Angry Replace Miseq reagent kit v3 by v2?

Hi all!
I am very new to NGS.
I am using Miseq, Veriseq for PGS. It can do 24 samples at the same time, by using Miseq reagent kit V3, 150 cycles.
I want to run less than 24 sample (ex: 12 samples per run). Could i run 12 samples per run by using Miseq reagent kit V2, 50cycles.
Thank you a lot!
Truong
Truong Dang is offline   Reply With Quote
Old 05-02-2017, 04:40 AM   #2
GenoMax
Senior Member
 
Location: East Coast USA

Join Date: Feb 2008
Posts: 7,077
Default

As long as the reduced output still satisfies your need of certain reads per sample it should be fine. You want to go down from 150 cycles to 50 cycles or was that a typo? There may be subtle differences so if you are doing cross-comparisons you may want to keep that in mind.
GenoMax is offline   Reply With Quote
Old 05-02-2017, 07:05 AM   #3
Truong Dang
Junior Member
 
Location: Vietnam

Join Date: May 2017
Posts: 2
Default

Quote:
Originally Posted by GenoMax View Post
As long as the reduced output still satisfies your need of certain reads per sample it should be fine. You want to go down from 150 cycles to 50 cycles or was that a typo? There may be subtle differences so if you are doing cross-comparisons you may want to keep that in mind.
In one public: Development and validation of a next-generation sequencing–based protocol for 24-chromosome aneuploidy screening of embryos, by Francesco Fiorentino, 2014. They do for this kit. They using Miseq v2, dual in- dex single-end 36 bp reads.

So i think, i can use for our pgs. But we run on less samples (12 samples) replacing 24 samples by Miseq v3 (in the kit).

Could i do so?
Thanks!
Truong Dang is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:07 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO