Old 06-01-2011, 05:00 AM   #1
rahulr1485
Junior Member
 
Location: United Kingdom

Join Date: Apr 2011
Posts: 4
Default ChIP- Different DNA size for Input DNA and ChIP DNA on Agilent Bioanalyzer

I ran one of my input DNA and all my ChIPd DNA on Agilent Bioanalyzer today.And I am getting different DNA sizes for my input and ChIP DNA samples.Input shows a distribution around 100-300bp while ChIp samples is distributed around 900-6000bp. I processed them together ,decrosslinked overnight and cleaned up using Qiagen PCR clean up kit. I diluted the input 360 X to get 100pg/microlitre and the ChIPd DNA samples 10X to 300-500pg/microlitre before running on the Bioanalyzer. I am using Agilent High Sensitivity kit which expired on 3rd May 2011. The ladder ran correctly. So I am not sure if this inconsistency is due to the kit.
What could be the reason for this difference?


Input DNA


ChIP DNA

Last edited by rahulr1485; 06-01-2011 at 08:58 AM.
rahulr1485 is offline   Reply With Quote
Old 06-06-2011, 01:54 AM   #2
mudshark
Senior Member
 
Location: Munich

Join Date: Jan 2009
Posts: 138
Default

theoretically, your IP target binds to long pieces of shearing-resistant chromatin which are heavily under represented in your input material.
however, i guess there are other problems. what is your IP target? what's your total yield of DNA in the IP as compared to the input amount?
mudshark is offline   Reply With Quote
Old 06-06-2011, 03:11 AM   #3
rahulr1485
Junior Member
 
Location: United Kingdom

Join Date: Apr 2011
Posts: 4
Default

Quote:
Originally Posted by mudshark View Post
theoretically, your IP target binds to long pieces of shearing-resistant chromatin which are heavily under represented in your input material.
however, i guess there are other problems. what is your IP target? what's your total yield of DNA in the IP as compared to the input amount?
My IP target is transcription factor which was found to be induced during the treatment. Concentration of IP DNA was around 1 ng/microlitre and Input was around 35 ng/microlitre. Someone suggested that there could be proteins thats slowing down the migration, so I gave proteinase K treatment, cleaned up with qiagen kit and ran it again in triplicates on the agilent. The size distribution remained the same but the concentration of the fragments were different in same samples.Another ,member of my lab had earlier carried out the same experiment and had got size distribution around 100-500 bp along with some higher bp fragments. I also followed the same protocol, but I am getting a very different size distribution.
rahulr1485 is offline   Reply With Quote
Old 06-08-2011, 08:02 AM   #4
mudshark
Senior Member
 
Location: Munich

Join Date: Jan 2009
Posts: 138
Default

do you pre-clear your chromatin before IP? did you ever do a mock-IP + Bioanalyzer?
another possibility for your problem is unspecific DNA fragments attaching to your beads as you appear to get a hell of a lot of DNA in your IP.
mudshark is offline   Reply With Quote
Old 06-08-2011, 09:49 AM   #5
rahulr1485
Junior Member
 
Location: United Kingdom

Join Date: Apr 2011
Posts: 4
Default

Quote:
Originally Posted by mudshark View Post
do you pre-clear your chromatin before IP? did you ever do a mock-IP + Bioanalyzer?
another possibility for your problem is unspecific DNA fragments attaching to your beads as you appear to get a hell of a lot of DNA in your IP.
Thank you for your reply. Yes, I did preclear it for 2 hours. I didn't run mock IP on bioanalyser, however, I checked it on nanodrop and it showed some negative value.
rahulr1485 is offline   Reply With Quote
Old 04-12-2012, 07:45 AM   #6
Pila
Junior Member
 
Location: morges

Join Date: Apr 2012
Posts: 3
Default

Hello,
It is nice to participate to this forum!
I experienced the same problem...does anybody know if we can shear again the DNA after IP or if it is not recommended?
Thank you for your help!!
Pila is offline   Reply With Quote
Old 05-21-2012, 07:11 AM   #7
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

I've just shared a very similar issue that I encountered on my histone modifications ChIPs.

There : http://seqanswers.com/forums/showthread.php?p=74004

For your information.
MGineste is offline   Reply With Quote
Old 05-22-2012, 01:12 AM   #8
Veleno
Member
 
Location: UK

Join Date: Aug 2011
Posts: 16
Default

Quote:
Originally Posted by Pila View Post
Hello,
It is nice to participate to this forum!
I experienced the same problem...does anybody know if we can shear again the DNA after IP or if it is not recommended?
Thank you for your help!!
you could but then your peaks won't be centered around their original position
Veleno is offline   Reply With Quote
Old 05-22-2012, 02:16 AM   #9
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

Quote:
Originally Posted by Pila View Post
I experienced the same problem...does anybody know if we can shear again the DNA after IP or if it is not recommended?
I would not do that, as the identity of this peak is not clear. If you hypothesize it corresponds to unspecific binding, there is not point to spend time and money trying to do anything with it.
MGineste is offline   Reply With Quote
Old 05-22-2012, 11:15 AM   #10
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

Quote:
Originally Posted by rahulr1485 View Post
What could be the reason for this difference?
Naive but useful question : What (precise) type of beads did you use ?
MGineste is offline   Reply With Quote
Old 05-22-2012, 11:17 AM   #11
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

Quote:
Originally Posted by Pila View Post
I experienced the same problem...
Same problem, same question : What (precise) type of beads did you use ?
MGineste is offline   Reply With Quote
Old 05-22-2012, 12:43 PM   #12
Hamid
Senior Member
 
Location: San Diego, CA

Join Date: Sep 2009
Posts: 105
Default

Hi rahulr1485 and MGineste,

What you are noticing is unfortunately an ideal example of epitope destruction during high energy chromatin shearing. What is occurring is that you are over-shearing your samples to the extent that you are destroying the epitope by ripping off the proteins from the DNA. High energy shearing also transfers quite a bit of heat to the sample during processing, contributing the reversing of the cross links during shearing.
The gel/bioanalyzer results look great, but when you IP, you only pull down what has not been sheared extensively. A very good way to confirm that is to IP different shearing time points, and check the results on the Bioanalyzer. Since MGineste is working with histones, which are highly prevalent, a western of the sheared chromatin will not be a good confirmation. if rahulr1485 is working with transcription factors, then a western will illustrate the epitope destruction during the shearing process by running aliquots of the shearing time course on a SDS PAGE for western.
How are you fixing your samples? and how are you shearing chromatin?
Hamid is offline   Reply With Quote
Old 05-24-2012, 05:53 AM   #13
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

Quote:
Originally Posted by Hamid View Post
Hi rahulr1485 and MGineste,

What you are noticing is unfortunately an ideal example of epitope destruction during high energy chromatin shearing. What is occurring is that you are over-shearing your samples to the extent that you are destroying the epitope by ripping off the proteins from the DNA. High energy shearing also transfers quite a bit of heat to the sample during processing, contributing the reversing of the cross links during shearing.
The gel/bioanalyzer results look great, but when you IP, you only pull down what has not been sheared extensively. A very good way to confirm that is to IP different shearing time points, and check the results on the Bioanalyzer. Since MGineste is working with histones, which are highly prevalent, a western of the sheared chromatin will not be a good confirmation. if rahulr1485 is working with transcription factors, then a western will illustrate the epitope destruction during the shearing process by running aliquots of the shearing time course on a SDS PAGE for western.
How are you fixing your samples? and how are you shearing chromatin?
Mmmh... This a very technical as well as speculative explanation.

I may have a basic one : http://seqanswers.com/forums/showpos...10&postcount=2

This is why I'm insisting on the type of beads that have been used.
MGineste is offline   Reply With Quote
Old 05-25-2012, 12:41 PM   #14
Hamid
Senior Member
 
Location: San Diego, CA

Join Date: Sep 2009
Posts: 105
Default

Hi MGineste,

Thank you for that link and data. While your theory might be plausible for people who do ChIP with carrier DNA, quite a few people including my lab, do not use carrier DNA or DNA blocked magnetic beads for obvious reasons. The "technical" explanation I gave, is based on work by us, and several other groups who have noted similar issues.

Thank you

Hamid
Hamid is offline   Reply With Quote
Old 05-29-2012, 06:19 AM   #15
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

Quote:
Originally Posted by Hamid View Post
Hi MGineste,

Thank you for that link and data. While your theory might be plausible for people who do ChIP with carrier DNA, quite a few people including my lab, do not use carrier DNA or DNA blocked magnetic beads for obvious reasons.
The thing that is puzzling me at the moment is that I've been trained to ChIP in a lab very much used to ChIP-Seq approaches. They have always been using DNA blocked beads - even though we can easily foresee potential troubles using them, I agree - without encountering any issue.

Quote:
Originally Posted by Hamid View Post
The "technical" explanation I gave, is based on work by us, and several other groups who have noted similar issues.
I keep your observation in mind. Thanks for the feedback.
MGineste is offline   Reply With Quote
Old 10-11-2012, 12:18 AM   #16
Pila
Junior Member
 
Location: morges

Join Date: Apr 2012
Posts: 3
Default

Hi everybody,
We are using Dynabeads which are not blocked by any carrier DNA so I also think the reason may be due to something else...what I can tell you is that even if you get this strange pattern it does not mean that the sequencing will be bad...
Pila is offline   Reply With Quote
Old 10-11-2012, 12:25 AM   #17
Pila
Junior Member
 
Location: morges

Join Date: Apr 2012
Posts: 3
Default

Actually,
Could it be that what we see corresponds to ssDNA enriched by the IP...since I just red on this forum that ssDNA is running very slowly compare to dsDNA on the DNA chip?
Pila is offline   Reply With Quote
Old 10-11-2012, 10:58 AM   #18
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Post a picture of your gel/bioanalyzer chip. Yes, ssDNA can run larger than dsDNA on a bioanalyzer chip.

--
Phillip
pmiguel is offline   Reply With Quote
Old 11-09-2012, 04:49 AM   #19
MGineste
Member
 
Location: Montpellier, France

Join Date: Feb 2011
Posts: 21
Default

I reproducibly observed the same pattern. I may have an explanation for this, that I already posted there : http://seqanswers.com/forums/showthr...50&postcount=3
MGineste is offline   Reply With Quote
Old 05-17-2013, 08:14 PM   #20
Shubhamoy
Junior Member
 
Location: California

Join Date: May 2013
Posts: 1
Default

Can somebody please tell me what should be the ideal size of frangments after IP if samples are run at bioanalyser/ Agarose gel?
Shubhamoy is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 02:58 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO