Old 02-05-2019, 04:06 PM   #1
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Exclamation iSeq, High Q30% but Low PF%

So there's not a lot of data for iSeq since its pretty new - but if anyone has had this issue on MiSeq or in general sequencing please help!

We did an iSeq run today and the results had high Q30% (94%97%) but a VERY low PF% (19%)

What could cause this kind of variation? I've found answers for the other way around but not this.

We're considering the demultiplexing having gone wrong and we're working to correct that but if anyone else has experienced this error - what was the reason, and if it was demultiplexing how did you resolve or get a higher PF%? Is that something we can calculate as we can't get that data from the machine after our data work.
ShyN is offline   Reply With Quote
Old 02-06-2019, 09:42 AM   #2
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 42
Default

Load up the run with SAV and see what the reported occupancy is. If it's below 90%, you've got too little library in there and need to increase the loading concentration. If it's 96% or higher, you've way over-loaded library and need to do the opposite.
misterc is offline   Reply With Quote
Old 02-06-2019, 01:54 PM   #3
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Default

For some reason, we can't see the reported occupancy on SAV for iSeq but we can for NovaSeq

Last edited by ShyN; 02-06-2019 at 02:03 PM. Reason: Wrong response
ShyN is offline   Reply With Quote
Old 03-09-2019, 03:06 PM   #4
Ahutch08
Junior Member
 
Location: Bay Area, CA

Join Date: Mar 2019
Posts: 1
Default

You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

Did you determine whether your input library concentration was too low or too high?
Ahutch08 is offline   Reply With Quote
Old 03-09-2019, 07:51 PM   #5
kbp
Junior Member
 
Location: san francisco

Join Date: Mar 2019
Posts: 2
Default Used iSeq/MiSeq Cartridges

Hi ShyN and/or others,

Any chance you have a spent iSeq and/or MiSeq cartridge(s) that my colleagues and I could have or borrow, perhaps in exchange for a beer? We are in the Bay Area and could come by to pick up.

Thanks!
kbp is offline   Reply With Quote
Old 03-10-2019, 08:03 PM   #6
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 42
Default

Quote:
Originally Posted by Ahutch08 View Post
You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

Did you determine whether your input library concentration was too low or too high?
SAV 2.4.7 (now released) now gives the %occupancy without having to deal with BS.
misterc is offline   Reply With Quote
Old 03-13-2019, 02:56 PM   #7
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Default

Quote:
Originally Posted by kbp View Post
Hi ShyN and/or others,

Any chance you have a spent iSeq and/or MiSeq cartridge(s) that my colleagues and I could have or borrow, perhaps in exchange for a beer? We are in the Bay Area and could come by to pick up.

Thanks!
Hi there - yes! We have a couple used cartridges that we give you. I've sent you an email via SeqAnswers.
ShyN is offline   Reply With Quote
Old 03-13-2019, 09:52 PM   #8
kbp
Junior Member
 
Location: san francisco

Join Date: Mar 2019
Posts: 2
Default Used iSeq/MiSeq Cartridges

Thanks Shy! Feel free to ping me at kbpa00@gmail.com. Thanks!!!
kbp is offline   Reply With Quote
Old 03-21-2019, 08:05 AM   #9
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Smile

Quote:
Originally Posted by Ahutch08 View Post
You need to load the run into BaseSpace to see SAV metrics for iSeq. Ability to view SAV metrics on instrument is coming in later software update.

Did you determine whether your input library concentration was too low or too high?

We're going to try to so BaseSpace now - we used a local application before but I think its time we give base space a chance.

We ended up concluding that our loading concentration was wayyy too low. So out PF was low because of this. Our %Q30 was high because out of the samples that passed (low as they were), most of them were useable.

Thanks everyone for your input!

Last edited by ShyN; 03-21-2019 at 08:10 AM. Reason: Incorrect response
ShyN is offline   Reply With Quote
Old 03-21-2019, 04:24 PM   #10
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Default

Quote:
Originally Posted by misterc View Post
Load up the run with SAV and see what the reported occupancy is. If it's below 90%, you've got too little library in there and need to increase the loading concentration. If it's 96% or higher, you've way over-loaded library and need to do the opposite.
Hi, can you tell me where you found this data? Thanks!
ShyN is offline   Reply With Quote
Old 03-25-2019, 08:38 AM   #11
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 42
Default

It's more empirically driven. We've been running NovaSeq (also patterned flow cell with this occupancy metric generated by Illumina) for a couple years now. iSeq performs very, very similarly.
misterc is offline   Reply With Quote
Old 04-01-2019, 06:39 AM   #12
a.boese
Junior Member
 
Location: Canada

Join Date: Mar 2019
Posts: 1
Default

I recently started using the iSeq and I had the same problem. I felt that my library prep concentration was too high and I got these results by changing the concentrations (I was using PCR products fragmented with the NEBNext Ultra II FS library prep kit):

200pM % PF 39.26% AVG %Q30 82.64%
100pM % PF 67.54%; AVG %Q30 89.15%
50pM %PF 73.92%; AVG %Q30 91.81%
a.boese is offline   Reply With Quote
Old 04-23-2019, 08:36 AM   #13
ShyN
Junior Member
 
Location: Mountain View

Join Date: Feb 2019
Posts: 7
Smile

Quote:
Originally Posted by a.boese View Post
I recently started using the iSeq and I had the same problem. I felt that my library prep concentration was too high and I got these results by changing the concentrations (I was using PCR products fragmented with the NEBNext Ultra II FS library prep kit):

200pM % PF 39.26% AVG %Q30 82.64%
100pM % PF 67.54%; AVG %Q30 89.15%
50pM %PF 73.92%; AVG %Q30 91.81%
We had some more issues with PF and they were corrected with the loading concentration being changed (raised or lowered depending on the % occupancy). We readily have 70% to 80% PF and the %Q30 have been in the 90th percentile.

The problem really ended up being miscalculation of our starting concentrations via qPCR due to an equation error.

50pM is the perfect loading concentration! Glad you (and I) were able to figure it out
ShyN is offline   Reply With Quote
Old 04-24-2019, 12:45 PM   #14
yoderto
Junior Member
 
Location: Philadelphia

Join Date: Feb 2019
Posts: 1
Default

Hey ShyN-- We've had some yield issues we believe are related to loading concentration. I'm just curious, what plexities have you been running on your iSeq? Thanks!
yoderto is offline   Reply With Quote
Reply

Tags
illumina, iseq, miseq, pf%, q30

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 02:19 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO