Reply With Quote
Old 08-09-2016, 02:21 AM   #5
Simone78
Senior Member
 
Location: Basel (Switzerland)

Join Date: Oct 2010
Posts: 208
Default

Quote:
Originally Posted by thermophile View Post
We don't do enough to justify rewriting the protocol that much. Our last run with 1/2 reactions looked identical to full reactions and used the same procedural steps so I think I'll stick to that unless we suddenly are getting full plates of genomes to prep.

thanks for the replies
I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit (exactly the same protocol, no changes). Tagmentation is done in 2 ul, I add 0.5 ul NT (or 0.2% SDS) followed by 1 ul i5/i7 (1:5 dilution to avoid excess adaptors and save money) and 1.5 ul NPM. Input is 50-150 pg/rxn.
Even without a robot this is perfectly doable with a pipette.
I this way a 3000 USD kit will last for about 1000 samples instead of 96!

/Simone
Simone78 is offline   Reply With Quote
Old 08-09-2016, 07:24 AM   #6
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 243
Default

Thanks for that! Our liquid handler doesn't go under 1ul, but good to know you were able to just reduce vol.
thermophile is offline   Reply With Quote
Old 08-11-2016, 01:22 AM   #7
docsascha
Junior Member
 
Location: Austria

Join Date: Dec 2013
Posts: 5
Default

Quote:
Originally Posted by Simone78 View Post
I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit ....

/Simone
Dear Simone

Thanks for this info, which robot are you using, and how do you deal with magnetic beats separation and shaking, we have seen here some problems on our way for cutting cost

Does anybody have some tips on how to use a more time saving method than Qbit for DNA measurement? We would like to use a plate reader or similar to reduce hands-on time

Alexander
docsascha is offline   Reply With Quote
Old 08-11-2016, 05:28 AM   #8
gabrieltw
Member
 
Location: Boston

Join Date: Mar 2013
Posts: 24
Default

Quant-iT PicoGreen dsDNA Assay Kit can do the plate format
gabrieltw is offline   Reply With Quote
Old 08-11-2016, 08:29 AM   #9
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 243
Default

Further reduce costs for picogreen (which is already much cheaper than qubit). You can do 50 ul reactions of picogreen (25 dilute sample, 25 standard diluted reagent) in 384 well plates, rather than 200ul reactions that manufacturer suggests.
thermophile is offline   Reply With Quote
Old 09-26-2016, 05:55 PM   #10
SAV
Junior Member
 
Location: San Diego

Join Date: Nov 2012
Posts: 3
Default

Please see this publication showing nextera XT being prepared in 2, 4 and 8 uL and results for single cell RNAseq being compared and showing no loss of complexity.
http://jla.sagepub.com/content/early...216630741.full

And this one having samples prepared using same method but not mentioned in detail in methods.
http://www.cell.com/cell/pdfExtended...674(16)30740-1
libraries were normalized and prepared in 4 uL total volume.
SAV is offline   Reply With Quote
Old 09-26-2016, 11:59 PM   #11
Simone78
Senior Member
 
Location: Basel (Switzerland)

Join Date: Oct 2010
Posts: 208
Default

Quote:
Originally Posted by docsascha View Post
Dear Simone

Thanks for this info, which robot are you using, and how do you deal with magnetic beats separation and shaking, we have seen here some problems on our way for cutting cost

Does anybody have some tips on how to use a more time saving method than Qbit for DNA measurement? We would like to use a plate reader or similar to reduce hands-on time

Alexander
I use a Tecan Freedom Evo 200 with a 384 head. To save time and avoid (decrease) cDNA losses I don not perform any ethanol wash after binding. I just remove the supernatant and elute the cDNA from the beads. Since I am resuspending the cDNA in 15 ul (and I dilute it even more for the library prep) the leftover salt from the RT/preampl does not affect the efficiency of the downstream reactions.
Simone78 is offline   Reply With Quote
Old 05-22-2017, 02:10 AM   #12
Arthur45
Junior Member
 
Location: London

Join Date: May 2017
Posts: 7
Default

Quote:
Originally Posted by Simone78 View Post
I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit (exactly the same protocol, no changes). Tagmentation is done in 2 ul, I add 0.5 ul NT (or 0.2% SDS) followed by 1 ul i5/i7 (1:5 dilution to avoid excess adaptors and save money) and 1.5 ul NPM. Input is 50-150 pg/rxn.
Even without a robot this is perfectly doable with a pipette.
I this way a 3000 USD kit will last for about 1000 samples instead of 96!

/Simone
Dear Simone and other Nextera users,

I was wondering how to go about the Tagmented DNA Clean-up in the Nextera protocol. It uses a Zymo spin column, which I guess will not be possible for a smaller volume (e.g. 2-10 ul). What do you do at this stage? Do you clean with beads? Skip the whole clean-up? Dilute the Tagmented DNA in buffer to make up for the original 50ul volume needed?

Thanks,

Arthur


Would it b
Arthur45 is offline   Reply With Quote
Old 05-22-2017, 06:19 AM   #13
Simone78
Senior Member
 
Location: Basel (Switzerland)

Join Date: Oct 2010
Posts: 208
Default

Quote:
Originally Posted by Arthur45 View Post
Dear Simone and other Nextera users,

I was wondering how to go about the Tagmented DNA Clean-up in the Nextera protocol. It uses a Zymo spin column, which I guess will not be possible for a smaller volume (e.g. 2-10 ul). What do you do at this stage? Do you clean with beads? Skip the whole clean-up? Dilute the Tagmented DNA in buffer to make up for the original 50ul volume needed?

Thanks,

Arthur


Would it b

Hi Arthur,
no need to clean up the tagmented DNA before the final PCR. I do the tagmentation reaction in 2 ul, add 0.5 ul of NT buffer (or 0.2% SDS) and simply go on with the PCR mix (2.5 ul). 0.2% SDS will remove the Tn5 from the DNA and, after adding the PCR mix, will be diluted down and wonīt interfere with the final PCR.
Only after PCR you will do a clean up and, if you want to multiplex tens or hundreds of cells, I would suggest to pool all of them and then do the cleanup, not the other way around.
Best,
Simone
Simone78 is offline   Reply With Quote
Old 05-23-2017, 09:25 AM   #14
Arthur45
Junior Member
 
Location: London

Join Date: May 2017
Posts: 7
Default

Quote:
Originally Posted by Simone78 View Post
Hi Arthur,
no need to clean up the tagmented DNA before the final PCR. I do the tagmentation reaction in 2 ul, add 0.5 ul of NT buffer (or 0.2% SDS) and simply go on with the PCR mix (2.5 ul). 0.2% SDS will remove the Tn5 from the DNA and, after adding the PCR mix, will be diluted down and wonīt interfere with the final PCR.
Only after PCR you will do a clean up and, if you want to multiplex tens or hundreds of cells, I would suggest to pool all of them and then do the cleanup, not the other way around.
Best,
Simone
Hi Simone,

thank you for the message - this is so much easier than expected!

You mention after the PCR I should do a bead clean-up. The original protocol mentions to add twice the beads to the PCR product :

4 Add 30 μl AMPure XP beads to NAP2.
5 Add 30 μl AMPure XP beads to NAP2.
To me, this ratio of 20:30+30 looks like a mistake. Say I have 11ul of PCR products, would I add twice 16.5ul?

I hope someone can help me out there.

Cheers,
Arthur
Arthur45 is offline   Reply With Quote
Old 08-04-2017, 09:49 AM   #15
Arthur45
Junior Member
 
Location: London

Join Date: May 2017
Posts: 7
Default

Hi Simone,

thank you for your answer. I experimented with scaling-down the Nextera protocol, with mixed results.

My main worry is that the transposome activity is not behaving as expected. See the TapeStation traces, before and after tagmentation. The 1kb peak has disappeared during tagmentation, which is good news (we are aiming for a 300bp library). But I fail to be convinced that there is any DNA left after tagmentation, looking at the TapeStation electropherogram.

What do you think about it?

Cheers,

arthur
Attached Images
File Type: png 2017_08_before_tag.png (17.5 KB, 16 views)
File Type: png 2017_08_02_after_tag.png (18.7 KB, 13 views)
Arthur45 is offline   Reply With Quote
Old 08-04-2017, 10:52 AM   #16
ScienceGrrl
Junior Member
 
Location: Indianapolis, IN

Join Date: Jul 2015
Posts: 8
Default

Hello Arthur45,

Make sure you have a very accurate quant of your input samples. Nextera is a mass driven setup so if you are off on your initial quant at all, can get results of too large/too small fragments very easily.
ScienceGrrl is offline   Reply With Quote
Old 08-04-2017, 02:23 PM   #17
Simone78
Senior Member
 
Location: Basel (Switzerland)

Join Date: Oct 2010
Posts: 208
Default

hi Arthur 45,
sorry for the misunderstanding. You do need to clean up the cDNA after the first PCR (pre-amplification for the Smart-seq2 protocol). I am not really familiar with the latest version of the Nextera protocol but the 30 + 30 ul beads sounds a bit strange to me.
I know that Nextera is using the Zymo and Nextera XT is not. We decided to not use a column purifcation because itīs cheap and equally effective.
That said, I am not sure what you have there. A single fragment at 1 kb doesnīt look like a standard library (is it a ladder or similar?). Why donīt you just take some tot RNA , letīs say 1 ng, do some pre-ampl and try to tagment that? I agree that the Tape after tagmentation has no library. Donīt really know why, though
Simone78 is offline   Reply With Quote
Old 08-05-2017, 02:18 AM   #18
Arthur45
Junior Member
 
Location: London

Join Date: May 2017
Posts: 7
Default

Hi both,

thanks for your answers

About the 30+30 ul steps: I called Illumina since and the representative agreed that this *might* be a typo. He raised an issue for their protocol team. This team will change the protocol when the next version of the protocol will be out.

About my graphs: the graph with the 1kb peaks is my DNA before starting the library. The other graph is right after the tagmentation. It looks like there is no more DNA after tagmentation. I am starting to think that my input is not of the right concentration like ScienceGrrl said. And also, I read somewhere that DNA should be clean before starting a library, so I will try to clean my phenol-chloroform extractions with a spin column, just in case.

Cheers,
arthur
Arthur45 is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:42 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2021, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO