Default Blue Pippen problems

Is anyone else having problems with Blue Pippen's accuracy? Our error % from all our control DNAs tests range from 3-13%.

One Example:
Tight Setting(250bp)
Bioanalyzer average size (bp) (282bp)
12.8% error

What is an acceptable error rate for the Blue Pippen?

We have been using a 20bp offset for our samples (both amplicon and sheared gDNA) and BP control DNA. (So for a 390bp extraction, we use the 370bp tight setting.) This has been working for a while, but not anymore. We contacted Sage Science, who has been helping us a little, but we are still running through loops. Sage says our BP instrument log files are great and that it could be a cassette issue. They also think that since we are using PCR product, we may have single-stranded DNA that may be interfering with how we visualize peaks on the Bioanalzyer. This may be true but our sheared gDNA does not size accurately either.

Sage Science suggests that, for every new sample type, we run a test run to determine the best setting. Does anyone do this?

What other problems have you seen with the Blue Pippen? Is there a better alternative?
kgoglin is offline   Reply With Quote
Old 07-14-2016, 06:44 PM   #2
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,225
Default

One has to take into account size distribution CV, reproducibility% and accuracy% for each Cassette type. Loading quantity, structure of DNA (blunt ended, Y shape, staggered ends) also can affect the size range distribution. On top of all of these also BA has its own specifications and deviation from them also will affect the sizing.

Your example of tight setting at 370bp will collect fragments in 337-403bp. If you use range setting of 350-390 it will not collect 370bp as that setting is narrower than tightest setting possible for which the software gives a narrow warning.

I agree with Sage explanation and their platform is much better than another one that I have tried.
nucacidhunter is offline   Reply With Quote
Old 07-18-2016, 11:32 AM   #3
kerplunk412
Senior Member
 
Location: Bioo Scientific, Austin, TX, USA

Join Date: Jun 2012
Posts: 119
Default

I know that to select small RNA libraries of ~150 bp and exclude adapter-dimer at ~130 bp you must use settings way different from expected (115 bp-170 bp collection range), so I am not surprised that a little optimization may be necessary.

Regarding a better alternative, the only other thing I have tried is the E-gel, and Pippin is so much better it isn't even a comparison. I think the Caliper LabChip is probably a better comparison but I haven't had a chance to try it. I did find this Googling a little though.
kerplunk412 is offline   Reply With Quote
Old 07-19-2016, 12:13 AM   #4
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,225
Default

I had a similar result to the linked paper in kerplunk412 post comparing Pippin and LabChip.
nucacidhunter is offline   Reply With Quote
Old 07-22-2016, 10:38 AM   #5
kgoglin
Member
 
Location: San Diego

Join Date: Dec 2014
Posts: 15
Default

Thank you for responding

Sage Science said that they like to see less than 10% error from the control DNA. This seems pretty high to me but if that's the case, then I guess we will have to test more than one setting for each sample type.
kgoglin is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 12:16 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO