Default MiSeq vs HiSeq cluster densities

Those of you have just gotten MiSeq and intending to use them to help dial in cluster densities for their HiSeqs, check your MyIllumina: Bulletin. Especially

Quote:
Cluster efficiency Due to differences in the cluster generation chemistry, the MiSeq flowcells typically cluster at approximately 50% higher cluster density than the same library concentration loaded on a HiSeq flow cell (i.e. a library concentration that yields 750/mm2 clusters on a MiSeq will result in approximately 500/mm2 on a HiSeq).
Can anyone support or contradict this? We just got the MiSeqs, so this would be mission critical for us. I guess we tend to load our HiSeq (or, previously, HiScanSQ) at 12 pM [wrong, it was loaded at 10 pM] (assay by SYBR-green KAPA qPCR with Illumina phiX as our concentration standard.) That tended to land us in the 600-800 Kclusters/mm^2 on the HiScanSQ. On the HiSeq, the phiX control lanes at 12 pM came in way low (450 K clusters/mm^2) whereas the libraries we titrated with them came in higher than expected: 700-1100K clusters/mm^2. But that could just be normal variation.

--
Phillip

Last edited by pmiguel; 05-05-2012 at 07:21 AM. Reason: Corrected loading concentration of phiX used.
pmiguel is offline   Reply With Quote
Old 05-04-2012, 08:34 PM   #2
Heisman
Senior Member
 
Location: St. Louis

Join Date: Dec 2010
Posts: 535
Default

Quote:
Originally Posted by pmiguel View Post
Those of you have just gotten MiSeq and intending to use them to help dial in cluster densities for their HiSeqs, check your MyIllumina: Bulletin. Especially



Can anyone support or contradict this? We just got the MiSeqs, so this would be mission critical for us. I guess we tend to load our HiSeq (or, previously, HiScanSQ) at 12 pM (assay by SYBR-green KAPA qPCR with Illumina phiX as our concentration standard.) That tended to land us in the 600-800 Kclusters/mm^2 on the HiScanSQ. On the HiSeq, the phiX control lanes at 12 pM came in way low (450 K clusters/mm^2) whereas the libraries we titrated with them came in higher than expected: 700-1100K clusters/mm^2. But that could just be normal variation.

--
Phillip
Our sequencing core has used a 1:1.4 ratio between MiSeq:HiSeq (similar to the 1:1.5 ratio you posted above), and they are considering going even higher in their recommendations. Not being part of the core I can't really give details beyond that.
Heisman is offline   Reply With Quote
Old 05-15-2012, 06:34 PM   #3
megalodon
Member
 
Location: Pennsylvania

Join Date: Sep 2010
Posts: 20
Default

Hi Philip,

we have followed Illumina's recommendation of using a 1:1.5 ratio between MiSeq:HiSeq and by doing that, we have been achieving roughly the same cluster densities on the HiSeq as for previous MiSeq runs for the same sample.

Daniela
megalodon is offline   Reply With Quote
Old 05-17-2012, 02:18 AM   #4
RPiddock
Lab Monkey
 
Location: UK

Join Date: Jun 2011
Posts: 16
Default

We've been loading our MiSeq at 6pM and HiSeq at 10pM and getting good results, so very similar all round.

Rachel
RPiddock is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 02:27 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO