Old 07-11-2014, 12:00 PM   #1
Krina796
Junior Member
 
Location: Washington DC

Join Date: Jul 2014
Posts: 7
Default MiSeq Wash question

We just had our MiSeq installed about a month ago, and haven't yet sequenced anything on it. I've just been performing maintenance washes weekly to keep the fluidics system in good shape.

I just performed a Standby wash this week since the instrument has been idle for over 30 days. Now that I've done the Standby wash, is it necessary to keep doing Maintenance washes every week given that we haven't run any DNA yet on the machine? Or can we just let it go for now, until we actually start running samples? (at that point I think we'd perform a maintenance wash before running the DNA)
Krina796 is offline   Reply With Quote
Old 07-11-2014, 03:45 PM   #2
bunce
Member
 
Location: Perth

Join Date: Sep 2012
Posts: 55
Default

Hi Krina,
Just keep following the wash instructions on the MiSeq - the bottom line is that the more you use the fluidics system the less chance you will have of developing biofilms or blockages. I would continue to do maintenance washes each week.

On a side note we have been using the Illumina bleach wash protocols after runs and have found this to be easy and is a good safeguard to (at least minimise) carry-over. Cheers, Mike
bunce is offline   Reply With Quote
Old 07-14-2014, 05:41 AM   #3
Krina796
Junior Member
 
Location: Washington DC

Join Date: Jul 2014
Posts: 7
Default

Hi bunce,

Thank you for your help, I really appreciate it. Did you find the bleach wash protocol on the Illumina website?

Best,
Krina
Krina796 is offline   Reply With Quote
Old 07-14-2014, 06:14 AM   #4
bunce
Member
 
Location: Perth

Join Date: Sep 2012
Posts: 55
Default

Hi Krina, the wash protocol is attached together with the xml - there maybe a more up-to-date version of this now? I am not sure how easy this is to find on the illumina site. You just need to make sure the xml finds it into the correct place (v2, v3 etc) - some users were putting it in the wrong location apparently.

We have replaced the post-run wash with the bleach protocol... then after the bleach was we run a 'pseudo'-maintainance wash (three times with the same wash solution). We have done the latter to make sure the bleach is flushed - it is probably not necessary but we have always done it.
bunce is offline   Reply With Quote
Old 07-16-2014, 04:48 AM   #5
Krina796
Junior Member
 
Location: Washington DC

Join Date: Jul 2014
Posts: 7
Default

This is very helpful, thank you very much bunce!
Krina796 is offline   Reply With Quote
Old 08-15-2014, 04:54 AM   #6
james hadfield
Moderator
Cambridge, UK
Community Forum
 
Location: Cambridge, UK

Join Date: Feb 2008
Posts: 221
Default How to get the bleach protocol

Hi, I just wrote a new post on my blog, http://core-genomics.blogspot.co.uk/...ng-bleach.html which describes the bleach protocol and how to get it.
james hadfield is offline   Reply With Quote
Old 08-26-2015, 10:53 AM   #7
GA-J
Member
 
Location: USA

Join Date: Jul 2015
Posts: 28
Default Miseq wash question

Does anyone know where I may find miseq wash record from miseq computer? Any folder or file? Thanks in advance.

Another one: if I have next run maybe in 2 or 3 weeks, I don't want to do stand-by wash, what will you suggest? Thanks again.
GA-J is offline   Reply With Quote
Old 08-27-2015, 07:32 AM   #8
Jessica_L
Senior Member
 
Location: Washington, D.C. metro area

Join Date: Feb 2010
Posts: 116
Default

I'm not sure about a log for the wash dates and times on the MiSeq.

If you plan to not run anything for 2 weeks, just do a post-run wash each week and you should be fine. I've heard lately that Illumina does not recommend doing the standby wash and then leaving the instrument idle for multiple weeks-- apparently they prefer if you just do a post-run wash each week with monthly maintenance washes. As others have said already, the more you use the fluidics system, the less likely the instrument is to develop blockages or films in the lines.
Jessica_L is online now   Reply With Quote
Old 08-27-2015, 08:14 AM   #9
GA-J
Member
 
Location: USA

Join Date: Jul 2015
Posts: 28
Default

Thank you, Jessica L, sounds very clear and simple to follow.
GA-J is offline   Reply With Quote
Old 02-24-2016, 05:44 AM   #10
mlo
Junior Member
 
Location: Oxford

Join Date: Feb 2016
Posts: 1
Default

Does anyone know where I may find WashStatus.xml files on the MiSeq?
mlo is offline   Reply With Quote
Old 02-24-2016, 09:29 AM   #11
misterc
Member
 
Location: Livermore, CA

Join Date: Jan 2016
Posts: 31
Default

I think it's root level on the C:\ or D:\ drive.
misterc is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 05:54 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO