Old 11-01-2019, 08:39 AM   #1
CPHSEQer
Junior Member
 
Location: British Columbia

Join Date: Sep 2019
Posts: 4
Default Unexpected power failure during run

Hey NextSeq500 users, has anyone experienced an unexpected power failure during their run? Have you determined any causes? A instrument power failure occurred 10 hours into my last run.

Tracy
CPHSEQer is offline   Reply With Quote
Old 11-01-2019, 09:44 AM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,015
Default

As in the sequencer just died while the power was still on in building? Do you not use UPS's on your sequencers to filter out power anomalies?
GenoMax is offline   Reply With Quote
Old 11-01-2019, 09:55 AM   #3
CPHSEQer
Junior Member
 
Location: British Columbia

Join Date: Sep 2019
Posts: 4
Default

The power failure was not a result of a utility failure. Yes, the instrument was connected to a UPS. It appears to be an instrument power failure.
CPHSEQer is offline   Reply With Quote
Old 11-01-2019, 10:16 AM   #4
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,015
Default

I assume you have eliminated a faulty UPS as a possible cause? If your sequencer is under a maintenance contract you should contact Illumina tech support ASAP.

Never seen a sequencer just die (out of 20 over years) so something is not right.
GenoMax is offline   Reply With Quote
Old 12-03-2019, 09:28 AM   #5
CPHSEQer
Junior Member
 
Location: British Columbia

Join Date: Sep 2019
Posts: 4
Default Update: NextSeq500 Power Failure

The power failure turned out to be a failure of one the computer fans in the instrument. After 10 hours into the run when the computer started heating up with the processing of data, the processing unit overheated and shutdown.
CPHSEQer is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:26 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO