Old 11-21-2011, 08:18 AM   #1
shawpa
Member
 
Location: Pittsburgh

Join Date: Aug 2011
Posts: 72
Default library gel purification with SYBR gold problems

We are using the Illumina truseq DNA adapter prep kit. We are now doing whole genome and went to using the SYBR gold as illumina suggested. I had used SYBR safe and gotten low yield so they suggested SYBR gold as per protocol. We have done it twice and our ladders and/or samples are not running properly. Our 100bp ladder just looks like a big smear. These were about 2% agarose TAE gels. The ladder is from bioexpress. Is there something about SYBR gold that would be causing this? Is there some change we need to implement? Just to clarify, we add 15ul SYBR gold to about 120ml agarose pre gel-electrophoresis.
shawpa is offline   Reply With Quote
Old 11-21-2011, 10:54 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

We never tried this--the instructions for SYBR-gold imply that it must be used only as a post-stain. I was surprised to see Illumina recommending that it be used during the run.

Also, I don't see how using SYBR-gold instead of SYBR-safe would increase your yield. I had only seen SYBR-gold suggested as a method to avoid the forked-adapter migration artifacts that one sometimes sees. But I would much prefer ETHANol's doing a few cycles of PCR prior to running the gel method to swapping out my running dye.

--
Phillip
pmiguel is offline   Reply With Quote
Old 11-22-2011, 04:59 AM   #3
bbeitzel
Member
 
Location: Ft. Detrick, MD

Join Date: Aug 2008
Posts: 50
Default

You may need to cut back on the amount of ladder that you are loading. I have seen that when using very sensitive dyes (GelStar, GelRed, most of the Sybr stains) you need to cut back on the amount of ladder that you use by 5-10x. If you use the same amount as you would for an EtBr gel, you end up with a big smear.

I agree with pmiguel, too, as I thought Sybr-Gold is supposed to be used for post staining only.
bbeitzel is offline   Reply With Quote
Old 01-16-2012, 12:25 PM   #4
Gina_P
Member
 
Location: New York

Join Date: Dec 2011
Posts: 20
Default

I have used SYBR gold with great success, mixed into the gel as Illumina suggests. I used Benchtop ladder, which was also suggested in the manual, which does resolve quite clearly. It could be that the Bioexpress ladder is created in such a way that it does not resolve well with SYBR gold staining (maybe it has overhangs) - try looking up that information or maybe using less ladder as the others suggest.
__________________
- Gina
Gina_P is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:44 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO