Hi everyone,

we have troubles with our MiSeq Runs. At about cycle 80 we have a massive decrease in Q30scores.In general the quality of the runs is not good. Q30 scores of 50 or lower. In addition, during my last two runs- the intensity does not increase . I am very new to this technique and have therefore attached some Plots from basespace.
Maybe someone had the same problem? Or has an idea about what is going wrong- what are the possible causes? I went through the troubleshooting guides, but could find yet any answer.

thanks :-)
Attached Images
File Type: png Intensities.png (75.1 KB, 39 views)
File Type: png Run Overview.png (103.0 KB, 45 views)
File Type: png Run Overview II.png (56.5 KB, 37 views)
File Type: png Q30 score drop.png (43.3 KB, 40 views)
ATGCT is offline   Reply With Quote
Old 10-31-2017, 03:16 AM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,049
Default

What kind of libraries are these? Are you spiking in phiX and if so at what %? Do you expect to have low nucleotide diversity after cycle 80 in sequencing? This run is also borderline overclustered and that could be one the reasons for this.
GenoMax is offline   Reply With Quote
Old 10-31-2017, 04:14 AM   #3
HESmith
Senior Member
 
Location: Bethesda MD

Join Date: Oct 2009
Posts: 505
Default

The run is significantly over-clustered, as indicated by the low PF (only 68%; should be mid-80s to mid-90s) in combination with the high error rate (4.4%; should be <1%). Quantify your libraries by qPCR and load at the recommended concentration (actually, a bit lower for PE-300bp runs).
HESmith is offline   Reply With Quote
Old 10-31-2017, 07:59 AM   #4
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 243
Default

Are you sure your template is more than 80bp?
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
thermophile is offline   Reply With Quote
Old 11-01-2017, 04:54 AM   #5
MU Core
Member
 
Location: Columbia, Missouri

Join Date: Apr 2008
Posts: 57
Default

Is there a possibility of adapter dimers? May need to perform an additional library clean-up.
MU Core is offline   Reply With Quote
Old 11-01-2017, 05:13 AM   #6
HESmith
Senior Member
 
Location: Bethesda MD

Join Date: Oct 2009
Posts: 505
Default

Short templates/adapter dimers produce a drop-off in signal intensity, which @ATGCT doesn't observe (first thumbnail). And adapter dimers produce a diagnostic spiky signal, also not observed.
HESmith is offline   Reply With Quote
Old 11-01-2017, 11:11 PM   #7
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,232
Default

I wonder if you can post %Base from Data by cycle pan.
nucacidhunter is offline   Reply With Quote
Old 11-06-2017, 10:23 PM   #8
ATGCT
Junior Member
 
Location: Austria

Join Date: Oct 2017
Posts: 4
Default

Hi, thanks for your answers !!!
We are doing Rad-seq libraries. Yes, I am quite sure, that my templates are bigger than 80bp- but we are checking this now using the bioanalyzer kit.
We already contacted Illumina, and they said, that we should increase PhiX concentration and lower our library concetration (from 8pM)- I think because of low diversity. I attached now % Base- at first it looks like problem of low intensities- then I dont know what happens
Attached Images
File Type: png %Base.png (134.0 KB, 31 views)
ATGCT is offline   Reply With Quote
Old 11-06-2017, 10:28 PM   #9
ATGCT
Junior Member
 
Location: Austria

Join Date: Oct 2017
Posts: 4
Default

Ah yes @ GenoMax - we spike with 5% PhiX. I thin we should increase to 10%
ATGCT is offline   Reply With Quote
Old 11-07-2017, 01:08 AM   #10
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,232
Default

Two issues:
1- low sequence diversity at the start which has resulted in low PF%. This should not happen if following standard RAD-seq protocol

2- short inserts and adapter-dimers that start reading into adapter in cycles 80-120

Both can be fixed by more careful library prep.
nucacidhunter is offline   Reply With Quote
Old 11-07-2017, 04:05 AM   #11
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,049
Default

@ATGCT: You should take this data and use BBMap to estimate the insert size. You can do it in different ways. See this link for examples. You likely have very short inserts as @nucacidhunter has said. This will just confirm that observation.
GenoMax is offline   Reply With Quote
Old 11-07-2017, 04:33 AM   #12
ATGCT
Junior Member
 
Location: Austria

Join Date: Oct 2017
Posts: 4
Default

Hi, thanks :-) I will have a look at it. But the insert size should be around 500 bp - I cut them out of a Agarose gel - so I have seen the size- unless something happens after gel extraction. We will have our Bioanalyzer results soon. Then I can also have a look at that. But I will try also BBMap.
Thanks for your suggestions :-) !!!
ATGCT is offline   Reply With Quote
Old 11-07-2017, 04:44 AM   #13
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,049
Default

Don't be surprised. Keep in mind that small inserts preferentially cluster so if they are in there they are attaching to flowcell first.
GenoMax is offline   Reply With Quote
Old 11-07-2017, 07:29 AM   #14
thermophile
Senior Member
 
Location: CT

Join Date: Apr 2015
Posts: 243
Default

your %base also suggest short fragments and reading into the flowcell.
__________________
Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
thermophile is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:43 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO