Default NextSeq550_LowClusterDensity_Low%PF?

A 1.8 pM diverse library was loaded on a high-output reagent kit on the NextSeq550 platform.

I understand that an over-clustered run will result in poor signal purity and a low percentage of clusters passing filter (%PF).

But what would cause a severely under-clustered run (~15 K/mm2) to result in a low %PF (60%)?

I have checked the best focus images of this NextSeq550 run, and they definitely appear to be under-clustered, so the reported cluster density does not seem to be '...an underestimation of the true density due to saturation...'.

Furthermore, other than potential issues with sodium hydroxide concentration, quantification, and the actual sequencing primer, what other causes could result in reduced cluster densities?

A phix run had been run on the instrument and achieved optimal clustering thus ruling out any machine related issues.
7891_ZA is offline   Reply With Quote
Old 12-12-2019, 08:29 AM   #2
Johnwang
Member
 
Location: NE, USA

Join Date: Jun 2016
Posts: 17
Default

how did you quantify your libs for sequencing ? by my experience, only qPCR method works well for lib quantification. I use illumina lib quantification kit from KAPA.
Johnwang is offline   Reply With Quote
Reply

Tags
low % pf, nextseq550, underclustering

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 04:13 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO