Old 01-26-2014, 03:34 AM   #1
HarryHaller89
Junior Member
 
Location: Germany

Join Date: Jan 2014
Posts: 5
Default Bioanalyzer: Multiple Peaks after library PCR

Hello,

in some of my prepared sequencing libraries I observe multiple peaks in the bioanalyzer run after library pcr (see attached file). Many of the peaks have an average distance of arround 60 bp. Are these adaptor concatameres or something like this?

Is there any way to avoid these peaks?

I know the peak at 126 bp are adaptor dimers... But this is a different story.

Best Harry
Attached Files
File Type: pdf Bioanalyzer.pdf (53.3 KB, 307 views)
HarryHaller89 is offline   Reply With Quote
Old 01-28-2014, 09:43 AM   #2
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

That is the strangest library I've ever seen. What type of library is this?
NextGenSeq is offline   Reply With Quote
Old 01-28-2014, 09:52 AM   #3
HarryHaller89
Junior Member
 
Location: Germany

Join Date: Jan 2014
Posts: 5
Default

It's a ChIP-Seq library prepared with Illumina's TrueSeq ChIP Prep kit - prepared according to the protocol. I have it in both input samples and IP samples. However, from around 50 samples it happened only in 6.

My first idea was that it depends on the initial amount of DNA used for library preparation, but there is absolutely no correlation (it happend both for the maximal amount 10 ng and for a lower input amount 5-6 ng). The bioanalyzer traces are reproducible, excluding problems with the run. I personally can not believe that these are adaptor concatamers - since these peaks appear up to 1800 bp (would basically mean 30 adaptors ligated...).
HarryHaller89 is offline   Reply With Quote
Old 01-28-2014, 10:16 AM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,236
Default

Did you happen to use the spike-in standards for library construction? Those can give some strange patterns.

--
Phillip
pmiguel is offline   Reply With Quote
Old 01-28-2014, 10:28 AM   #5
HarryHaller89
Junior Member
 
Location: Germany

Join Date: Jan 2014
Posts: 5
Default

No, nothing was added during library preparation.

Samples were prepared in parallel from the same fragmented chromatin and only few of them showed this pattern. All other libraries look fine. It also does not depend on particular adaptors.

Could these peaks be the result of contaminations in the samples?
HarryHaller89 is offline   Reply With Quote
Old 01-28-2014, 10:42 AM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,236
Default

Quote:
Originally Posted by HarryHaller89 View Post
No, nothing was added during library preparation.

Samples were prepared in parallel from the same fragmented chromatin and only few of them showed this pattern. All other libraries look fine. It also does not depend on particular adaptors.

Could these peaks be the result of contaminations in the samples?
Sure. Was there any gel purification step before the ligation? Some of the samples might have been near to a molecular weight marker that contaminated the resulting sample.

--
Phillip
pmiguel is offline   Reply With Quote
Old 01-28-2014, 10:49 AM   #7
HarryHaller89
Junior Member
 
Location: Germany

Join Date: Jan 2014
Posts: 5
Default

No the gel purification was performed after the ligation step. The strange samples were not near the markers (50 bp ladder) on the gels.
HarryHaller89 is offline   Reply With Quote
Old 01-28-2014, 11:46 AM   #8
barrmur
Member
 
Location: Nottingham, England

Join Date: Nov 2009
Posts: 26
Default

Silly question first. Have you run them on the chip a second time? It almost looks like bleed through from the ladder to be honest. Adapter concatamers is also possible but we have done a lot of chip preps and I have never seen this.

Last edited by barrmur; 01-28-2014 at 11:50 AM.
barrmur is offline   Reply With Quote
Old 01-28-2014, 09:59 PM   #9
HarryHaller89
Junior Member
 
Location: Germany

Join Date: Jan 2014
Posts: 5
Default

Quote:
Originally Posted by barrmur View Post
Silly question first. Have you run them on the chip a second time? It almost looks like bleed through from the ladder to be honest. Adapter concatamers is also possible but we have done a lot of chip preps and I have never seen this.
Yes, the pattern is absolutely reproducible and not dependent on a particular position on the chip or anything like this.
HarryHaller89 is offline   Reply With Quote
Old 01-31-2014, 07:33 AM   #10
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

The 126 bp peak is an adapter dimer. The other peaks are likely concatamers of the adapter dimer.

I wouldn't bother sequencing this library. I would just remake it.
NextGenSeq is offline   Reply With Quote
Old 06-20-2016, 11:50 PM   #11
windnature03
Junior Member
 
Location: Hong Kong

Join Date: Nov 2014
Posts: 3
Default

how about sequencing the sample??
Were there any strange result or it is just a problem on bioanalyzer??
windnature03 is offline   Reply With Quote
Reply

Tags
bioanalyzer, library preparation, quality control

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:38 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO