Question Bioanalyzer, empty wells show profiles.

Dear All.
It has occurred in the last couple of run that empty wells in the bioanalyzer actually showed a RNA profile (indistinguishable from a real sample)
Here some details:
-RNA pico chip
-It occurred 3 times (weeks apart) every time in a different well.
-I run a test with only empty wells and one positive control in the central well, there was no contamination.
-The "ghost samples" in one case was very similar to a real sample even though these were not adjoin (well 6 and 10 ).

Has anyone experienced anything like that before that could not be due to a mis-pipetting ( even though I can imagine a mis-pipetting I really don't think so) ?

Thank you very much and have a good day!
Spleen is offline   Reply With Quote
Old 04-13-2015, 10:43 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,236
Default

Quote:
Originally Posted by Spleen View Post
Dear All.
It has occurred in the last couple of run that empty wells in the bioanalyzer actually showed a RNA profile (indistinguishable from a real sample)
Here some details:
-RNA pico chip
-It occurred 3 times (weeks apart) every time in a different well.
-I run a test with only empty wells and one positive control in the central well, there was no contamination.
-The "ghost samples" in one case was very similar to a real sample even though these were not adjoin (well 6 and 10 ).

Has anyone experienced anything like that before that could not be due to a mis-pipetting ( even though I can imagine a mis-pipetting I really don't think so) ?

Thank you very much and have a good day!
If you used the same electrode for a nano-RNA chip prior to running a pico-RNA chip then the pins would need to be very carefully washed. The amount of RNA needed to visualize on a pico-RNA chip is very small.

--
Phillip

Last edited by pmiguel; 04-13-2015 at 10:47 AM.
pmiguel is offline   Reply With Quote
Old 03-04-2016, 11:14 AM   #3
sajoshi
Member
 
Location: cleveland

Join Date: Dec 2015
Posts: 26
Default

Clean the pins with a brush. Alternate with propanol and milliq water a few times and try again.
sajoshi is offline   Reply With Quote
Reply

Tags
bioanalyzer, contamination

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:45 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO