#1 |
Member
Location: UK Join Date: Jan 2014
Posts: 75
|
![]()
Hello,
We are using the 75 cycle 400M kit for paired-end sequencing. Our understanding is that there are 75 read cycles, plus another 16 cycles (2x8 cycles) available for barcode detection. Total of 91 cycles. I know that it is possible to split the read cycles into a *minimum* of 2x37. However, our samples are only single-indexed and only use 6 bases for the index. Thus, in theory, there are another 10-11 cycles of chemistry in these kits, suggesting that 2x42 bp should be possible - plus 6 cycles for indexing. Does anyone have experience with setting up such a custom run? Would we simply specify 2x42 bp in the sample sheet? Many thanks! |
![]() |
![]() |
![]() |
#2 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,080
|
![]()
Here is Illumina's official take on the amount of reagents available in all kits.
|
![]() |
![]() |
![]() |
#3 |
Member
Location: UK Join Date: Jan 2014
Posts: 75
|
![]()
Thanks for the link: it looks like 2x42 + 6 = 90 is within range (91 cycles).
I'll report back if there is a problem. |
![]() |
![]() |
![]() |
#4 |
Member
Location: UK Join Date: Jan 2014
Posts: 75
|
![]()
2x42bp + 6bp index worked perfectly. Approx 190 K/mm clusters, and >95% PF, yielded a total of 480M paired reads. Q-scores look excellent.
I am tempted to try incrementing the read length in successive runs to see at what point the cartridge runs out of reagents...there is bound to be an excess. |
![]() |
![]() |
![]() |
#5 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,080
|
![]()
Don't push your luck. If air gets sucked into sippers etc you could ruin something trying to get that extra cycle of sequence.
|
![]() |
![]() |
![]() |
#6 |
Member
Location: UK Join Date: Jan 2014
Posts: 75
|
![]()
Fair comment. Certainly not wanting to upset the machine. I did see a user on here extend the sequencing run on a MiSeq reagent kit and got more than an extra 50 cycles.
We are currently struggling trying to obtain the FASTQs from BaseSpace...they don't seem to be available for some reason on BaseSpace or BaseMount, and requeueing the job has resulted a FASTQ failure message. We were not the original owners of the run, but ownership was transferred fine (it seemed). Any ideas? |
![]() |
![]() |
![]() |
#7 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,080
|
![]()
Contact Illumina tech support for issues related to BaseSpace permissions (techsupport at illumina.com).
|
![]() |
![]() |
![]() |
#8 |
Junior Member
Location: Australia Join Date: May 2018
Posts: 1
|
![]()
Hey there, just wondering if you ever played around with this to work out how many cycles you could squeeze out of a 75 cycle Nextseq kit?
|
![]() |
![]() |
![]() |