  • #46
    Closer examination of optical and current signals indicates:
    1- Current drops at the start of elution which is normal
    2- Elution times are 7:31 and 52:56 for lanes 2 and 3 , respectively, this also is consistent for selected size ranges
    3- There is no optical signal after start of elution, abnormal

    Suggested actions:
    1- Since you do not have access to a BA, if fragments have adapters you may check size range by running PCR using a fraction of eluates and visualising amplicon in gel.
    2- Reviewing if the correct loading buffer was used for the cassette type
    3- Contacting Sage Science tech support (there could be a hardware malfunction)
    Last edited by nucacidhunter; 01-19-2016, 04:11 AM.

    Comment


    • #47
      Thank you very much. I will follow your advises.

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Essential Discoveries and Tools in Epitranscriptomics
        by seqadmin




        The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
        Yesterday, 07:01 AM
      • seqadmin
        Current Approaches to Protein Sequencing
        by seqadmin


        Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
        04-04-2024, 04:25 PM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 04-11-2024, 12:08 PM
      0 responses
      57 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 10:19 PM
      0 responses
      53 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 09:21 AM
      0 responses
      45 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-04-2024, 09:00 AM
      0 responses
      55 views
      0 likes
      Last Post seqadmin  
      Working...
      X