Old 02-06-2012, 10:59 PM   #1
nancysch
Junior Member
 
Location: Pleasanton, CA

Join Date: Feb 2012
Posts: 3
Default amplicon sequencing on MiSeq

I am a new Illumina user. What is the best way to do targeted seqeuncing (<100 amplicons) on the MiSeq?
nancysch is offline   Reply With Quote
Old 02-06-2012, 11:19 PM   #2
ECO
--Site Admin--
 
Location: SF Bay Area, CA, USA

Join Date: Oct 2007
Posts: 1,358
Default

How many samples and what size are your amplicons?
ECO is offline   Reply With Quote
Old 02-06-2012, 11:23 PM   #3
nancysch
Junior Member
 
Location: Pleasanton, CA

Join Date: Feb 2012
Posts: 3
Default

to start, just 4 amplicons of sizes 130-200 (without adaptors), and 32 samples.
nancysch is offline   Reply With Quote
Old 02-07-2012, 07:54 AM   #4
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

That's a very low complexity library. You'll probably have to spike a lot of PhiX in to get good results.
NextGenSeq is offline   Reply With Quote
Old 02-07-2012, 08:26 AM   #5
TonyBrooks
Senior Member
 
Location: London

Join Date: Jun 2009
Posts: 298
Default

I've often wondered if you couldn't include some Ns in the primer for low complexity libraries which you could then remove post sequencing along with your target primer. So your fwd primer would look like this

[P5][Seq Primer]NNNNN[Target Primer]

In theory the first 5 bases would be random which would give decent cluster identification.

There's probably some obvious reason why this wouldn't work that I haven't thought of yet.

Or maybe designing PCR's to both strands would increase the complexity enough?
TonyBrooks is offline   Reply With Quote
Old 02-07-2012, 11:35 AM   #6
krobison
Senior Member
 
Location: Boston area

Join Date: Nov 2007
Posts: 747
Default

Quote:
Originally Posted by TonyBrooks View Post
I've often wondered if you couldn't include some Ns in the primer for low complexity libraries which you could then remove post sequencing along with your target primer. So your fwd primer would look like this

[P5][Seq Primer]NNNNN[Target Primer]

In theory the first 5 bases would be random which would give decent cluster identification.

There's probably some obvious reason why this wouldn't work that I haven't thought of yet.

Or maybe designing PCR's to both strands would increase the complexity enough?
A related trick I saw published was to use a variable length sequence between the Illumina primers & targeting sequence; this way the targeting sequences aren't all in phase & the complexity is significantly increased in the eye of the cluster caller.
krobison is offline   Reply With Quote
Old 02-09-2012, 09:31 PM   #7
CoastsideOldRider
Junior Member
 
Location: Coastside

Join Date: Mar 2010
Posts: 1
Default

Do you have the reference for the paper mentioned in your post?

Thanks!
Marc
CoastsideOldRider is offline   Reply With Quote
Old 02-14-2012, 12:14 PM   #8
nancysch
Junior Member
 
Location: Pleasanton, CA

Join Date: Feb 2012
Posts: 3
Default

is it this one?

Kindle et al., Detection and Quantification of Rare Variants with Massively Parallel Sequencing. PNAS doi:10.1073 (April 2011)
nancysch is offline   Reply With Quote
Old 03-28-2012, 12:38 PM   #9
ShiveringFire
Junior Member
 
Location: Athens, GA

Join Date: Oct 2009
Posts: 7
Default Cluster caller?

Quote:
Originally Posted by krobison View Post
A related trick I saw published was to use a variable length sequence between the Illumina primers & targeting sequence; this way the targeting sequences aren't all in phase & the complexity is significantly increased in the eye of the cluster caller.
Can you educate me more on "cluster caller"? I recently got results from MiSeq Paired-End run (150bp) and I suspect there's a double read problem. Qscores are rather wavy at the beginning.

My samples are multiplexed targeted re-sequencing of exomes from 16 genotypes. Could this be due to a low complexity issue discussed on this thread?

I also heard rumors that if the first 4bp of a read is identical (which is very likely in targeted re-sequencing) it will be assigned to the same cluster. Is this true?
ShiveringFire is offline   Reply With Quote
Old 03-29-2012, 09:55 AM   #10
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by ShiveringFire View Post
Can you educate me more on "cluster caller"? I recently got results from MiSeq Paired-End run (150bp) and I suspect there's a double read problem. Qscores are rather wavy at the beginning.

My samples are multiplexed targeted re-sequencing of exomes from 16 genotypes. Could this be due to a low complexity issue discussed on this thread?

I also heard rumors that if the first 4bp of a read is identical (which is very likely in targeted re-sequencing) it will be assigned to the same cluster. Is this true?
I am thinking about this a lot. Especially with respect to possibly streamlining our operations by dumping our GS-FLX.

Our MiSeq is apparently 6 weeks from being delivered, so I can only extrapolate from the performance of our HiScanSQ (which is similar to a HiSeq). There are two issues masquerading as a single "problem".

(1) The first is that the actual instrument focusing software goes nuts and misfocuses if you have an empty tile (or set of tiles?) early in the run. Probably mainly during the first cycle. For our HiScanSQ this means if you don't have a fair number of clusters in the G&T channel and the A&C channel, you have a good chance that the instrument will pick the wrong focal plan to image.
Doesn't anyone have connections inside Illumina they could point out this blunder to? If focus is good for one channel and bad for the other, why not just use the good focal point?
Anyway, I think this is only an issue during the first few cycles (maybe only the first) of a read. Then the focal points seem to be carried along with little (or no) adjustments.
The effect? Once you are out of focus, you are hosed for that tile for the run, I think. (Or "psuedo-tile" -- but I'll just presume that everyone reading this realizes that no masonry is involved in this process and leave it at "tile".) This applies no matter how perfect your cluster spacing is.

(2) If the instrument did not hose up its focal plane during the first cycle, you can still have diminished results if you have a low complexity of base calls. This is the one everyone thinks about -- two adjacent clusters give the same base calls for 4 bases. If they are very close together, then neither of them may "register" (if I use the terminology correctly), well at all. This is a little more understandable. Except that Illumina foreclosed the use of methodology that could circumvent this issue on the GA-IIx for the HiSeq by making the early layers of data processing possible only on the instrument console.

So if issue 1 doesn't kill you, issue 2 probably will. Work-arounds are required.

--
Phillip
pmiguel is offline   Reply With Quote
Old 04-05-2012, 10:16 AM   #11
ShiveringFire
Junior Member
 
Location: Athens, GA

Join Date: Oct 2009
Posts: 7
Default no more deferred cluster calling

Thanks Philip,

I came across this paper formally defining the problem and a possible solution:
http://www.plosone.org/article/info%...l.pone.0016607

Sadly I was told that MiSeq doesn't save raw image files so a "deferred cluster calling" is not possible anymore. This sounds like getting rid of the evidence for a dirty job...

So spiking a lot of pHiX seems like the only solution for low complexity. Can one spike phiX even in custom made libraries?
ShiveringFire is offline   Reply With Quote
Old 04-06-2012, 07:40 AM   #12
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

Yes, you just buy the PhiX control from Illumina and spike it in.
NextGenSeq is offline   Reply With Quote
Old 07-24-2012, 07:48 AM   #13
gnomers
Junior Member
 
Location: Toronto

Join Date: Dec 2009
Posts: 4
Default

I am looking for an alternative to spiking in 50% or more PhiX while sequencing low diversity amplicon libraries.

I am wondering if using longer barcodes with balanced representation of bases (up to 24 bases) would be sufficient.

Additionally, does anyone know if I can specify barcodes of different lengths on a MiSeq sample sheet?
gnomers is offline   Reply With Quote
Old 07-24-2012, 11:25 AM   #14
bbeitzel
Member
 
Location: Ft. Detrick, MD

Join Date: Aug 2008
Posts: 50
Default

I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.

I am pretty sure (but don't quote me on this) that you can specify indeces of different lengths on the sample sheet. I think that the 2 index reads use a 6 base and a 7 base index.
bbeitzel is offline   Reply With Quote
Old 07-24-2012, 11:34 AM   #15
gnomers
Junior Member
 
Location: Toronto

Join Date: Dec 2009
Posts: 4
Default

Quote:
Originally Posted by bbeitzel View Post
I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.
Thanks for your answer. I should have clarified that I am not using truseq/nextera indexing reads; my barcodes compose the very first bases of read1. So, it sounds like I could use balanced barcodes, in lieu of the stretch of 12 Ns they recommended to you.
gnomers is offline   Reply With Quote
Old 08-03-2012, 04:53 PM   #16
groad
Junior Member
 
Location: San franscisco

Join Date: Jul 2012
Posts: 4
Default

Quote:
Originally Posted by bbeitzel View Post
I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.

I am pretty sure (but don't quote me on this) that you can specify indeces of different lengths on the sample sheet. I think that the 2 index reads use a 6 base and a 7 base index.
Hi bbeitzel,

Do you mean to add 12 degenerated nucleotides to the 5' end of gene specific primer? Thanks.

Jim
groad is offline   Reply With Quote
Old 08-05-2012, 06:45 AM   #17
ramsemm
Member
 
Location: USA

Join Date: Jan 2012
Posts: 12
Default

Hi!

I am doing some similar sequencing - 30 samples of 1 amplified region. I was wondering if anyone has a citation or comment on what level is the limit for assigning taxonomy to 100 bp sequences from bacterial 16S (V4) amplicons?

Thank you in advance,
Molli
ramsemm is offline   Reply With Quote
Old 08-06-2012, 01:08 PM   #18
bbeitzel
Member
 
Location: Ft. Detrick, MD

Join Date: Aug 2008
Posts: 50
Default

groad,

If you have incorporated the Illumina-specific sequences (sequence read primer binding sites, bridge PCR sequences, index read primer site) into your primers, then the 12 Ns would come between the Illumina stuff and the gene-specific sequence. If you are adding the Illumina adapters after PCR (ie. running amplicons through the TruSeq protocol), then you would put the 12 Ns at the 5' ends of both PCR primers. They have to be on both primers since TruSeq will not distinguish one end from the other.

Hope this helps.
bbeitzel is offline   Reply With Quote
Old 11-12-2012, 03:15 AM   #19
vl80
Member
 
Location: Asia

Join Date: Nov 2012
Posts: 34
Default

Apologies for reactivating an old thread. I posted a similar question yesterday and thought may be asking here may get more answers.
I plan to sequence fragments ranging from 90 bp - 450 bp using Miseq as a pilot study. Per sample, I would have 2 amplicons. My PCR is a multiplex with > 10 primer pairs per reaction. I am completely new to this field and my theorising was to use the Trueseq kit, skip the fragmentation and end repair steps and proceed with the other steps. However, reading this thread may me doubt the appropriateness of that method. I am worried about suggetion to modify the PCR primers with an N sequence as this may cause mispriming in my multiplex PCR. Any suggestions are welcome. Thanks a lot
vl80 is offline   Reply With Quote
Old 11-14-2012, 04:09 AM   #20
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Modulo any special issues with v2 hardware, the "spike in 50% phiX" seems to be a cure-all. Using v1 hardware we did a v3 loop sequence where the first 20 bases were 100% identical. But the addition of a genomic DNA library allowed this to work fine. So just use that method initially and you should be fine.

--
Phillip
pmiguel is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 12:21 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO