Old 04-16-2014, 04:34 PM   #1
arilindsey
Junior Member
 
Location: California

Join Date: Apr 2014
Posts: 1
Default Nextera Kit -- DNA fragments too small

Hello --

I have been working with the Nextera kit for Illumina to create DNA libraries for genome sequencing. I initially used the recommended 50ng input DNA (quantified with a Qubit). However, what should have been my final library had no DNA, and instead all the DNA was left in the supernatant that comes off the first Ampure XP bead clean up step. When I run that same supernatant on a gel, the DNA fragments range from 80 - 200 bp. I have tried higher concentrations of input DNA (up to 90ng) and still have the same problem. Not sure what to try next -- has anyone experienced this? My input DNA has a high molecular weight -- 20kb+ fragments.

Any help is much appreciated.
Cheers,
Amelia
arilindsey is offline   Reply With Quote
Old 05-19-2015, 07:13 AM   #2
Jfly7
Junior Member
 
Location: Montreal, Canada

Join Date: Nov 2014
Posts: 4
Default

Hi Amelia,
I found this question and I am wondering if you had any luck making adjustments to make longer libraries with Nextera? I prepared libraries with Nextera last year and also got shorter fragment sizes than ideal (200-250 bp) when I would like 300-400 bp. I am wondering if it would be best to use a higher concentration of DNA or lower concentration of enzyme?
Jfly7 is offline   Reply With Quote
Old 05-19-2015, 08:02 AM   #3
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

Single cell RNA-Seq libraries with an input of 1 ng or less of cDNA give Nextera fragment sizes of over 500 bp.

Note with the HiSeq 3000/4000 all libraries have to be less than 500 bp so you might want to reconsider whether you want larger fragments.
NextGenSeq is offline   Reply With Quote
Old 05-19-2015, 08:11 AM   #4
Jfly7
Junior Member
 
Location: Montreal, Canada

Join Date: Nov 2014
Posts: 4
Default

Thanks for the reply.
I am doing genomic DNA library prep and am using HiSeq 2500 100bp PE reads. The reason I want libraries a bit longer than what I got is because a lot of fragments were short and therefore sequenced into the adapter, so we lost a fair amount of sequencing coverage. I would like to avoid this this time around.
Jfly7 is offline   Reply With Quote
Old 05-19-2015, 09:02 AM   #5
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

What I find most annoying about Nextera is the insertion bias.
Every Nextera I've ever seen shows unusual GC bias in the first 15 basepairs (see attached).
Attached Images
File Type: png Nextera.PNG (76.5 KB, 88 views)
NextGenSeq is offline   Reply With Quote
Old 05-19-2015, 09:08 AM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by Jfly7 View Post
Thanks for the reply.
I am doing genomic DNA library prep and am using HiSeq 2500 100bp PE reads. The reason I want libraries a bit longer than what I got is because a lot of fragments were short and therefore sequenced into the adapter, so we lost a fair amount of sequencing coverage. I would like to avoid this this time around.
Just use less Ampure during the clean up.
You will get a wide range of sizes using Nextera. Getting the insert sizes of amplicons that are sequenced to increase in length is just a matter of removing the shorter amplicons. (Shorter amplicons seem to cluster preferentially.)

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-19-2015, 09:17 AM   #7
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by NextGenSeq View Post
What I find most annoying about Nextera is the insertion bias.
Every Nextera I've ever seen shows unusual GC bias in the first 15 basepairs (see attached).
Nextera is Tn5. When you recruit a natural protein to do a job for you, you recruit its natural proclivities as well. Sure they have engineered it a little, but not enough to completely remove its insertion site preferences.

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-21-2015, 11:18 PM   #8
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,230
Default

An interesting patent by Agilent was recently published about role of bivalent cations on Nextera activity.
http://patft1.uspto.gov/netacgi/nph-...S=PN/9,005,935
nucacidhunter is offline   Reply With Quote
Old 05-22-2015, 07:37 AM   #9
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by nucacidhunter View Post
An interesting patent by Agilent was recently published about role of bivalent cations on Nextera activity.
http://patft1.uspto.gov/netacgi/nph-...S=PN/9,005,935
Very interesting!
Does Agilent actually sell this Vibhar Tn kit?
So it is likely that including Mn in Nextera reactions would reduce the site specificity. Interesting...

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-22-2015, 09:48 AM   #10
Simone78
Senior Member
 
Location: Basel (Switzerland)

Join Date: Oct 2010
Posts: 205
Default

Interesting! When using our home-made Tn5 the tagmentation was actually INHIBITED by MnCl2 and worked only with MgCl2 (with 5 mM final being the optimal conc, while higher conc had a deleterious effect).
Thanks for sharing,
Simone
Simone78 is offline   Reply With Quote
Old 05-23-2015, 12:58 AM   #11
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,230
Default

Quote:
Originally Posted by pmiguel View Post
Does Agilent actually sell this Vibhar Tn kit?

--
Phillip

They offer two transposase based kits one for SureSelect QXT capture and another for WGS library prep. There is no info in manuals on transposase used, but for sequencing one must use custom reads and index sequencing primers and use “CTGTCTCTTGATCACA” for adapter trimming which could be full or partial recognition sequence of transposase. Nucleotide BLAST for adapter sequence brings up lots of hits.
nucacidhunter is offline   Reply With Quote
Old 05-25-2015, 07:53 AM   #12
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,315
Default

Quote:
Originally Posted by nucacidhunter View Post
They offer two transposase based kits one for SureSelect QXT capture and another for WGS library prep. There is no info in manuals on transposase used, but for sequencing one must use custom reads and index sequencing primers and use “CTGTCTCTTGATCACA” for adapter trimming which could be full or partial recognition sequence of transposase. Nucleotide BLAST for adapter sequence brings up lots of hits.
From the text of the patent application:

Quote:
Using bioinformatics means, the IRL sequence for the transposase was identified as 5'-ctgtctcttatacacaat-3' (SEQ ID NO:9) and the IRR sequence as 5'-acttgtgatcaagagacag-3' (SEQ ID NO:4). Through experimentation, a modified 19 bp IRR sequence 5'-agatgtgatcaagagacag-3' (SEQ ID NO:5) proved to be more efficient and was used in most experiments.
So the sequence you provide is consistent with Agilent using the Vibrio harveyi IS4-like element.

But, I guess the question still remains -- how non-biased is its insertion? It could be that by adding Mn ions the insertion specificity became lower. It certainly looks that way from the figures in the patent. But it may be just as high as Tn5 in Nextera.

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-25-2015, 08:41 PM   #13
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,230
Default

I have not seen any data comparing insertion bias to sheared DNA or Nextera. But comparing couple of Nextera and QXT FastQC plots (attached image) shows that it also has insertion bias but different from Nextera.
Nextera vs QXT.pdf
nucacidhunter is offline   Reply With Quote
Old 05-26-2015, 09:14 AM   #14
SNPsaurus
Registered Vendor
 
Location: Eugene, OR

Join Date: May 2013
Posts: 505
Default

Thanks for the info, nucacidhunter. I am curious about the QXT kmer plot, though. Looking at the base% plot, the bias is GC at the first 2 nucs then mildly T. How come there are no kmers with that at the start? I guess the kmers shown have some representation across the read. I like to compare kmers at the start to kmers selected from the middle of the read to see differences after normalizing the kmer content of the genome. Pretty interesting, though!
__________________
Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com
SNPsaurus is offline   Reply With Quote
Old 05-26-2015, 09:26 AM   #15
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

Quote:
Originally Posted by nucacidhunter View Post
I have not seen any data comparing insertion bias to sheared DNA or Nextera. But comparing couple of Nextera and QXT FastQC plots (attached image) shows that it also has insertion bias but different from Nextera.
Attachment 3809
That looks like unbalanced barcodes to me.
NextGenSeq is offline   Reply With Quote
Old 05-28-2015, 09:01 AM   #16
DRYTCYV
Member
 
Location: Portugal

Join Date: Apr 2011
Posts: 50
Default

Hi,

Anyone tried KAPA HyperPlus Kits?
They say that this is the direct nextera competitor.

Cheers,
DRYTCYV is offline   Reply With Quote
Old 05-28-2015, 11:31 PM   #17
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,230
Default

It could be considered a competitor to NEB dsDNA Fragmentase but not Nextera. Nextera does not require end repair and adapter ligation unlike Hyper.
nucacidhunter is offline   Reply With Quote
Old 05-29-2015, 10:17 AM   #18
SNPsaurus
Registered Vendor
 
Location: Eugene, OR

Join Date: May 2013
Posts: 505
Default

Our local facility looked at the Hyper kit and it looked less biased than Nextera in terms of k-mer preference of the ends.

I think it is a competitor to Nextera since it approaches it in terms of workflow speed and amount of hands-on time needed. At least I am evaluating it as a replacement for our Nextera-based libraries.
__________________
Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com
SNPsaurus is offline   Reply With Quote
Old 09-13-2019, 02:32 PM   #19
SCIL2019
Junior Member
 
Location: Melbourne, Australia

Join Date: May 2019
Posts: 7
Default Larger fragemtns atac

Hi all,

I am having a bit of an issue with this library/sample. I am getting larger fragments peak and then 200 bp peak is gone. This is using 10x ATAC kit.

I usually don;t see such a nhigfh large peak yield, likely because I remove it with dual sided selection. here, I also did dual sided selection but for some reason it's still there...and what's more odd is that I can see the 75 bp dimer too.

Ideas?

Thank you!
Attached Images
File Type: jpg Screen Shot 2019-09-14 at 7.26.12 am copy.jpg (94.6 KB, 5 views)
SCIL2019 is offline   Reply With Quote
Old 09-14-2019, 02:24 AM   #20
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,230
Default

Regardless of size distribution, library does not seem to have typical periodicity expected in ATAC-seq libraries indicating that there was ambient DNA in nuclei stock or nuclei disintegrated during transposition. Presence of 74bp fragment also raises question about accuracy of size selection.
nucacidhunter is offline   Reply With Quote
Reply

Tags
nextera sample prep, whole genome sequencing

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:27 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO