Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Long primers for illumina sequencing

    Hello, I was trying to run a pcr for downstream illumina sequencing; and the primers are specifically designed for this purpose (with Illumina adapters and barcodes), so the primers end up been very long (>65 bp).

    The regular primers (without Illumina adapters and barcodes) worked pretty well. However, the products from Illumina primers look like smears but not clear bands on the gel. Anyone have encountered the same problem before? Any ideas about the optimization would help. Thanks!

  • #2
    A simple way to get around any problems associated with long primers is to only use the long primers for as few cycles as possible, before doing the majority of your cycles using shorter primers at the very ends of your amplicons.

    So, you can do say 1-4 cycles with your full-length primers (the fewer the better), purify, then amplify just with the outermost sequences. You can even do it all in the same reaction vessel, with the long primers in a limiting concentration so the shorter primers outcompete after their binding site has been added.

    Comment


    • #3
      Illumina just published a bulletin with a 16s metagenomics protocol - https://my.illumina.com/MyIllumina/B...mic-sequencing. You can easily adapt it for any locus. We've been successfully using an adaptation of this for a while.

      Comment


      • #4
        What are you attempting to amplify?
        We successfully use our ~65bp primers to generate metagenomic libraries without problems (insert ~370bp).
        The Illumina metagenomic method is way too expensive for us (twice as many PCR reactions plus Ilumina adapters are EXPENSIVE).

        Comment


        • #5
          Hi,

          I have used 515F/926R primers for amplifying hypervariable regions of 16SrRNA gene. I see multiple bands in the gel. Why is this so? Has anyone else experienced this? What is the other band(s) ?

          Thanks
          Richa

          Comment


          • #6
            Originally posted by TonyBrooks View Post
            What are you attempting to amplify?
            We successfully use our ~65bp primers to generate metagenomic libraries without problems (insert ~370bp).
            The Illumina metagenomic method is way too expensive for us (twice as many PCR reactions plus Ilumina adapters are EXPENSIVE).
            Hi Tony,
            The step-out adapters are just oligos, so you could order them from some place other than Illumina.

            Are the extra PCR reactions costly? It might work to add the internal, locus-specific, primers and the step-out (index + distal adapter) oligos straight into the initial reaction. Or would that just make a mess?

            --
            Phillip

            Comment


            • #7
              Are you trying to do something like TraDIS? We had a go and ended up using abridged illumina primers for it to work successfully. But even with this when quantified using bioanyliser they were all different lengths. However it did produce data.

              Comment


              • #8
                Originally posted by pmiguel View Post
                Hi Tony,
                The step-out adapters are just oligos, so you could order them from some place other than Illumina.

                Are the extra PCR reactions costly? It might work to add the internal, locus-specific, primers and the step-out (index + distal adapter) oligos straight into the initial reaction. Or would that just make a mess?

                --
                Phillip
                We could have ordered the oligos separately and not used the Illumina ones.
                PCR reactions aren't costly when you do a few - when you have 5,000 samples to process, then it becomes an issue and saving even small amounts per reaction add up to being able to do extra sequencing runs.
                We did also thought about adding all four primers into one reaction but we were worried that funky things might start happening.
                Our dual oligo/dual index approach is working well though, no problems with duff oligos or freeze-thaw issues yet and getting >80% >Q30 on a 500 cycle v2 kit.

                Comment


                • #9
                  Originally posted by TonyBrooks View Post
                  What are you attempting to amplify?
                  We successfully use our ~65bp primers to generate metagenomic libraries without problems (insert ~370bp).
                  The Illumina metagenomic method is way too expensive for us (twice as many PCR reactions plus Ilumina adapters are EXPENSIVE).
                  Hi Tony Brooks

                  I was wondering if you could let me know about the protocol you used for amplicon sequencing on MiSeq?

                  Thanks
                  Mahdi

                  Comment


                  • #10
                    I'll summarise below

                    The primers are as follows. I won't put the P5/P7 sequences in here as the are (c) Illumina. You should be able to find them by Googling. Make sure you use the paired end P5 & P7

                    P5[i5 Index]ACGTACGTACGTGGATTAGATACCCBRGTAGTC
                    P7[i7 Index]AGTCAGTCAGCCACGTCRTCCCCDCCTTCCTC

                    Once PCR is done, we clean up using 0.8X Ampure beads. We do two rounds of clean up, but because we're cheap, we reuse the beads. After you elute in 10mM Tris, add 0.8X of 2.5M NaCL/20% PEG-8000 to re-bind the DNA to the beads. Incubate for 5 mins, remove supernatent, wash 2X with 80% EtOH and finally elute in 30µL. We QC a random selection on the Bioanalyser DNA 1000 chip to make sure we have product and no dimer. Then we quantify with the Qubit and pool for sequencing.

                    The template specific part of the primers are in red. The sequences shown here are 16S 785F and 1175R for interrogation of V5-7. You can replace these with whatever you target, but you'll also need to redesign your read primers if you do.

                    The sequence upstream of this is random sequence that finds no match when we blast it. It allows us to also use long custom primers with the required Tm > 65ºC

                    Read 1 ACGTACGTACGTGGATTAGATACCCBRGTAGTC - spike 5µL of 100µM into well #12 of MiSeq cartridge
                    Index Read GAGGAAGGHGGGGAYGACGTGGCTGACTGACT - spike 5µL of 100µM primer into well #13 of MiSeq cartridge (rev comp of read 2)
                    Read 2 AGTCAGTCAGCCACGTCRTCCCCDCCTTCCTC - spike 5µL of 100µM into well #14 of MiSeq cartridge

                    We use extra long P10 tips in order to be sure we dispense the primers into the mix already in the cartridge. You have to spike in rather than use custom mixes so that the PhiX spike in (we use 10%) is also primed and read.

                    DON'T SET THE MISEQ UP TO USE CUSTOM PRIMERS!! This will make the MiSeq attempt to aspirate from wells #18-#20 which will be empty.

                    Then set the run up as per the MiSeq protocol.

                    Comment


                    • #11
                      Originally posted by TonyBrooks View Post
                      I'll summarise below

                      The primers are as follows. I won't put the P5/P7 sequences in here as the are (c) Illumina. You should be able to find them by Googling. Make sure you use the paired end P5 & P7

                      P5[i5 Index]ACGTACGTACGTGGATTAGATACCCBRGTAGTC
                      P7[i7 Index]AGTCAGTCAGCCACGTCRTCCCCDCCTTCCTC

                      Once PCR is done, we clean up using 0.8X Ampure beads. We do two rounds of clean up, but because we're cheap, we reuse the beads. After you elute in 10mM Tris, add 0.8X of 2.5M NaCL/20% PEG-8000 to re-bind the DNA to the beads. Incubate for 5 mins, remove supernatent, wash 2X with 80% EtOH and finally elute in 30µL. We QC a random selection on the Bioanalyser DNA 1000 chip to make sure we have product and no dimer. Then we quantify with the Qubit and pool for sequencing.

                      The template specific part of the primers are in red. The sequences shown here are 16S 785F and 1175R for interrogation of V5-7. You can replace these with whatever you target, but you'll also need to redesign your read primers if you do.

                      The sequence upstream of this is random sequence that finds no match when we blast it. It allows us to also use long custom primers with the required Tm > 65ºC

                      Read 1 ACGTACGTACGTGGATTAGATACCCBRGTAGTC - spike 5µL of 100µM into well #12 of MiSeq cartridge
                      Index Read GAGGAAGGHGGGGAYGACGTGGCTGACTGACT - spike 5µL of 100µM primer into well #13 of MiSeq cartridge (rev comp of read 2)
                      Read 2 AGTCAGTCAGCCACGTCRTCCCCDCCTTCCTC - spike 5µL of 100µM into well #14 of MiSeq cartridge

                      We use extra long P10 tips in order to be sure we dispense the primers into the mix already in the cartridge. You have to spike in rather than use custom mixes so that the PhiX spike in (we use 10%) is also primed and read.

                      DON'T SET THE MISEQ UP TO USE CUSTOM PRIMERS!! This will make the MiSeq attempt to aspirate from wells #18-#20 which will be empty.

                      Then set the run up as per the MiSeq protocol.

                      Many thanks Tony!

                      Comment

                      Latest Articles

                      Collapse

                      • seqadmin
                        Essential Discoveries and Tools in Epitranscriptomics
                        by seqadmin




                        The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                        04-22-2024, 07:01 AM
                      • seqadmin
                        Current Approaches to Protein Sequencing
                        by seqadmin


                        Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                        04-04-2024, 04:25 PM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by seqadmin, Yesterday, 08:47 AM
                      0 responses
                      12 views
                      0 likes
                      Last Post seqadmin  
                      Started by seqadmin, 04-11-2024, 12:08 PM
                      0 responses
                      60 views
                      0 likes
                      Last Post seqadmin  
                      Started by seqadmin, 04-10-2024, 10:19 PM
                      0 responses
                      59 views
                      0 likes
                      Last Post seqadmin  
                      Started by seqadmin, 04-10-2024, 09:21 AM
                      0 responses
                      54 views
                      0 likes
                      Last Post seqadmin  
                      Working...
                      X