  • #16
    Well, if you are getting a high number of "adapter dimers" -- P1-P2 ligation products with no, or very short insert it becomes an issue. Basically they are the devil to get rid of after enrichment PCR. First they will amplify preferentially during PCR, because they are small. Second, even if you do a size selection after PCR, you can't remove all of them, because some of them will anneal back to full length library molecules.

    That is why I would recommend adding a QC step after P2 ligation. If you see some P1-P2 (adapter dimer) length products at that point, I suggest an additional size selection. Ampure would probably do the trick.

    In principle, you could do the size selection after enrichment PCR. But you would want to force the "hitch-hiking" adapter dimers to leave their "hiding place" by running a denaturing gel. But I have never tried this -- probably more trouble that it is worth. Multiple cycles of ampure also help. But this would be for cases where less than 50% of the amplicons are adapter dimers.

    --
    Phillip

    Comment


    • #17
      What do you think is the cause of these unwanted passengers? Titration of P1 down is not helping. So there must be something that is not optimal for either P1 or P2 ligations. Could you think of anything ...? Will titration of P2 make any sense?

      Before optimizing any of these two ligations there is no point adding a QC step or using expensive AMPure beads. If I get a library that is contaminated then AMPure might be a choice but if there is no library amplification then .

      Comment


      • #18
        please help me!

        does anyone here do the detection of GMO using RADSeq?
        am very hurry to do this project but have no idea about it.
        how to use RADSeq to detect GMOs?
        i will be appreciated if anyone can talk with me about this!
        thx!

        Comment


        • #19
          suzhebeibei, RAD-Seq produces sequences from random loci across a genome. It is not usually used to check for the presence of a single region. Would a PCR test work instead?
          Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

          Comment


          • #20
            Originally posted by SNPsaurus View Post
            suzhebeibei, RAD-Seq produces sequences from random loci across a genome. It is not usually used to check for the presence of a single region. Would a PCR test work instead?
            i know,but we can choose the enzymes to control this situation.
            The most urgent things now is that i don not konw how to analyse the data sequenced.would u help me about that?
            PCR is very useful,but it is also time-consuming,low through-put,so we want to search for a innovation method to solve the problem.

            Comment


            • #21
              If you pick an enzyme that digests the DNA of the GMO material, then the RAD-Seq library will have that sequence plus all the other genomic loci that have the cut site. So to analyze the data you could take the RAD-Seq reads and align them to a reference sequence of the GMO. I am still unsure if it offers any advantage over existing methods, though.
              Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

              Comment


              • #22
                Originally posted by SNPsaurus View Post
                If you pick an enzyme that digests the DNA of the GMO material, then the RAD-Seq library will have that sequence plus all the other genomic loci that have the cut site. So to analyze the data you could take the RAD-Seq reads and align them to a reference sequence of the GMO. I am still unsure if it offers any advantage over existing methods, though.

                yeah,you are right.and this is where i also feel confused.
                when i get the sequence,i will firstly align them to the insertion sequence,and pick up the correct reads,and then align them to the genome to pick up the other correct reads.
                all of these reads are what we want to analyse the insertion sites,but the question is how should i analyse these datas,and which kinds of softwares i should use,confused........i have never known about bioinformation before,so.......

                Comment


                • #23
                  Originally posted by jsteed View Post
                  I have been researching RAD sequencing myself, take a look at this website

                  I have found it very useful.

                  hello,i have searched the RADSeq home page,and i want to know how to use the Stacks,can u help me?thx a lot!

                  Comment


                  • #24
                    Gana, I am confused why you shear (Corvaris) your sample. For genotyping by sequencing (Elshire protocol) the restricted fragments are ligated to the barcoded adapters and then amplified with Illumina primers. A titration experiment is run (for each enzyme/species combination) to determine the optimal adapter concentration to use for library construction. If this is old news I apologize, but I would like to know the value of shearing. Thanks!

                    Comment


                    • #25
                      rosatoc, I thought Gana was carrying out RAD-Seq, not GBS. RAD-Seq shears the genomic DNA after ligation of the first adapters so that there are a full set of markers at every fragment size, avoiding any problems from having different fragment size ranges selected between library preps of different sets of individuals.
                      Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

                      Comment


                      • #26
                        Gana, your problem probably started with the use of the NanoDrop right at the beginning when you were determining how much DNA to digest. Use a fluorescent plate reader to quant your DNA.

                        Also, we've put a biotin on the end of the p1 adapters, which when used in combination with streptavidin-coated dynabeads, allows us to visualize ligation efficiency using simple gel electrophoresis.


                        Originally posted by Gana View Post
                        What do you think is the cause of these unwanted passengers? Titration of P1 down is not helping. So there must be something that is not optimal for either P1 or P2 ligations. Could you think of anything ...? Will titration of P2 make any sense?

                        Before optimizing any of these two ligations there is no point adding a QC step or using expensive AMPure beads. If I get a library that is contaminated then AMPure might be a choice but if there is no library amplification then .

                        Comment


                        • #27
                          Originally posted by suzhebeibei View Post
                          does anyone here do the detection of GMO using RADSeq?
                          am very hurry to do this project but have no idea about it.
                          how to use RADSeq to detect GMOs?
                          i will be appreciated if anyone can talk with me about this!
                          thx!
                          You can do this for many GMOs using this kit from BioRad. It uses PCR to amplify sequences frequently inserted as GMOs are made.

                          Does this food contain a genetically modified organism (GMO)? Use PCR and gel electrophoresis to find out!

                          Comment


                          • #28
                            RADseq new cut sites?

                            Does anyone have a script that can use aligned reads from RADseq libraries to identify putative mutations that have created new cut sites or disrupted cut sites between samples?

                            Comment


                            • #29
                              For RAD-Seq libraries, to be sequenced on Illumina MiSeq, what concentration do people typically use? Something akin to 16S metagenomics sequencing? I.E. 4.0 pM with a 10% spike-in of PhiX control?

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Strategies for Sequencing Challenging Samples
                                by seqadmin


                                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                                03-22-2024, 06:39 AM
                              • seqadmin
                                Techniques and Challenges in Conservation Genomics
                                by seqadmin



                                The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                                Avian Conservation
                                Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                                03-08-2024, 10:41 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Yesterday, 06:37 PM
                              0 responses
                              10 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, Yesterday, 06:07 PM
                              0 responses
                              10 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-22-2024, 10:03 AM
                              0 responses
                              51 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-21-2024, 07:32 AM
                              0 responses
                              67 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X