  • Double peak on Truseq RNA-seq library

    Hi All,

    I have prepared some Truseq RNA-seq libraries and i have 2 peaks.

    I have been reading some previous threads that it could be bead carry-over, but I do not really see any beads in my library and I think some of the libraries look a bit different. Do you have any suggestions to what the problem could be?

    I used another protocol than usual to remove rRNA but I do not think it should be the problem (Epicentre - Ribo Zero : http://www.illumina.com/products/rib...-kit-faqs.html)
    Attached Files

  • #2
    Possible reasons:
    1-Too many PCR cycles
    2- Too much input

    I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.

    Comment


    • #3
      Bubble products.
      The good news is that they usually sequence fine.
      TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

      --
      Phillip

      Comment


      • #4
        Originally posted by nucacidhunter View Post
        Possible reasons:
        1-Too many PCR cycles
        2- Too much input

        I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.
        Not a whole lot was done differently, between 4-5 and the others. I believe that the only difference was that 4-5 had 8 cycles and 1-3 had 10 cycles and 6 had 9 cycles. But this was due to the lower amplification test score before amplifying the libraries.

        We use 1000 ng as starting material, is this too high, or is it a too much input for the bioanalyzer you think of?



        Originally posted by pmiguel View Post
        Bubble products.
        The good news is that they usually sequence fine.
        TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

        --
        Phillip
        I like your phrasing - Or so the lore goes. Is there any way to control if this is the case, or can I avoid this by any means?


        THANKS TO BOTH OF YOU FOR YOUR HELP!

        Comment


        • #5
          1000 ng total RNA as input is OK but as you can see the size of larger peak is proportional to the PCR cycle number indicating over cycling. I would think that 7 cycles would have been enough.

          Using Good quality RNA results in high PCR yields and I would suggest using BA DNA1000 kit or diluting the libraries to below 1 ng/ul for running on HS Chip.

          Keeping PCR cycles the same for the samples that are going to be compared for differential expression is also a good practice to avoid PCR induced biases. This is less important if gene expression analysis is not the aim.

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Strategies for Sequencing Challenging Samples
            by seqadmin


            Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
            03-22-2024, 06:39 AM
          • seqadmin
            Techniques and Challenges in Conservation Genomics
            by seqadmin



            The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

            Avian Conservation
            Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
            03-08-2024, 10:41 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, 03-27-2024, 06:37 PM
          0 responses
          12 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-27-2024, 06:07 PM
          0 responses
          11 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-22-2024, 10:03 AM
          0 responses
          53 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-21-2024, 07:32 AM
          0 responses
          69 views
          0 likes
          Last Post seqadmin  
          Working...
          X