  • Maximum clustered flow cell life?

    What is the longest you have waited after clustering a flow cell (especially a PE flow cell) before running it?

    I have a PE (v3 chemistry) flow cell that was clustered 4 weeks ago (December 21st, 2011). We thought it defective at the time because edge finding failed every time we attempted to run it. Subsequently we found the edge calling issue was a result of a combination of new edge/autotilt methodology in HCS 1.5.0 and an install failure to update a config file with a parameter critical to the new software.

    We plan to re-hybe the 4 week old flow cell tomorrow and attempt to run it. It has been sitting in buffer at 4 oC since December 21st. Hopefully the first base report will tell us whether that is going to be productive or not. I guess I am most concerned that the first read will look okay, but turn around will perform poorly.

    Illumina tech support thinks it is worth a shot. Specifications call for a flow cell to be rehybed (sequencing primer re-annealed) after 10 days. But as to whether a flow cell is still functional at 30 days post-clustering, there seems to be no information.
    --
    Phillip

  • #2
    Well, unless there's DNase or something in there I don't see why the clusters wouldn't still be there. I mean, aren't they covalently linked to the surface? As long as you're doing a rehyb it should still work. Keep us posted, though...

    Comment


    • #3
      There is some additional chemistry embedded in the oligos on the surface of the flow cell that is used during "turn around". I don't know what that is, or if it "degrades" over time.

      --
      Phillip

      Comment


      • #4
        The First Base Report was marginal. I can see clusters and the focus scores are quite high. But the G and T intensities are mostly near the bottom edge of acceptable.

        Decided to run at least until tomorrow. Then recluster if it appears necessary.

        --
        Phillip
        Last edited by pmiguel; 01-23-2012, 04:25 AM.

        Comment


        • #5
          Read1 (101 nt) completed. Results are acceptable. Got >50 billion >Q30 bases from the flow cell surface. (The HiScanSQ will only image the top of the flow cell.) Index read now in progress. Cautiously optimistic.

          Although G signal started very low (just above 1000), it dropped very little during the run. Whereas A and C signal were where one expects them to be, but dropped substantially (as is normal) during the run.

          --
          Phillip

          Comment


          • #6
            Forgot to give the final results: pretty good. The final data set did not appear to suffer much, if at all, from having the flow cell sitting at 4 oC for a month before it was run. Of course, your mileage may vary...

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Essential Discoveries and Tools in Epitranscriptomics
              by seqadmin




              The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
              04-22-2024, 07:01 AM
            • seqadmin
              Current Approaches to Protein Sequencing
              by seqadmin


              Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
              04-04-2024, 04:25 PM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, Yesterday, 08:47 AM
            0 responses
            16 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-11-2024, 12:08 PM
            0 responses
            60 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 10:19 PM
            0 responses
            60 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 09:21 AM
            0 responses
            54 views
            0 likes
            Last Post seqadmin  
            Working...
            X