Old 08-02-2014, 11:59 AM   #21
bilyl
Member
 
Location: USA

Join Date: Aug 2013
Posts: 52
Default

Quote:
Originally Posted by nucacidhunter View Post
Illumina's stand on max of 1mM NaOH is pretty firm, so it must make a difference. It is easy to achieve this and follow Illumina's instruction. If library concentration is low it can be concentrated. You can also increase loading to 15-16 pM (maybe in small steps). Just have to make sure that low cluster number is not due to over-clustering where RTA struggles to recognise very closely located clusters and do not use them for template generation.
Has anyone ever tried using the NextSeq style of denaturing libraries for the HiSeq/MiSeq? You neutralize the NaOH with an equal amount of 200mM Tris HCl pH 7, then dilute with HT1.
bilyl is offline   Reply With Quote
Old 08-20-2014, 03:31 PM   #22
LabRat89
Junior Member
 
Location: California

Join Date: Jul 2014
Posts: 3
Default

Quote:
Originally Posted by nucacidhunter View Post
Illumina's stand on max of 1mM NaOH is pretty firm, so it must make a difference. It is easy to achieve this and follow Illumina's instruction. If library concentration is low it can be concentrated. You can also increase loading to 15-16 pM (maybe in small steps). Just have to make sure that low cluster number is not due to over-clustering where RTA struggles to recognise very closely located clusters and do not use them for template generation.
Thank you for the response! I ran a different sample set and got roughly 733k/mm with 14 million reads. This is still really low on the V2 considering how we're supposed to get 18M reads. After upping the concentration to 16pM, the cluster density generated was even less at 599k/mm. I'm guessing there's over-clustering now?

Additionally, has anyone used kits past expiration date and still have it work? If so, what is the time frame?

Thanks again!!
LabRat89 is offline   Reply With Quote
Old 08-20-2014, 05:29 PM   #23
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,179
Default

Cluster densities given for Illumina systems are for sequencing libraries prepared with Illumina kits and does not apply to third party kits or methods. This attachment is a useful source for troubleshooting cluster densities issues:
Attached Files
File Type: pdf Optimizing Cluster Densities.pdf (2.77 MB, 189 views)
nucacidhunter is offline   Reply With Quote
Old 08-21-2014, 03:12 PM   #24
LabRat89
Junior Member
 
Location: California

Join Date: Jul 2014
Posts: 3
Exclamation "Needs Attention" Message on BaseSpace

Thank you nucacidhunter for the attachment.

I encountered another problem that I've never seen before. Under BaseSpace, the run is showing "Needs Attention" even though the MiSeq instrument displays successful completion of the run.

Please see attached image. Have you or anyone else encountered this before?

Thank you again!
Attached Images
File Type: png MiSeq BaseSpace.png (5.9 KB, 34 views)
LabRat89 is offline   Reply With Quote
Old 08-26-2014, 10:18 AM   #25
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

The low cluster density is usually caused by your NaOH. Either your NaOH pH is off or you are adding too much Na+ ions. which interfere with library hybridization to the flowcell.

We always check the pH of our NaOH prior to using it by making up a 0.1 M solution of our stock. The pH of 0.1M NaOH should be at least 12.6 or higher.
CO2 from the atmosphere will acidify the NaOH solution over time. Also, make sure you are using the more recent denaturation protocol (attached).

Denature DNA for 4 nM Library
1 Combine the following volumes of sample DNA and freshly diluted 0.2 N NaOH in
a microcentrifuge tube:
• 4 nM sample DNA (5 μl)
• 0.2 N NaOH (5 μl)
2 Discard the remaining dilution of 0.2 N NaOH or set aside to prepare a PhiX control
within the next 12 hours.
3 Vortex briefly to mix the sample solution, and then centrifuge the sample solution to
280 g for 1 minute.
Denature and Dilute DNA
Preparing Libraries for Sequencing on the MiSeq 9
4 Incubate for 5 minutes at room temperature to denature the DNA into single strands.
5 Add the following volume of pre-chilled HT1 to the tube containing denatured DNA:
• Denatured DNA (10 μl)
• Pre-chilled HT1 (990 μl)
The result is a 20 pM denatured library in 1 mM NaOH.
6 Place the denatured DNA on ice until you are ready to proceed to final dilution.

Last edited by NextGenSeq; 08-26-2014 at 10:20 AM.
NextGenSeq is offline   Reply With Quote
Old 08-26-2014, 10:20 AM   #26
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,291
Default

Quote:
Originally Posted by NextGenSeq View Post
The low cluster density is usually caused by your NaOH. Either your NaOH pH is off or you are adding too much Na+ ions. which interfere with library hybridization to the flowcell.

We always check the pH of our NaOH prior to using it by making up a 0.1 M solution of our stock. The pH of 0.1M NaOH should be at least 12.6 or higher.
CO2 from the atmosphere will acidify the NaOH solution over time.
I don't think the Na+ ions cause the issue. It is the OH- ions that are problematic above a certain concentration.

--
Phillip
pmiguel is offline   Reply With Quote
Old 09-10-2014, 02:15 PM   #27
PrestoMKBC
Junior Member
 
Location: Oregon

Join Date: Mar 2013
Posts: 2
Default

Quote:
Originally Posted by LabRat89 View Post
Additionally, has anyone used kits past expiration date and still have it work? If so, what is the time frame?
LabRat89 - I didn't see a response to your last question, but thought I'd chime in - we have tested a couple of "expired" MiSeq kits for a validation experiment, and after a month, you start getting clustering errors. We had a support Tech looking at a different issue with our machine, and he was okay using a flow cell that was a month old, but not the cartridge.
PrestoMKBC is offline   Reply With Quote
Old 07-07-2015, 07:18 AM   #28
lpalacios
Junior Member
 
Location: Bilbao (Spain)

Join Date: Apr 2011
Posts: 5
Default

Dear all,
I am having problems with cluster density in the last 8 libraries prepared with Agilent Sureselect XT kit. I have run more than 40 libraries prepared with this kits with an average of 800 clusters when I loaded 8pM but in the last runs they have decreased to about 250-300 clusters! I have tried reducing NaOH in the denaturalization, the protocol starting with 2 or 4nM, to load 8, 12.5, 15 or 20pM of library... And nothing worked. I have run other libraries prepared with other kits and they work ok, so it is not an issue of the MiSeq. The final purification is done with agentcourt beads and the bioanalyzer traces are ok, similar to runs that worked fine.
Could anybody help me?

Thanks a lot
lpalacios is offline   Reply With Quote
Old 07-08-2015, 07:36 AM   #29
Jessica_L
Senior Member
 
Location: Washington, D.C. metro area

Join Date: Feb 2010
Posts: 116
Default

lpalacios-- if you're fairly certain it's the Sureselect kit, I'd consider the number of freeze/thaws the kit components have been through. If you're using a brand new kit, maybe check whether this is a new kit lot number compared to the other ones you've used. I would assume, though, that if the problem was with the kit, you'd see some problems on the bioanalyzer-- reduced yield, etc.

The other question to ask is how are you quantitating your libraries after construction? Are you using a qPCR method?
Jessica_L is offline   Reply With Quote
Old 07-08-2015, 09:08 AM   #30
lpalacios
Junior Member
 
Location: Bilbao (Spain)

Join Date: Apr 2011
Posts: 5
Default

Dear Jessica, thanks for your replay. I am not absolutely sure about the problem is the agilent kit but other libraries prepared with other kits and quantified and denatured in the same way are working properly. We have prepared 2 batches of libraries with 2 different lots and both failed, they were not used before. In fact, we have repeated 4 samples with a new kit sent by techsupport and we have no clusters... The quantification has been done with Qubit and bioanalyzer, in the same way as the 200 libraries that worked fine in previous runs. In fact, we have tested several concentrations and all of them had the same low number of clusters. Perhaps there is somthing in the pcr2 primer sequences, something that makes the DNA more sensible to denaturation... I don't know but today I have launched a new run and I have obtained low clusters again...
lpalacios is offline   Reply With Quote
Old 07-08-2015, 10:42 AM   #31
Jessica_L
Senior Member
 
Location: Washington, D.C. metro area

Join Date: Feb 2010
Posts: 116
Default

If you're not quantitating via qPCR, you run the risk of this issue as discussed above:

Quote:
Originally Posted by pmiguel View Post

(3)One big problem with fluorimetry over qPCR is that intact amplicons look no different than non-intact amplicons. So if your library is only 10% full amplicons through a failure in ligation, etc. then qPCR will tell you that. But fluorimetry will not.
It's hard to say definitively that this is why your cluster densities are low, but if the problem is not with the library prep kits or with your MiSeq reagents, then library quantitation is worth looking at. I understand that you've not had problems with other libraries, but maybe there's something about the Sureselect method that causes it to be less efficient. qPCR would tell you for sure one way or the other if that was your problem.
Jessica_L is offline   Reply With Quote
Reply

Tags
cluster density, low quantity

Thread Tools

Posting Rules
You may not post new threads