  • Atac-Seq - Side qPCR question - cycle number

    Hello, I am a new user who has tried searching for an answer to a problem I am having with Atac-Seq library prep.

    Let me start off by saying I am using the Buenrostro 2013 Nature Protocols paper and using 50,000 FACS purified mouse cells.

    The issue I am having is with the qPCR side reaction requiring MANY more cycles than reported, with others suggesting ~10-15 addition cycles, but my samples requiring ~35-40 extra cycles!

    I have attached two images:

    (1) Amplification plot showing that I need to add ~37 addition cycles to get 1/3 of the peak fluorescence. You will also notice that the first 30 cycles show almost no amplification and then the reaction takes off -- not sure what this means -- potentially very low starting material? I will note that I decided for these samples to run the qPCR using a 2X SYBR Green Master Mix that contains its own Taq (Thermo Power SYBR Green Master Mix, Cat#4368577) that my lab already had as opposed to the SYBR Green I alone, but otherwise followed everything from the Buenrostro manuscript. I did this because someone else at my institute previous did the same and their side qPCR reactions worked fine and only required ~10 extra cycles. I realize now that this may have been a mistake?

    (2) Gel of my amplified final library preps after all PCRs showing that I get a decent smear as reported by some, but don't necessarily have the distinct ~150bp cutoff (sorry for ripping the gel). I end up cutting out 100-1,000bp fragment and gel purifying, followed by PCR purification column, QC, and eventually Seq.


    I have yet to send these samples to seq as they were my test samples before I run my more precious samples. I guess the main question I have is: If this was your experiment would you bother sending them to Seq or does taking this many cycles ruin the experiment?


    I appreciate any advice the community can offer!

    Thanks,

    -J
    Attached Files

Latest Articles

Collapse

  • seqadmin
    Current Approaches to Protein Sequencing
    by seqadmin


    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
    04-04-2024, 04:25 PM
  • seqadmin
    Strategies for Sequencing Challenging Samples
    by seqadmin


    Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
    03-22-2024, 06:39 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by seqadmin, 04-11-2024, 12:08 PM
0 responses
24 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-10-2024, 10:19 PM
0 responses
25 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-10-2024, 09:21 AM
0 responses
22 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-04-2024, 09:00 AM
0 responses
52 views
0 likes
Last Post seqadmin  
Working...
X