Old 09-15-2011, 03:31 PM   #1
Anthony.287
Member
 
Location: Ohio

Join Date: Dec 2010
Posts: 95
Default Enrichment Numbers

I have a few new questions about enrichment beads that Im hoping you folks can help me with. In the SV protocols, the number of capture beads in the Lib-A manuals is different from those in the Lib-L manuals (2.0x10^6 vs. 2.4x10^6). Only the SV numbers are different, not MV or LV. Does anyone know if the numbers are actually different, or if this is a typo?
Also, during our training, we were taught to calculate bead enrichment values differently than the manuals describe it, and were given Excel spreadsheets with the calculations we were taught. Basically, the manuals use the enriched beads/total number of input beads; we were taught to use the enriched beads/actual beads recovered. I was wondering how other users are determining enrichment %s, and what percentage range tends to give good results with the calc we were taught?
Thanks!
Anthony
Anthony.287 is offline   Reply With Quote
Old 09-16-2011, 06:21 AM   #2
uhany
Junior Member
 
Location: York

Join Date: Apr 2010
Posts: 6
Default

Hi Anthony,

Yes capture beads in Lib A protocol is different and that is ok. I determine by enriched beads/total no. of beads and usually get good results from up to 30% recovered emulsions. thanks,
uhany is offline   Reply With Quote
Old 09-19-2011, 01:38 PM   #3
Nicole 454 Sequencing
Member
 
Location: Branford

Join Date: Apr 2011
Posts: 19
Default

Hi Anthony,

The numbers in the protocols are accurate and the Small Volume Lib-A protocol does actually use fewer beads.

As far as enrichment percentage, either method can be utilized. The calculation in the manual is assuming there is 100% bead recovery. As the manual states, calculating the % breaking efficiency (Section 3.6.1, Step 8; GS FLX Titanium Series emPCR Method Manual – Lib-A SV) is an optional step, therefore the final calculations are based off the assumed recovery of 100% (2.0E6 million Capture Beads).

For a single Small Volume emulsion, you will want to be in the range of 100,000 – 400,000 enriched Capture Beads for best quality.

Feel free to contact your local Roche Representative with any further inquiries. http://my454.com/contact-us/technical-support.asp

Nicole
___________________________
Technical Support Scientist
454 Life Sciences, A Roche Company
Nicole 454 Sequencing is offline   Reply With Quote
Old 09-20-2011, 04:09 AM   #4
yaximik
Senior Member
 
Location: Oregon

Join Date: Apr 2011
Posts: 205
Default

What is the number of input beads for GS Jr Lib-L then? The manual gives an example of caclulation for 10e6 - is this the actual input number of capture beads?
yaximik is offline   Reply With Quote
Old 09-22-2011, 05:55 AM   #5
Nicole 454 Sequencing
Member
 
Location: Branford

Join Date: Apr 2011
Posts: 19
Default

Hi Yazimik,

The number of capture beads supplied within the GS Junior emPCR Reagents kit is 10 million. The enrichment calculations given in the manual are based off of this input and suggest between 500,000 and 2,000,000 enriched beads for best sequencing quality.

This information can be found in our Junior user manuals located at http://my454.com/my454/documentation...df-manuals.asp.

Regards,

Nicole
_________________________
Technical Support Scientist
454 Life Sciences, A Roche Company
Nicole 454 Sequencing is offline   Reply With Quote
Old 09-22-2011, 06:26 AM   #6
suefo
Member
 
Location: Ann Arbor, MI

Join Date: Jul 2011
Posts: 36
Default

We run many GS Junior emPCR Lib-L kits. When determining % initial bead recovery we use the 10e6 number. However, we adjust our %enrichment based on the recovered beads and not the original 10e6. This number will trend higher, but hopefully by only a couple of percentage points. For example our %enrichment based on initial bead recovery may be 20.2% (but if we had used the 10e6 number it would have been 18-19%).

PS... I would take any advise "Nicole 454 Sequencing" suggests. She rocks!
suefo is offline   Reply With Quote
Old 09-26-2011, 05:04 AM   #7
Dynamac
Member
 
Location: Cincinnati, OH

Join Date: Aug 2011
Posts: 46
Default Enrichment Too High

During a MV emPCR workup, I have washed the enriched beads up to 15 times to remove the white un-enriched beads, yet I still get enrichments up around 45%, which apparently is too high for good results. Anyone have thoughts of this?

Plus, I consistently underload the PTP regions even though I'm supposedly loading 2E+6 beads per region. Is this somehow related to the enrichment problem? I'm confident that the Coulter Counter is reading correctly; I checked it against concentration standards provided by Beckman.

Barry
Dynamac is offline   Reply With Quote
Old 09-26-2011, 05:09 AM   #8
suefo
Member
 
Location: Ann Arbor, MI

Join Date: Jul 2011
Posts: 36
Default

Quote:
Originally Posted by Dynamac View Post
During a MV emPCR workup, I have washed the enriched beads up to 15 times to remove the white un-enriched beads, yet I still get enrichments up around 45%, which apparently is too high for good results. Anyone have thoughts of this?

Plus, I consistently underload the PTP regions even though I'm supposedly loading 2E+6 beads per region. Is this somehow related to the enrichment problem? I'm confident that the Coulter Counter is reading correctly; I checked it against concentration standards provided by Beckman.

Barry
OK.. I know this is going to sound silly but it's true. Originally we were using a Matrix pipette (set at the slowest possible speed) to transfer our emPCRs to the plate and we got terrible %enrichment (30-50%). We did a side-by-side with a repeater pipette (recommended by 454) and what a difference! We went from 30% enrichment down to 8%. AND - since we made the switch - our SVs pretty much predict the MVs and LVs now within a couple of percent.

As for the loading of the PTP... you load 2million but typically you'll only see a little over 1million. I think that is standard - and if you look at Roche's benchmark, I think they even indicate that.

Good luck!
suefo is offline   Reply With Quote
Old 09-26-2011, 05:38 AM   #9
Dynamac
Member
 
Location: Cincinnati, OH

Join Date: Aug 2011
Posts: 46
Default

Interesting. I used a 1 mL pipette and tried to estimate 100 L per well because I was doing a MV emPCR and the repeater tip wouldn't fit into the MV tube! I suppose I could transfer the emulsion into another tube for the MV runs.

As for the underloading, I'm getting about 750,000 wells per region. So I'm losing about half a million wells per plate.

Thanks for the help.

Barry
Dynamac is offline   Reply With Quote
Old 09-26-2011, 05:45 AM   #10
suefo
Member
 
Location: Ann Arbor, MI

Join Date: Jul 2011
Posts: 36
Default

Quote:
Originally Posted by Dynamac View Post
Interesting. I used a 1 mL pipette and tried to estimate 100 L per well because I was doing a MV emPCR and the repeater tip wouldn't fit into the MV tube! I suppose I could transfer the emulsion into another tube for the MV runs.

As for the underloading, I'm getting about 750,000 wells per region. So I'm losing about half a million wells per plate.

Thanks for the help.

Barry
We're using 1 mL combi-tips for SVs, 2.5 mL combi's for MVs and 5 mL combi's for LVs. And so far the diameter of the barrel seems to fit.
suefo is offline   Reply With Quote
Old 09-26-2011, 08:44 AM   #11
Zwartbie
Junior Member
 
Location: Wales

Join Date: Oct 2010
Posts: 2
Default

Quote:
AND - since we made the switch - our SVs pretty much predict the MVs and LVs now within a couple of percent.
I've been having terrible trouble with the c.p.b values suggested by the SV titration not scaling up to MV - I keep getting far too high enrichment. I'm now off to buy some different tips for my multipipette so it fits in the MV tubes!

Thanks for posting your explanation!
Zwartbie is offline   Reply With Quote
Old 09-26-2011, 09:22 AM   #12
Dynamac
Member
 
Location: Cincinnati, OH

Join Date: Aug 2011
Posts: 46
Default

Let us know how it works out.
Dynamac is offline   Reply With Quote
Old 11-18-2011, 11:18 AM   #13
Anthony.287
Member
 
Location: Ohio

Join Date: Dec 2010
Posts: 95
Default

Thanks for all of the great help and info!
I have some questions about the comparison between the Matrix pipette and the repeater...was the Matrix set to pick up more volume than there actually was? I've noticed with our Matrix, which was programmed to fill 1250ul, that once it drew up all of the emulsion in the tube, it would continue to fill, since there was less than 1250ul of emulsion. This would cause the emulsion to bubble at the top of the tip, and I wonder if this could be breaking the micro-reactors. I too have not had any consistency between emPCR's, even from one SV to another, with the exact same library dilutions that have not gone through any freeze-thaw cycles between emPCRs. It would be wonderful if I could use the data from a titration to perform a bulk run, but that doesn't seem to be the case.
Also, doesn't the repeater dispense a lot faster than the Matrix? We have one that I just started using for wash steps, and it would be quite tricky to dispense slowly, with ours at least.
Anthony.287 is offline   Reply With Quote
Old 11-18-2011, 11:25 AM   #14
suefo
Member
 
Location: Ann Arbor, MI

Join Date: Jul 2011
Posts: 36
Default

Anthony,
The Matrix was VERY slow and yes, on the final pickup, we overshot the volume on the draw. The repeator is much much faster - we don't try to slow it down. We just shoot into the wells and move on.
I just did an SV titration this week, selected a cpb and then made a new dilution series for 2 LVs. I'm attaching the linear graph of that. We were pretty happy with the reproducibility of the repeator pipettor. The triangles are the LVs and the blue diamonds are the SVs.
Attached Images
File Type: jpg SV vs LV titration.jpg (58.6 KB, 18 views)
suefo is offline   Reply With Quote
Old 03-21-2012, 10:29 AM   #15
Christina85
Junior Member
 
Location: Greece

Join Date: Aug 2011
Posts: 7
Default

@Anthony --> I am determining enrichment % based on the actual beads receovered, because I think that assuming 100% bead recovery is not the best way to proceed.

@Dynamac --> In my SV workup, I have stopped washing the enriched beads at the 6th wash (because I was following the protocol that recommended 3-6 times) even though there were still white un-enriched beads. In the end, the enrichments were very high (40%-70%). Do you think I should have kept washing?

@suefo --> Do you think that the decrease in the enrichment was due to the repeater pipette and it wasn't anything else? I am not questioning you, I am asking because I am very stressed in finding out what went wrong with my emulsions.

I have posted all the details of my emPCR on "Calculation of % Bead enrichment (454 FLX Titanium)": http://seqanswers.com/forums/showthread.php?t=18588

Thank you!!
Christina85 is offline   Reply With Quote
Old 03-21-2012, 10:53 AM   #16
Dynamac
Member
 
Location: Cincinnati, OH

Join Date: Aug 2011
Posts: 46
Default

Christina - Those enrichment numbers seem a little high to me, so I would keep washing. As for my problem, it seemed to work itself out when I went from 5 copies/bead down to 2 copies/bead during amplicon sequencing (did not do a titration).
Dynamac is offline   Reply With Quote
Old 03-21-2012, 11:11 AM   #17
suefo
Member
 
Location: Ann Arbor, MI

Join Date: Jul 2011
Posts: 36
Default

Christina85... I am ABSOLUTELY sure it was the pipettor. I have photos and data to support this is a side-by-side experiment we did. We checked everything before hand as well... pcr plate, thermal cyclers, caps for the pcr plates, etc and we had Roche is 3 times. It wasn't until 454 came in with Roche and made this recommendation - now everything works like a dream. And there is correlations between our SVs and LVs (almost perfect). I can send you data / photos if you want to see them. Email me at sfoltin@umich.edu.
suefo is offline   Reply With Quote
Old 03-21-2012, 01:11 PM   #18
Anthony.287
Member
 
Location: Ohio

Join Date: Dec 2010
Posts: 95
Default

One thing I've been doing to increase accuracy (I hope) is perform 2 volume measurements and 2 bead counts for each tube, and then use the averages. Measuring ~1200ul with a pipette is pretty tricky, and I've found that 2 back-to-back measurements, with the same pipette, tip, and user can give differences up to 50ul, and two back-to-back bead counts (using different Accuvettes and different 3ul aliquots) can give differences up to 500,000 beads. Little deviations like these can certainly add up and make troubleshooting difficult, so I'm doing it all in duplicate.
Anthony.287 is offline   Reply With Quote
Old 03-22-2012, 04:30 AM   #19
Christina85
Junior Member
 
Location: Greece

Join Date: Aug 2011
Posts: 7
Default

What do you mean exactly by
Quote:
Originally Posted by Dynamac View Post
I went from 5 copies/bead down to 2 copies/bead during amplicon sequencing (did not do a titration).
Christina85 is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 09:52 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO