  • Sequencing beyond the end of the fragment

    Hi everyone

    We've had a problem with a sequencing run and I would like to ask for a re-run with justification.

    We submitted 4 libraries to size-select via pippin prep, pool and run SE 150 on a MiSeq. The pippin prep was to select between 200-600bp fragments, and apparently worked well (image of one of the libraries attached, the others looked the same).

    The run came back and it's pretty obvious that something went wrong. between 70-80% of the reads are primer dimers, sized 135bp. I'm attaching the FastQC per-base quality. The first 6bp were in-line barcode. After that there is a big drop in quality at the 77bp mark, which is the end of the primer dimer fragment. However, the fragments keep giving base calls, albeit at very poor quality. Is this what the software does when the reads go beyond the end of the fragment?

    The discrepancy between the reads and the bioanalyzer profile post-pippinprep makes me think that they messed up and loaded the pre-pippinprep sample into the flow cell. But I can also imagine that the bioanalyzer could give a wrong size profile in some conditions, or that there could be some primer dimer contamination that would not be visible on bioanalyzer.

    Does anyone have experience in this kind of problem? Should the read lengths from the MiSeq run be concordant with the bioanalyzer profile, or is that too much to ask?

    Thank you in advance!
    Attached Files
    Last edited by zaratieg; 02-28-2019, 07:01 AM.

  • #2
    The trace that you are presenting here, shows fragments that go beyond 600 bp that are even overlapping with the upper marker. Is this picture from the library before size-selection? That's what it looks like.

    Comment


    • #3
      Thanks for your reply. No, this is supposed to be post size selection. I didn't pay attention to that part, I thought that was their size marker. I saw the well defined bump between 200-600 and assumed that that's what was selected. Does that mean that the pippin prep purification failed?

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Current Approaches to Protein Sequencing
        by seqadmin


        Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
        04-04-2024, 04:25 PM
      • seqadmin
        Strategies for Sequencing Challenging Samples
        by seqadmin


        Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
        03-22-2024, 06:39 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 04-11-2024, 12:08 PM
      0 responses
      18 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 10:19 PM
      0 responses
      22 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-10-2024, 09:21 AM
      0 responses
      17 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 04-04-2024, 09:00 AM
      0 responses
      49 views
      0 likes
      Last Post seqadmin  
      Working...
      X