Default How to define RNA quality for Drosophila RNA using Agilent Bioanalyzer?

Hi all,

I extracted RNA from Drosophila heads, and want to do RNAseq on them. I used Agilent's Bioanalyzer to assess RNA quality, and found an expected double peak at the 18S region; I read that this is common in insects, and due to 28S denaturation, so I am not worried about that. However, I think that due to this double-peak all my RINs are N/A, and I want to ask Drosophilists or insect scientists with experience how they assess RNA quality for library preparation.

I attached an example sample to this thread; all my samples are similar, even though the RNA concentration varies greatly, and the attached sample is my highest concentration (~400 ng/ul); my lowest concentration was about 30 ng/ul, but, again, shows the same characteristic double-peak and no 'shoulder' in the fast region.


Best,

Pat
Attached Images
File Type: png Screen Shot 2017-04-07 at 12.23.37.png (128.1 KB, 31 views)
patkrat is offline   Reply With Quote
Old 04-10-2017, 07:30 AM   #2
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,212
Default

You won't get a RIN score until you clear the error (whatever it is) for the lane.

--
Phillip
pmiguel is offline   Reply With Quote
Old 06-30-2017, 07:50 AM   #3
vandykitty
Junior Member
 
Location: charlotte,nc

Join Date: Dec 2012
Posts: 5
Default

Here is the answer I received from Agilent: Total RNA from several insect species will appear to be lacking 26/28S rRNA. This pattern can be explained by the cleavage of the 28S rRNA during its maturation in the cytoplasm [ Jordan,BR, et al., J. Mol. Biol 101(1):85-105 ]. In Drosophila, the doublet peak results from the processing of the 28S rRNA into 28Sa and 28Sb mature forms that migrate in a similar manner to the 18S rRNA. Unfortunately if that is the case, it won’t be possible to get an accurate RIN from these samples
vandykitty is offline   Reply With Quote
Old 06-30-2017, 11:20 AM   #4
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,212
Default

Quote:
Originally Posted by vandykitty View Post
Here is the answer I received from Agilent: Total RNA from several insect species will appear to be lacking 26/28S rRNA. This pattern can be explained by the cleavage of the 28S rRNA during its maturation in the cytoplasm [ Jordan,BR, et al., J. Mol. Biol 101(1):85-105 ]. In Drosophila, the doublet peak results from the processing of the 28S rRNA into 28Sa and 28Sb mature forms that migrate in a similar manner to the 18S rRNA.
This is all true.

Quote:
Originally Posted by vandykitty View Post
Unfortunately if that is the case, it won’t be possible to get an accurate RIN from these samples
This isn't true. At least we typically see RIN scores from from cleaved large subunit rRNA samples like drosophila and they seem reasonably accurate to me.
However the red dot at the top of the OP's supplied image, above the lane, denotes an error. Until errors are cleared, no RIN will be generated.

--
Phillip
pmiguel is offline   Reply With Quote
Old 06-30-2017, 11:39 AM   #5
vandykitty
Junior Member
 
Location: charlotte,nc

Join Date: Dec 2012
Posts: 5
Default

Thanks, I did have the same problem as the OP and had the error message that there was an unexpected signal in the fast region, so I changed that threshold, resulting in unexpected ribosomal ratio, so I changed that and the error cleared but it did not give me a re-calculated RIN. Do you know if there is something else I can change... because apparently the Agilent rep did not know how to help me troubleshoot.. thank you for any help you can provide.
vandykitty is offline   Reply With Quote
Old 06-30-2017, 11:45 AM   #6
vandykitty
Junior Member
 
Location: charlotte,nc

Join Date: Dec 2012
Posts: 5
Default

Ok, I figured it out, responded too soon
vandykitty is offline   Reply With Quote
Old 06-30-2017, 11:46 AM   #7
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,212
Default

Quote:
Originally Posted by vandykitty View Post
Thanks, I did have the same problem as the OP and had the error message that there was an unexpected signal in the fast region, so I changed that threshold, resulting in unexpected ribosomal ratio, so I changed that and the error cleared but it did not give me a re-calculated RIN. Do you know if there is something else I can change... because apparently the Agilent rep did not know how to help me troubleshoot.. thank you for any help you can provide.
Yes, that is also true -- generally if you don't get a RIN score right away the program will never calculate one.

--
Phillip
pmiguel is offline   Reply With Quote
Old 07-07-2017, 10:06 AM   #8
Chinanicole
Junior Member
 
Location: Philadelphia

Join Date: Jun 2017
Posts: 8
Default

I'm also isolating total RNA from drosophila heads and my quality check involves both a bioanalyzer run and 1% agarose gel. I look for signs of rRNA degradation (smearing on gel, weird peaks on bioA) instead of relying on the RIN score.

For the 1% gel, good quality RNA should have single large band around 1200-800 bp which corresponds to the 18S unit AND the 28S unit (in drosophila, 28S migrates with the 18S: https://www.thermofisher.com/us/en/h...rna-sizes.html). In degraded samples, the band intensity around the 18S region will be diminished AND there is a smear in the region. See attached image: the sample in lane 5 is degraded, while lanes 2,3,4 and 6 are of good quality.

On a bioanalzyer: good quality RNA you'll see two peaks around 2kb corresponding to rRNA and, if you are isolating total RNA, you will also see much smaller peaks concentrated around <200 nt and between 200 and 12kb. This is the RNA. Now compare a degraded RNA sample (sample 6) to a non-degraded sample (sample 7). See all the excess peaks - those are degraded rRNA fragments? In addition, the amount of rRNA expected for the amount of starting tissue was much lower (FU) in sample 6, again indicating degradation. Note that sample 6 is 50 heads, sample 7 is 20 heads.
Attached Images
File Type: jpg sample_6.jpg (34.5 KB, 7 views)
File Type: jpg sample_7.jpg (33.0 KB, 4 views)
File Type: jpg UVP01446July42017-p1bkf3fepb37rc75lq4r9f7ck.jpg (22.3 KB, 5 views)
Chinanicole is offline   Reply With Quote
Old 07-07-2017, 10:13 AM   #9
Chinanicole
Junior Member
 
Location: Philadelphia

Join Date: Jun 2017
Posts: 8
Default

I would also go back to your data in the bioanalyzer software and zoom in on the FU axis in the electropherogram to properly visualize the small peaks corresponding to RNA like in the photos I showed above.
Chinanicole is offline   Reply With Quote
Old 07-07-2017, 03:51 PM   #10
nucacidhunter
Senior Member
 
Location: Iran

Join Date: Jan 2013
Posts: 1,013
Default

Quote:
Originally Posted by Chinanicole View Post
I'm also isolating total RNA from drosophila heads and my quality check involves both a bioanalyzer run and 1% agarose gel. I look for signs of rRNA degradation (smearing on gel, weird peaks on bioA) instead of relying on the RIN score.

For the 1% gel, good quality RNA should have single large band around 1200-800 bp which corresponds to the 18S unit AND the 28S unit (in drosophila, 28S migrates with the 18S: https://www.thermofisher.com/us/en/h...rna-sizes.html). In degraded samples, the band intensity around the 18S region will be diminished AND there is a smear in the region. See attached image: the sample in lane 5 is degraded, while lanes 2,3,4 and 6 are of good quality.
Loading on the gel is uneven making comparisons difficult. From the gel I think that sample in lane 4 is degraded and it is difficult to comment on lane 5 due to lower loading.

Edit: lane counting from samples not the ladder

Last edited by nucacidhunter; 07-07-2017 at 03:55 PM.
nucacidhunter is offline   Reply With Quote
Old 07-10-2017, 05:19 AM   #11
Chinanicole
Junior Member
 
Location: Philadelphia

Join Date: Jun 2017
Posts: 8
Default

Samples were loaded by volume, not RNA content so yes, they are uneven. Samples were also from different amounts of starting tissue. The point was the sample in lane 5 (including the ladder or lane 4 if you ignore the ladder) is degraded.
Chinanicole is offline   Reply With Quote
Reply

Tags
bioanalyser rna quality, bioanalyzer 2100, rna bioanalyzer, rna extraction, rna integrity

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:34 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2017, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO